Hi Jeong Hwan Kim,
Another good reason to attend our VNA Tools training course and learn more about the implications of using only the polynomial definitions of the calibration standards 😊
There is no method for directly converting a calibration kit file from Keysight, Rohde & Schwarz, or Anritsu. And personally, I think it doesn't make sense, because those calibration kit files typically don't contains any information about the complete uncertainty and correlation information of each standard.
It is therefore better
to first have your calibration kit characterized by a National Metrology Institute
(e.g., METAS) or a calibration laboratory that can provide sdatb files for each
standard or a scolcv file for the complete calibration kit.
In the meantime, several cal kit manufacturers offer now their definitions using the sdatb
or sdatcv files or at least by request.
You
should also check whether the important systematic connector effects are
included in their traceability chain. Especially since many NMIs do not yet
take these effects into account. Not even in the declared CMC uncertainties!
As a work around, VNA Tools offers in the Database to create under 'Calibration Standard / New Standard' an 'Agilent Model Standard'. This allows to enter the polynomial definitions based on the Keysight model and to add electrical specifications.
Best
Regards,
Juerg