Hi, I am posting an explanation from the Clinical Education Manager of Newport Medical. Thanks Cyndy!
"Active circuit" implies that there is an
exhalation valve in the circuit and that valve is used in conjunction with flow
delivery to control inspiration/exhalation of breath delivery and hold PEEP
pressure in the circuit in between breaths if the machine is set to deliver
PEEP. If a mask is used in place of a
trach tube, the mask is usually required to be non-vented (usually blue elbow).
The mask should be relatively leak free and not contain an anti suffocation
valve. Pressure and volume ventilation are used interchangeably with this kind
of system and they are used with or without PEEP.
"Passive circuit" implies that there is just a
length of tubing between the machine and the patient and no exhalation
valve. The system must be vented so
extra gas that is delivered from the machine can escape and also, most
importantly, so the patient can exhale.
The venting aspect of this system is usually accomplished with some kind
of swivel or elbow that contains holes.
If a mask is used, the mask might also contain holes. Also there is usually some kind of anti
suffocation valve so the patient can still breathe if gas delivery stops (not
needed on the active system because the exhalation valve serves this
purpose). Inspiration/exhalation is
managed by regulating the gas flow to the patient. Pressure (like PEEP) is held in the system in
between breaths to make sure that the carbon dioxide that the patient exhales
will go out through the vents and not be re-inhaled. It is most common to use pressure ventilation
with this kind of system and it is always with PEEP. Inspiratory pressure and expiratory pressure
are usually called IPAP (inspiratory positive airway pressure) and EPAP
(expiratory positive airway pressure).