Instrument ID Query fails.... N5230C & E8362C

125 views
Skip to first unread message

cca...@gmail.com

unread,
Dec 7, 2020, 1:29:45 PM12/7/20
to VNA Tools
Hello,

If I use a USB GPIB adaptor I have no problems. 

If I try to connect over ethernet, I always get the ID Query fails. I have no problems sending commands through test panels, LV, or .NET apps. 

Any ideas?
-Chris


1.  viOpenDefaultRM (0x00001001)
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7816      Call Duration 00:00:00.0000
Status: 0 (VI_SUCCESS)

2.  viParseRsrcEx (0x00001001, "TCPIP0::A-E8362C-20804::inst0::INSTR", 6 (0x6), 0 (0x0), "INSTR", "TCPIP0::A-E8362C-20804::inst0::INSTR", "A-E8362C-20804")
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7816      Call Duration 00:00:00.0020
Status: 0 (VI_SUCCESS)

3.  viParseRsrcEx (0x00001001, "TCPIP0::A-E8362C-20804::inst0::INSTR", 6 (0x6), 0 (0x0), "INSTR", "TCPIP0::A-E8362C-20804::inst0::INSTR", "A-E8362C-20804")
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7856      Call Duration 00:00:00.0020
Status: 0 (VI_SUCCESS)

4.  viOpenDefaultRM (0x00001007)
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7876      Call Duration 00:00:00.0000
Status: 0 (VI_SUCCESS)

5.  viOpen (0x00001007, "TCPIP0::A-E8362C-20804::inst0::INSTR", 0 (0x0), 2000 (0x7D0), 0x00000006)
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7876      Call Duration 00:00:00.0099
Status: 0 (VI_SUCCESS)

6.  viInstallHandler (TCPIP0::A-E8362C-20804::inst0::INSTR (0x00000006), 0x3FFF2009 (VI_EVENT_IO_COMPLETION), 0x00000281F186A62C, NULL)
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7976      Call Duration 00:00:00.0000
Status: 0 (VI_SUCCESS)

7.  viClose (0x00001001)
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7976      Call Duration 00:00:00.0010
Status: 0 (VI_SUCCESS)

8.  viWrite (TCPIP0::A-E8362C-20804::inst0::INSTR (0x00000006), "*IDN?", 5 (0x5), 5 (0x5))
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.7986      Call Duration 00:00:00.0019
Status: 0 (VI_SUCCESS)

9.  viRead (TCPIP0::A-E8362C-20804::inst0::INSTR (0x00000006), "Agilent Technologies,E8362C,MY43020804,A.09.42.22.", 16384 (0x4000), 50 (0x32))
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.8006      Call Duration 00:00:00.0000
Status: 0 (VI_SUCCESS)

10.  viSetAttribute (TCPIP0::A-E8362C-20804::inst0::INSTR (0x00000006), 0x3FFF001A (VI_ATTR_TMO_VALUE), 3000 (0x0000000000000BB8))
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:13.8006      Call Duration 00:00:00.0000
Status: 0 (VI_SUCCESS)

> 11.  viEnableEvent (TCPIP0::A-E8362C-20804::inst0::INSTR (0x00000006), 0x3FFF200B (VI_EVENT_SERVICE_REQ), 1 (0x1), 0 (0x0))
> Process ID: 0x00000F78         Thread ID: 0x000032B4
> Start Time: 13:20:13.8006      Call Duration 00:00:04.0061
> Status: 0xBFFF0015 (VI_ERROR_TMO)

12.  viStatusDesc (TCPIP0::A-E8362C-20804::inst0::INSTR (0x00000006), 0xBFFF0015 (VI_ERROR_TMO), "Timeout expired before operation completed.")
Process ID: 0x00000F78         Thread ID: 0x000032B4
Start Time: 13:20:17.8068      Call Duration 00:00:00.0010
Status: 0 (VI_SUCCESS)

Michael Wollensack METAS

unread,
Dec 8, 2020, 2:23:54 AM12/8/20
to VNA Tools
Hi,

If it works over GPIB and not over Ethernet then it could be a Firewall issue.

Starting from METAS VNA Tools V2.3.2 the installer adds an rule for METAS VNA Tools to allow VXI-11 with service requests.

If you just use NI MAX to communicate your are not using SRQs.

I'll sent you the newest version of VNA Tools in a private mail.

Please let me know if it solved the problem.

Regards
Michael

Chris C

unread,
Dec 9, 2020, 10:51:03 AM12/9/20
to VNA Tools
Hi Michael,

The new version solved my issues connecting via Ethernet, thank you! 

Somewhat strange was our N5230C had no issues with using the default LAN Device name, however, our E8362C had to be changed to "hpib7,16". 

Our PXI P9371A will not work, also with a ID Query Failed error. I don't believe you support this device based on the PNA driver code but I had thought that the P9XXXX and P5XXXX are on the same FW, I do not have a P5XXXX to test but was thinking of modifying the code to allow the P9XXXX to see if it will work. Any thoughts on this?

if (idn.IndexOf("Agilent Technologies,E836") < 0 &&
                idn.IndexOf("Agilent Technologies,N52") < 0 &&
                idn.IndexOf("Keysight Technologies,N52") < 0 &&
                idn.IndexOf("Keysight Technologies,E508") < 0 &&
                idn.IndexOf("Keysight Technologies,M937") < 0 && // PXI over HiSLIP
                idn.IndexOf("Keysight Technologies,M948") < 0 && // PXI over HiSLIP
                idn.IndexOf("Keysight Technologies,M980") < 0 && // PXI over HiSLIP
                idn.IndexOf("Keysight Technologies,P50") < 0) // USB over HiSLIP
            {
                throw new System.Exception("Instrument identification query failed.");
            }

Michael Wollensack METAS

unread,
Dec 10, 2020, 3:50:09 AM12/10/20
to VNA Tools
Hi,

> Somewhat strange was our N5230C had no issues with using the default LAN Device name, however, our E8362C had to be changed to "hpib7,16"

Please ask Keysight why

> Our PXI P9371A will not work, also with a ID Query Failed error.

I've added the P937 series to the PNA driver. I'll send you another private email with the updated driver dll (just replace the dll under 'C:\Users\Public\Documents\Metas.Instr\Drivers').

Please report back. If it works it will be part of VN Tools V2.4.3.

General remark, please use HiSLIP and not VXI-11 when it's possible, see https://scdn.rohde-schwarz.com/ur/pws/dl_downloads/dl_application/application_notes/1ma208/1MA208_2e_HiSLIP.pdf  

Regards
Michael

Chris C

unread,
Dec 14, 2020, 9:08:29 AM12/14/20
to VNA Tools
Michael,

Didn't see the update and had already modified the driver to work. I will replace with the one you e-mailed me to make sure I am using your code and not modified code. 

Initial testing looks promising; no errors or warnings. Hopefully later this week I will find time to go through an entire calibration including noise characterization of the VNA to verify all functions. 

Thank you very much for the support! Hoping to visit METAS again in 2021. 

-CC

Reply all