Hi Will,
good question. As far as I can see, DICOM CP-2559 "Correction of VM for
Inconsistent Data Elements in PS3.6" is still work in progress (the official
status is "Assigned"), but I haven't seen any progress since June last year.
The person assigned to this CP, i.e., the member of WG-06, is David Clunie, so
I'll send him an email about it.
Regards,
Jörg
-------------------------------------------------------------
Subject: Re: [DICOM Forum] Re: UR VR and VM Inconsistency in DICOM Standards
Date: Monday, April 13, 2026, 7:48:16 PM CEST
From: William Sugarman <
will.s...@gmail.com>
To: DICOM Forum <
dicom...@googlegroups.com>
Hey David,
Was there any update within WG6 on these inconsistencies? I see them still
present in the latest 2026b. This includes:
- (0010,0047) Sex Parameters for Clinical Use Category Reference
- VR = UR and VM = 1-n
- (0014,4106) Wave Dimension Values Sequence
- VR = SQ and VM = 1-n
- (0014,6025) Data Streaming Protocol
- VR = ST and VM = 1-n
All three VRs explicitly forbid a VM != 1 in 6.4 Value Multiplicity (VM)
and Delimitation
<
https://dicom.nema.org/medical/dicom/current/output/chtml/part05/
sect_6.4.html>
.
-------------------------------------------------------------
--