Electroporation Wet Lab Style
Three Pieces to the Puzzle 1. Power Supply- Model 400 2. Invitrogen Electroporator 3. Oscilloscope
Power Supply Set Up Volts- 500v -modulating the voltage dial will result in faster or slower charing of the Electroporator Watts- 20 Watts -dial 12 o’clock Amps- 50mAmps -dial 3 o’clock
Make sure that the wiring diagram and parameters are set correctly Turn Power On DC -Push start button. Hold briefly. If it does not stay on something is not set up properly
Electroporation Charge Flip switch to CHARGE Load resistance- 500 Ohms Capacitance- 660v Armed- to pulse When charge light stays on (see notes if light does not go solid within 1 minute) Verify the safety circuit is closed Set trigger on Oscilloscope Flip switch to PULSE
Notes- When charging light is blinking the battery is charging. When light stays on it is fully charged. If you do not trigger immediately when fully charged. Charge will bleed off and will not fully charge without discharge. This can be regulated by speeding up or slowing down the charge with the voltage dial on the power supply.
GwInstek Oscilloscope GDS-820C 1. Trigger -Push {Menu} Set TYPE to PULSE -Use {F1} to change Set SOURCE to CHANNEL BEING USED -Use {F2} to change Set MODE to SINGLE -Use {F3} to change 2. Set {CH1} to 200mV using Yellow Volts/Div Knob 3. Set TRIGGER Level Knob to 200mV 4. Set HORIZONTAL TIME/DIV knob to 50mS 5. Push RUN/STOP to RESET TRIGGER