Hi,

We would like measure the input reflection of the spectrum analyser.
Do you have any idea how can we synchronize spectrum analyser with VNA?

Best Regards,
László

Michael Wollensack METAS

unread,
Mar 16, 2026, 9:35:39 AM (9 days ago) Mar 16
to VNA Tools
Hi,

what kind of spectrum analyzer do you want to measure?

For a R&S FSMR and FSW there are driver in VNA Tools.

Measurement steps:

  1. VNA Settings (setup frequency sweep)
  2. Measurement / Spectrum Analyzer / Settings (this will read out the VNA frequency list and set up the SA)
  3. Measurement / Spectrum Analyzer / Measurement (this will measure each frequency in CW mode and the frequency is set to the SA)
Regards
Michael

sle...@gmail.com

unread,
Mar 16, 2026, 10:45:42 AM (9 days ago) Mar 16
to VNA Tools
Hi,

R&S FSMR, it is good news, thank you for the information. 

In the setting the OK button is grey. 
Do we have to create a calibration standard for the SA, what type SA Table or SA database?

Regards.
László

Michael Wollensack METAS

unread,
Mar 16, 2026, 10:52:52 AM (9 days ago) Mar 16
to VNA Tools
Hi,

yes you have to add it to the database. See Test\SpectrumAnalyzer\TestSpectrumAnalyzerTable.calstd as an example.

Regards
Michael

sle...@gmail.com

unread,
Mar 18, 2026, 8:59:55 AM (7 days ago) Mar 18
to VNA Tools
Hi,

We managed to measure the S11 of the FSMR with Agilent E8363C in 20 MHz – 40 GHz frequency range.

Now we are measuring the lower frequency range 100 kHz – 160 MHz with the Agilent E5061B, we have problems with this frequency range.

Warnings, for example:
Warning: SA sweep (1245 ms) must be complete before the VNA sweep (666 ms). Increase VNA sweep points.

Error message, connection problem:
Error: Egy vagy több hiba történt.
   a következő helyen: System.Threading.Tasks.Task.FastWaitAll(Task[] tasks)
   a következő helyen: System.Threading.Tasks.Parallel.Invoke(ParallelOptions parallelOptions, Action[] actions)
   a következő helyen: Metas.Instr.Driver.SpectrumAnalyzer.SpectrumAnalyzerHelper.MeasureSingleFrequencyPoint(ISpectrumAnalyzer sa, IVna vna, Double freq, Int32 delay, Int32 period, Boolean checkTimming, Double[]& saData, VnaData`1& vnaData, Int64& tSA, Int64& tVNA, Int64& tTotal)
   a következő helyen: Metas.Instr.Driver.SpectrumAnalyzer.SpectrumAnalyzerHelper.MeasureAllFrequencyPointsSAandVNA(Double[][]& saData, VnaData`1[]& vnaData, BackgroundWorker worker, DoWorkEventArgs e, String reportMessage)
   a következő helyen: Metas.Vna.Journal.MeasurementSpectrumAnalyzerSeries.Start(BackgroundWorker worker, DoWorkEventArgs e)
   a következő helyen: System.ComponentModel.BackgroundWorker.OnDoWork(DoWorkEventArgs e)
   a következő helyen: System.ComponentModel.BackgroundWorker.WorkerThreadStart(Object argument)
Error: The connection for the given session has been lost.
   a következő helyen: NationalInstruments.Visa.Internal.ErrorChecker.CheckStatusAndThrow(Int32 status, SafeVisaHandle visaObject)
   a következő helyen: NationalInstruments.Visa.Internal.MessageBasedSessionImpl.Write(Byte* buffer, Int64 index, Int64 count)
   a következő helyen: NationalInstruments.Visa.Internal.MessageBasedSessionImpl.Write(Byte[] buffer, Int64 index, Int64 count)
   a következő helyen: Metas.Instr.VisaNS.MessageBasedSessionNS.Query(String dataToWrite)
   a következő helyen: Metas.Instr.Driver.SpectrumAnalyzer.Rohde_Schwarz_FSMR.TriggerSingleStart() hely: D:\svn_work\217.01 S-Parameter\Software\Metas.Instr\source\Metas.Instr.Driver.SpectrumAnalyzer\RohdeSchwarz_FSMR.cs, sor: 138
   a következő helyen: Metas.Instr.Driver.SpectrumAnalyzer.SpectrumAnalyzerHelper.<>c__DisplayClass51_0.<MeasureSingleFrequencyPoint>b__0()
   a következő helyen: System.Threading.Tasks.Task.Execute()

I attached the copy of console.

Regards.
László
errors_E5061B.txt

Michael Wollensack METAS

unread,
Mar 18, 2026, 9:18:01 AM (7 days ago) Mar 18
to VNA Tools
Hi, 

try to increase the VNA seep points per frequency. So that a VNA sweep at one frequency takes longer than the SA sweep.

Regards
Michael

sle...@gmail.com

unread,
Mar 18, 2026, 11:20:24 AM (7 days ago) Mar 18
to VNA Tools
Hi,

We increased the VNA sweep points to 48 and now it  works fine.
Thank you for the help.

Regards
László