Hi,
it depends in which calibration model (switched vs generic) you are.
VNA Switched Calibration Model (only used for SOLT calibration)

Switched Error Terms 1 (Forward Error Terms, Source switched to port 1)
| E11: Directivity Port 1 | | E13: Reflection Tracking Port 1 | |
| E21: Crosstalk from Port 1 to Port 2 | | | E24: Transmission Tracking from Port 1 to Port 2 |
| E31: 1 | | E33: Source Match Port 1 | |
| | | | E44: Load Match Port 2 |
Switched Error Terms 2 (Reverse Error Terms, Source switched to port 2)
| | E12: Crosstalk from Port 2 to Port 1 | E13: Transmission Tracking from Port 2 to Port 1 | |
| | E22: Directivity Port 2 | | E24: Reflection Tracking Port 2 |
| | | E33: Load Match Port 1 | |
| | E42: 1 | | E44: Source Match Port 2 |
VNA Generic Calibration Model (used for all other calibration)
Generic Error Terms
| E11: Directivity Port 1 | E12: Crosstalk from Port 2 to Port 1 | E13: Tracking | |
| E21: Crosstalk from Port 1 to Port 2 | E22: Directivity Port 2 | | E24: Tracking / Symmetry |
| E31: 1 | | E33: Match Port 1 | |
| | E42: Symmetry | | E44: Match Port 2 |
The generic error terms can be converted using the switch terms to the switched error terms. This can be used for comparison reasons.
Hope this helps.
Regards
Michael