Hello Everyone!

I have been put in charge of maintaining our processing of raw MRS, and I haven't been able to get the files from one of our sites inputted into Tarquin.

This is the output I get when running the terminal version:
"
tarquin --input P58368.7 --format ge --output_pdf ex.pdf

2017-Nov-17 19:01:15 INFO   : TARQUIN 4.3.7 Started
2017-Nov-17 19:01:15 INFO   : Loading data file : P58368.7
2017-Nov-17 19:01:15 INFO   : Rdb reader rev : 26.002001
2017-Nov-17 19:01:15 INFO   : P-file offset  : 0
2017-Nov-17 19:01:15 INFO   : Frame size : -57344
2017-Nov-17 19:01:15 INFO   : Offset to data : -57344
2017-Nov-17 19:01:15 INFO   : Nechoes : 0
2017-Nov-17 19:01:15 INFO   : Water ref frames : 1.000000
error loading FID: Unrealistic number of reference frames, check data type.

2017-Nov-17 19:01:15 INFO   : TARQUIN Finished
"

I created a small video showing what happens when I try through the GUI, though, admittedly, it's not very informative. Both GE files that I opened in the video came from the same machine, a GE Sigma 750W, except both acquisitions were separated by a few months. I get the same error output when I try to run through any of the new acquisitions.

Any thoughts on possible solutions? I am new to processing MRS through Tarquin, so any guidance at all would be greatly appreciated!

Thank you so much for your time,

Best

John Duncan

Matt Lyon

unread,
Jan 11, 2018, 5:13:33 AM1/11/18
to TARQUIN users' group
Hi Duncan,

The problem is due to the header of the GE P-file being changed in DV26.
The line below refers to which version.


2017-Nov-17 19:01:15 INFO   : Rdb reader rev : 26.002001

Unfortunately I am experiencing this exact issue, and have not found a way to fix this. At the moment I am trying to find a way to edit the header of the P-file to a version that was working with TARQUIN such as DV24.
Will update when I have found a fix

Regards

Matt

Martin

unread,
Jan 14, 2018, 10:31:01 AM1/14/18
to TARQUIN users' group
Hi both,

I can confirm GE have changed the format of their p-files. I haven't been able to figure out the winning combination of new offsets yet - it looks like a more significant change than usual...

Martin

Nicole V

unread,
Feb 8, 2018, 7:48:36 PM2/8/18
to TARQUIN users' group
Hi Martin,

Do you have any updates on this issue? I have been trying to process p-files from a GE Discovery MR750 3T system with no luck.

Thank you!

Nicole 

Martin Wilson

unread,
Feb 9, 2018, 4:22:53 AM2/9/18
to TARQUIN users' group
Hi Nicole,

I'm about 90% sure this is fixed in the source code, but no release yet. Are you using the Linux binary version of TARQUIN?

Martin

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.
To post to this group, send email to tarquin_users_group@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

Nicole V

unread,
Feb 9, 2018, 10:42:18 AM2/9/18
to TARQUIN users' group
Hi Martin,

Thanks for your quick reply. Yes, I am using the Linux binary version!

Thank you,

Nicole
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.

Martin Wilson

unread,
Feb 12, 2018, 8:20:03 AM2/12/18
to TARQUIN users' group
Hi Nicole,

Try this download and let me know if your problem is resolved:

https://www.dropbox.com/s/fbumo3oy5vewlpj/TARQUIN_Linux_4.3.11_RC.tar.gz?dl=0

Martin



To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_users_group@googlegroups.com.

Nicole V

unread,
Feb 13, 2018, 9:35:49 AM2/13/18
to TARQUIN users' group
Hi Martin,

Everything seems to be working! Do you by chance have an updated Mac version as well?

Thank you!

Nicole
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.

researc...@gmail.com

unread,
Feb 14, 2018, 9:55:10 PM2/14/18
to TARQUIN users' group
Hi Martin,

Just wanted to say thank you for working on this!

My group uses the Mac version currently, though we're hoping to move to the linux version relatively soon. Any chance there's an updated Mac Version as well?
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.

Martin Wilson

unread,
Feb 15, 2018, 2:58:11 AM2/15/18
to tarquin_u...@googlegroups.com
Yes, I’ll do a proper release now I know the GE issue is resolved.

Martin


To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_g...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_g...@googlegroups.com.

To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_g...@googlegroups.com.

Martin Wilson

unread,
Feb 21, 2018, 5:51:54 AM2/21/18
to TARQUIN users' group
Mac version now available from here:

https://sourceforge.net/projects/tarquin/files/TARQUIN_4.3.11/

Martin

To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.

To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_users_group@googlegroups.com.

researc...@gmail.com

unread,
Feb 26, 2018, 2:25:46 PM2/26/18
to TARQUIN users' group
Hey Martin, 

sorry to bring bad news, but I'm still getting an error. 

Here is the output:
`
2018-Feb-26 11:22:46 INFO   : TARQUIN 4.3.11 Started
2018-Feb-26 11:22:46 INFO   : Loading data file : P33792.7
2018-Feb-26 11:22:46 INFO   : Rdb reader rev : 26.002001
2018-Feb-26 11:22:46 INFO   : P-file offset  : 213684
2018-Feb-26 11:22:46 INFO   : Frame size : 32768
2018-Feb-26 11:22:46 INFO   : Offset to data : 246452
2018-Feb-26 11:22:46 INFO   : Nechoes : 1
2018-Feb-26 11:22:46 INFO   : Water ref frames : 2.000000
2018-Feb-26 11:22:46 INFO   : Water sup frames : 8.000000
2018-Feb-26 11:22:46 INFO   : Number of coils : 12
2018-Feb-26 11:22:46 INFO   : x dim : 1
2018-Feb-26 11:22:46 INFO   : y dim : 1
2018-Feb-26 11:22:46 INFO   : z dim : 1
2018-Feb-26 11:22:46 INFO   : nStart = 50
2018-Feb-26 11:22:46 INFO   : nEnd   = 2048
2018-Feb-26 11:22:46 INFO   : Averaging metabolite FIDs
2018-Feb-26 11:22:46 INFO   : Averaging across 8 of 8 FIDs
TARQUIN Version 4.3.11

**** TARQUIN is a research tool and is NOT for clinical use. ****

2018-Feb-26 11:22:46 INFO   : Rdb reader rev : 26.002001
2018-Feb-26 11:22:46 INFO   : P-file offset  : 213684
2018-Feb-26 11:22:46 INFO   : Frame size : 32768
2018-Feb-26 11:22:46 INFO   : Offset to data : 246452
2018-Feb-26 11:22:46 INFO   : Nechoes : 1
2018-Feb-26 11:22:46 INFO   : Water ref frames : 2.000000
2018-Feb-26 11:22:46 INFO   : Water sup frames : 8.000000
2018-Feb-26 11:22:46 INFO   : Number of coils : 12
2018-Feb-26 11:22:46 INFO   : x dim : 1
2018-Feb-26 11:22:46 INFO   : y dim : 1
2018-Feb-26 11:22:46 INFO   : z dim : 1
2018-Feb-26 11:22:46 INFO   : Averaging water ref FIDs
2018-Feb-26 11:22:46 INFO   : Number of W fids does not match WS fids.
2018-Feb-26 11:22:46 INFO   : Averaging accross 2 water reference fids.
Parameters Before Preprocessing
-------------------------------
Input file:                 P33792.7
Water reference file:       P33792.7
Data points:                4096
Rows:                       1
Columns:                    1
Slices:                     1
Echo time:                  1606.39 s
Sampling frequency:         5000 Hz
Transmitter frequency:      1.27828e+08 Hz
Starting sample:            50
Ending sample:              2048
Removing signals in:        [-45,+45] Hz
Autophasing yes/no?:        yes
Water eddy correction?:     no
Autoreferencing yes/no?:    yes
Reference:                  4.65 ppm


2018-Feb-26 11:22:46 INFO   : Starting 'Preprocessing'
2018-Feb-26 11:22:46 INFO   : Preprocessing fid 1 of 1
2018-Feb-26 11:22:46 INFO   : Starting 'autophasing mag method water'
2018-Feb-26 11:22:46 INFO   : Finished 'autophasing mag method water'
2018-Feb-26 11:22:46 INFO   : Ref start = 4.650000 ppm
2018-Feb-26 11:22:46 INFO   : Ref new = 4.651194 ppm
2018-Feb-26 11:22:46 INFO   : Residual water freq. (Hz) : 0.152588
2018-Feb-26 11:22:46 INFO   : Starting 'building Hankel LP matrix'
2018-Feb-26 11:22:46 INFO   : Finished 'building Hankel LP matrix'
2018-Feb-26 11:22:46 INFO   : Starting 'SVD of LP matrix'
2018-Feb-26 11:22:47 INFO   : Finished 'SVD of LP matrix'
2018-Feb-26 11:22:47 INFO   : Starting 'estimating LP parameters'
2018-Feb-26 11:22:47 INFO   : Finished 'estimating LP parameters'
2018-Feb-26 11:22:47 INFO   : Starting 'estimating LP amplitudes'
2018-Feb-26 11:22:47 INFO   : Finished 'estimating LP amplitudes'
2018-Feb-26 11:22:47 INFO   : Starting 'autophasing (new)'
2018-Feb-26 11:22:47 INFO   : Finished 'autophasing (new)'
2018-Feb-26 11:22:47 DEBUG  : Max Phi0 : 0.00
2018-Feb-26 11:22:47 INFO   : Ref start = 4.650000 ppm
2018-Feb-26 11:22:47 INFO   : Ref new = 4.679842 ppm
2018-Feb-26 11:22:47 DEBUG  : Width pts     : 60.41
2018-Feb-26 11:22:47 DEBUG  : Width Hz est  : 6.15
2018-Feb-26 11:22:47 DEBUG  : Init beta est : 94.24
2018-Feb-26 11:22:47 INFO   : Ymax       = 8111561.067328
2018-Feb-26 11:22:47 INFO   : sdev noise = 204130.557947
2018-Feb-26 11:22:47 INFO   : SNR guess  = 19.8686
2018-Feb-26 11:22:47 INFO   : Finished 'Preprocessing'

Parameters After Preprocessing
------------------------------
Input file:                 P33792.7
Water reference file:       P33792.7
Data points:                4096
Rows:                       1
Columns:                    1
Slices:                     1
Echo time:                  1606.39 s
Sampling frequency:         5000 Hz
Transmitter frequency:      1.27828e+08 Hz
Starting sample:            50
Ending sample:              2048
Removing signals in:        [-45,+45] Hz
Autophasing yes/no?:        yes
Water eddy correction?:     no
Autoreferencing yes/no?:    yes
Reference:                  4.67984 ppm


2018-Feb-26 11:22:47 INFO   : Simulating using 12 threads.
2018-Feb-26 11:22:48 INFO   : All simulation threads stopped.
2018-Feb-26 11:22:48 ERROR  : incorrect echo time (1606.39 in FID does not match 0.00 in basis)
2018-Feb-26 11:22:48 INFO   : TARQUIN Finished

ERROR:   Basis set and fid parameter mismatch, please resimulate
`
And this is for both the Mac and Linux versions.

John
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.

Martin Wilson

unread,
Mar 2, 2018, 7:49:58 AM3/2/18
to TARQUIN users' group
Hi,

Thanks for the info. Can you work around it by manually setting the echo time (this is the real problem I think)?

Also, please send me the data if you're able.

Martin

To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.

To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.

To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_users_group@googlegroups.com.

Dave Lythgoe

unread,
Mar 2, 2018, 1:35:31 PM3/2/18
to tarquin_u...@googlegroups.com
Hi Martin,

I've just tried this with some DV25 data, which previously worked with the last version of Tarquin. It incorrectly reads the TE at 1.6 s, where it should have been 0.035 s. As you suggested, manually typing in the TE works fine. 


Cheers, 
Dave

To post to this group, send email to tarquin_users_group@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

Martin

unread,
Mar 12, 2018, 11:00:19 AM3/12/18
to TARQUIN users' group
Hi,

Thanks for the bug report, I've just tested some other data RDB_26.002001 that finds the correct TE. Are you able to send me an example?

Martin

researc...@gmail.com

unread,
Apr 19, 2018, 1:46:28 PM4/19/18
to TARQUIN users' group
Hi Martin!

Any update on reading GE p-files?

Best,
John Duncan

Martin Wilson

unread,
Apr 24, 2018, 4:24:06 AM4/24/18
to TARQUIN users' group
Hi John,

The data you sent me (TE=35ms) works ok with the latest version of the source code, so I'm wondering if I made a mistake with the binary release. Have you found a problem with the latest release?

Martin

Martin
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.

To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsubscribe...@googlegroups.com.

To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_users_group@googlegroups.com.

researc...@gmail.com

unread,
May 3, 2018, 10:06:21 PM5/3/18
to TARQUIN users' group
Yes, we are still having trouble reading it. Here is the output from command: tarquin --input P88064.7 --format ge --output_pdf test.pdf 

'
2018-May-03 19:02:22 INFO   : TARQUIN 4.3.11 Started
2018-May-03 19:02:22 INFO   : Loading data file : P85504.7
2018-May-03 19:02:22 INFO   : Rdb reader rev : 26.002001
2018-May-03 19:02:22 INFO   : P-file offset  : 213684
2018-May-03 19:02:22 INFO   : Frame size : 32768
2018-May-03 19:02:22 INFO   : Offset to data : 246452
2018-May-03 19:02:22 INFO   : Nechoes : 1
2018-May-03 19:02:22 INFO   : Water ref frames : 2.000000
2018-May-03 19:02:22 INFO   : Water sup frames : 8.000000
2018-May-03 19:02:22 INFO   : Number of coils : 12
2018-May-03 19:02:22 INFO   : x dim : 1
2018-May-03 19:02:22 INFO   : y dim : 1
2018-May-03 19:02:22 INFO   : z dim : 1
2018-May-03 19:02:22 INFO   : nStart = 50
2018-May-03 19:02:22 INFO   : nEnd   = 2048
2018-May-03 19:02:22 INFO   : Averaging metabolite FIDs
2018-May-03 19:02:22 INFO   : Averaging across 8 of 8 FIDs
TARQUIN Version 4.3.11

**** TARQUIN is a research tool and is NOT for clinical use. **** 

2018-May-03 19:02:22 INFO   : Rdb reader rev : 26.002001
2018-May-03 19:02:22 INFO   : P-file offset  : 213684
2018-May-03 19:02:22 INFO   : Frame size : 32768
2018-May-03 19:02:22 INFO   : Offset to data : 246452
2018-May-03 19:02:22 INFO   : Nechoes : 1
2018-May-03 19:02:22 INFO   : Water ref frames : 2.000000
2018-May-03 19:02:22 INFO   : Water sup frames : 8.000000
2018-May-03 19:02:22 INFO   : Number of coils : 12
2018-May-03 19:02:22 INFO   : x dim : 1
2018-May-03 19:02:22 INFO   : y dim : 1
2018-May-03 19:02:22 INFO   : z dim : 1
2018-May-03 19:02:22 INFO   : Averaging water ref FIDs
2018-May-03 19:02:22 INFO   : Number of W fids does not match WS fids.
2018-May-03 19:02:22 INFO   : Averaging accross 2 water reference fids.
Parameters Before Preprocessing
-------------------------------
Input file:                 P85504.7
Water reference file:       P85504.7
Data points:                4096
Rows:                       1
Columns:                    1
Slices:                     1
Echo time:                  1606.39 s
Sampling frequency:         5000 Hz
Transmitter frequency:      1.27828e+08 Hz
Starting sample:            50
Ending sample:              2048
Removing signals in:        [-45,+45] Hz
Autophasing yes/no?:        yes
Water eddy correction?:     no
Autoreferencing yes/no?:    yes
Reference:                  4.65 ppm


2018-May-03 19:02:22 INFO   : Starting 'Preprocessing'
2018-May-03 19:02:22 INFO   : Preprocessing fid 1 of 1
2018-May-03 19:02:22 INFO   : Starting 'autophasing mag method water'
2018-May-03 19:02:22 INFO   : Finished 'autophasing mag method water'
2018-May-03 19:02:22 INFO   : Ref start = 4.650000 ppm
2018-May-03 19:02:22 INFO   : Ref new = 4.664324 ppm
2018-May-03 19:02:22 INFO   : Residual water freq. (Hz) : 1.831055
2018-May-03 19:02:22 INFO   : Starting 'building Hankel LP matrix'
2018-May-03 19:02:22 INFO   : Finished 'building Hankel LP matrix'
2018-May-03 19:02:22 INFO   : Starting 'SVD of LP matrix'
2018-May-03 19:02:23 INFO   : Finished 'SVD of LP matrix'
2018-May-03 19:02:23 INFO   : Starting 'estimating LP parameters'
2018-May-03 19:02:23 INFO   : Finished 'estimating LP parameters'
2018-May-03 19:02:23 INFO   : Starting 'estimating LP amplitudes'
2018-May-03 19:02:23 INFO   : Finished 'estimating LP amplitudes'
2018-May-03 19:02:23 INFO   : Starting 'autophasing (new)'
2018-May-03 19:02:23 INFO   : Finished 'autophasing (new)'
2018-May-03 19:02:23 DEBUG  : Max Phi0 : 0.63
2018-May-03 19:02:23 INFO   : Ref start = 4.650000 ppm
2018-May-03 19:02:23 INFO   : Ref new = 4.639257 ppm
2018-May-03 19:02:23 DEBUG  : Width pts     : 23.83
2018-May-03 19:02:23 DEBUG  : Width Hz est  : 2.42
2018-May-03 19:02:23 DEBUG  : Init beta est : 7.22
2018-May-03 19:02:23 INFO   : Ymax       = 33880319.383140
2018-May-03 19:02:23 INFO   : sdev noise = 194853.587139
2018-May-03 19:02:23 INFO   : SNR guess  = 86.9379
2018-May-03 19:02:23 INFO   : Finished 'Preprocessing'

Parameters After Preprocessing
------------------------------
Input file:                 P85504.7
Water reference file:       P85504.7
Data points:                4096
Rows:                       1
Columns:                    1
Slices:                     1
Echo time:                  1606.39 s
Sampling frequency:         5000 Hz
Transmitter frequency:      1.27828e+08 Hz
Starting sample:            50
Ending sample:              2048
Removing signals in:        [-45,+45] Hz
Autophasing yes/no?:        yes
Water eddy correction?:     no
Autoreferencing yes/no?:    yes
Reference:                  4.63926 ppm


2018-May-03 19:02:23 INFO   : Simulating using 12 threads.
2018-May-03 19:02:23 INFO   : All simulation threads stopped.
2018-May-03 19:02:23 ERROR  : incorrect echo time (1606.39 in FID does not match 0.00 in basis)
2018-May-03 19:02:23 INFO   : TARQUIN Finished

ERROR:   Basis set and fid parameter mismatch, please resimulate
'


John
Martin
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.
To post to this group, send email to tarquin_u...@googlegroups.com.
Visit this group at https://groups.google.com/group/tarquin_users_group.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "TARQUIN users' group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to tarquin_users_group+unsub...@googlegroups.com.

Laura Garrison

unread,
Oct 10, 2019, 10:21:27 AM10/10/19
to TARQUIN users' group
I am also still having trouble reading the new p-file format, using the Mac OS binary version. I appreciate any help in fixing this, it seems to work fine on a colleague's linux installation, so we're wondering if this is mac-specific...

Here is the output from running the p-file with the v26 header format:

2019-Oct-10 14:22:25 INFO   : TARQUIN 4.3.11 Started
2019-Oct-10 14:22:25 INFO   : Loading data file : data/spectra/005_PROBE-SV_35_ve_front/P29184.7
2019-Oct-10 14:22:25 INFO   : Rdb reader rev : 26.002001
2019-Oct-10 14:22:25 INFO   : P-file offset  : 213684
2019-Oct-10 14:22:25 INFO   : Frame size : 32768
2019-Oct-10 14:22:25 INFO   : Offset to data : 246452
2019-Oct-10 14:22:25 INFO   : Nechoes : 1
2019-Oct-10 14:22:25 INFO   : Water ref frames : 2.000000
2019-Oct-10 14:22:25 INFO   : Water sup frames : 16.000000
2019-Oct-10 14:22:25 INFO   : Number of coils : 32
2019-Oct-10 14:22:25 INFO   : x dim : 1
2019-Oct-10 14:22:25 INFO   : y dim : 1
2019-Oct-10 14:22:25 INFO   : z dim : 1
2019-Oct-10 14:22:25 INFO   : nStart = 50
2019-Oct-10 14:22:25 INFO   : nEnd   = 2048
2019-Oct-10 14:22:25 INFO   : Averaging metabolite FIDs
2019-Oct-10 14:22:25 INFO   : Averaging across 16 of 16 FIDs
TARQUIN Version 4.3.11

**** TARQUIN is a research tool and is NOT for clinical use. **** 

2019-Oct-10 14:22:25 INFO   : Rdb reader rev : 26.002001
2019-Oct-10 14:22:25 INFO   : P-file offset  : 213684
2019-Oct-10 14:22:25 INFO   : Frame size : 32768
2019-Oct-10 14:22:25 INFO   : Offset to data : 246452
2019-Oct-10 14:22:25 INFO   : Nechoes : 1
2019-Oct-10 14:22:25 INFO   : Water ref frames : 2.000000
2019-Oct-10 14:22:25 INFO   : Water sup frames : 16.000000
2019-Oct-10 14:22:25 INFO   : Number of coils : 32
2019-Oct-10 14:22:25 INFO   : x dim : 1
2019-Oct-10 14:22:25 INFO   : y dim : 1
2019-Oct-10 14:22:25 INFO   : z dim : 1
2019-Oct-10 14:22:26 INFO   : Averaging water ref FIDs
2019-Oct-10 14:22:26 INFO   : Number of W fids does not match WS fids.
2019-Oct-10 14:22:26 INFO   : Averaging accross 2 water reference fids.
Parameters Before Preprocessing
-------------------------------
Input file:                 data/spectra/005_PROBE-SV_35_ve_front/P29184.7
Water reference file:       data/spectra/005_PROBE-SV_35_ve_front/P29184.7
Data points:                4096
Rows:                       1
Columns:                    1
Slices:                     1
Echo time:                  -272.66 s
Sampling frequency:         5000 Hz
Transmitter frequency:      1.27715e+08 Hz
Starting sample:            50
Ending sample:              2048
Removing signals in:        [-45,+45] Hz
Autophasing yes/no?:        yes
Water eddy correction?:     no
Autoreferencing yes/no?:    yes
Reference:                  4.65 ppm


2019-Oct-10 14:22:26 INFO   : Starting 'Preprocessing'
2019-Oct-10 14:22:26 INFO   : Preprocessing fid 1 of 1
2019-Oct-10 14:22:26 INFO   : Starting 'autophasing mag method water'
2019-Oct-10 14:22:26 INFO   : Finished 'autophasing mag method water'
2019-Oct-10 14:22:26 INFO   : Ref start = 4.650000 ppm
2019-Oct-10 14:22:26 INFO   : Ref new = 4.666727 ppm
2019-Oct-10 14:22:26 INFO   : Residual water freq. (Hz) : 2.136230
2019-Oct-10 14:22:26 INFO   : Starting 'building Hankel LP matrix'
2019-Oct-10 14:22:26 INFO   : Finished 'building Hankel LP matrix'
2019-Oct-10 14:22:26 INFO   : Starting 'SVD of LP matrix'
2019-Oct-10 14:22:26 INFO   : Finished 'SVD of LP matrix'
2019-Oct-10 14:22:27 INFO   : Starting 'estimating LP parameters'
2019-Oct-10 14:22:27 INFO   : Finished 'estimating LP parameters'
2019-Oct-10 14:22:27 INFO   : Starting 'estimating LP amplitudes'
2019-Oct-10 14:22:27 INFO   : Finished 'estimating LP amplitudes'
2019-Oct-10 14:22:27 INFO   : Starting 'autophasing (new)'
2019-Oct-10 14:22:27 INFO   : Finished 'autophasing (new)'
2019-Oct-10 14:22:27 DEBUG  : Max Phi0 : -0.25
2019-Oct-10 14:22:27 INFO   : Ref start = 4.650000 ppm
2019-Oct-10 14:22:27 INFO   : Ref new = 4.671506 ppm
2019-Oct-10 14:22:27 DEBUG  : Width pts     : 31.64
2019-Oct-10 14:22:27 DEBUG  : Width Hz est  : 3.22
2019-Oct-10 14:22:27 DEBUG  : Init beta est : 17.52
2019-Oct-10 14:22:27 INFO   : Ymax       = 3800616.626468
2019-Oct-10 14:22:27 INFO   : sdev noise = 26112.421192
2019-Oct-10 14:22:27 INFO   : SNR guess  = 72.7741
2019-Oct-10 14:22:27 INFO   : Finished 'Preprocessing'

Parameters After Preprocessing
------------------------------
Input file:                 data/spectra/005_PROBE-SV_35_ve_front/P29184.7
Water reference file:       data/spectra/005_PROBE-SV_35_ve_front/P29184.7
Data points:                4096
Rows:                       1
Columns:                    1
Slices:                     1
Echo time:                  -272.66 s
Sampling frequency:         5000 Hz
Transmitter frequency:      1.27715e+08 Hz
Starting sample:            50
Ending sample:              2048
Removing signals in:        [-45,+45] Hz
Autophasing yes/no?:        yes
Water eddy correction?:     no
Autoreferencing yes/no?:    yes
Reference:                  4.67151 ppm


2019-Oct-10 14:22:27 INFO   : Simulating using 8 threads.Warning, negative delay requsted!!Warning, negative delay requsted!!

Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!
Warning, negative delay requsted!!

2019-Oct-10 14:22:27 INFO   : All simulation threads stopped.
2019-Oct-10 14:22:27 ERROR  : incorrect echo time (-272.66 in FID does not match 0.00 in basis)
2019-Oct-10 14:22:27 INFO   : TARQUIN Finished

Alex

unread,
Oct 10, 2019, 10:52:58 AM10/10/19
to TARQUIN users' group
Hello Martin (and Laura, and group),

Per Laura's message, we've encountered an issue in the latest tarquin (4.3.11) reading certain recent versions of the GE P-file header inconsistently.

Quick fix, initialize te_int to 0 around lines 321 and 328 of common/fidio/CFIDReaderGE.cpp, ie:

int te_int=0;


On many contemporary architectures, the size of int is going to be 4 bytes rather than the assumed 2. The read statement is only filling the first two bytes of te_int, leaving the remaining bytes un-initialized. Often we're lucky enough that these are coincidentally zero and everything works beautifully, but there are various random cases where they're filled with junk, yielding bogus (often negative) TE values and causing everything to fail.

(I'll probably come back to you in some weeks/months with more GE header improvements, but this little fix would help greatly for now).

Best regards,
Alex.


Alex

unread,
Oct 10, 2019, 11:31:27 AM10/10/19
to TARQUIN users' group
Sorry, this is an incomplete 'fix'. Thinking about it some more, the TE is of course actually stored as a 32-bit int (for all header revisions). A 16-bit int would overflow for longer TEs.

So, it's a mystery to me why the reader was going to the trouble of doing a 2-byte read into te_int at all, instead of a 4-byte read straight into TE as with other revisions? Maybe endian issues with some obscure ancient releases? The following seems just fine for recent data (also note: fwiw, the field shifted in 25.002, not 25.001)...

common/fidio/CFIDReaderGE.cpp

319     else if ( (rdb_header_rev > 11.0) && (rdb_header_rev <= 25.001) )
320     {
321         // int te_int=0; // ARC remove
322         file.seekg(4*303, std::ios_base::beg);
323         file.read((char*)&TE, 4);
324         // TE = te_int; // ARC remove
325     }
326     else if ( rdb_header_rev > 25.001 )
327     {
328         //int te_int; // ARC remove
329         file.seekg(4*303-64, std::ios_base::beg);
330         file.read((char*)&TE, 4);
331         // TE = te_int; // ARC remove
332     }

Alex

unread,
Nov 27, 2019, 9:31:22 AM11/27/19
to TARQUIN users' group
Greetings,

For the benefit of anyone still experiencing this issue, here are some (unofficial and entirely unsupported) binaries for Linux and macOS, incorporating the fix for recent GE P-files.



I'm afraid it's command-line only for MacOS, due to archaic Qwt5 and finite patience on my part.

Hope this is useful for someone.

- Alex.

thak...@gmail.com

unread,
Oct 16, 2020, 10:33:58 AM10/16/20
to TARQUIN users' group
Dear all,

My apologies if i overlooked any response regarding similar questions. I used Tarquin for single-voxel data long ago and I think it is very user-friendly software. Recently, all data at our site is collected from MR750 3T scanners. I was trying to use GE p files as well as converted DCM files (single voxel as well as 2D CSI data) to analyze using this software. I do have issues reading the data. Looking for some help. Thank you

Sunitha

Nick Bolo

unread,
Jun 7, 2021, 9:23:42 AM6/7/21
to TARQUIN users' group
Hi Alex,

Thanks for providing the fix (tarquin.redist.macos.10.14.tar.gz), which I've just downloaded for Mac (I've got OS version 10.15.7).
I haven't tried it yet, but I have a question for you before installing.
I've got single voxel presscsi on a GE MR750 DV26.0.  In the supported version of Tarquin (V 4.3.11), the Tarquin GUI reads the TE (= 0.028 sec) correctly, but I get the same error message as in this e-mail train (incorrect echo time (-272.67 in FID does not match 0.00 in basis)) when I run in command line.  
Also, manually entering the echo time (--echo 0.028) doesn't change the error message.
I have a lot of spectra, so I am really interested in getting the command line to work, but of course the GUI is great for quickly inspecting spectra, and it seems to work.
My question is, to install the fix, where should I place all the files (tarquin, simulate, libcvm.a, etc...) that are in the "redist" folder?   
Are there separate versions of tarquin for the GUI and the command line?
(I guess that's actually 2 questions)
Thanks again.

Best,
Nick


On Wednesday, November 27, 2019 at 9:31:22 AM UTC-5 Alex wrote: