Procedure to Upgrade ACUSON SEQUOIA C512 v9.51 to v12.2

1,266 views
Skip to first unread message

UltraSound

unread,
Apr 14, 2010, 2:46:50 PM4/14/10
to ultrasoundsolutions
Procedure to Upgrade ACUSON SEQUOIA C512 v9.51 to v12.2

System Hardware are the same – NO CHANGE

A. Backup the Preset Systems to MO disk
B. Upgrade Acuson Sequoia C512 from v9.51 to v12.2
C. Restore the Systems from MO disk


A. Backup the Preset Systems to MO disk
1. Power up the Acuson Sequoia C512
2. Verify if your Acuson Sequoia C512 is running with version v9.51
3. Insert blank 230MB of MO disk into MO drive
4. While System Power up, Flip down the MultiHz key while hold down
the S key for about 5 seconds and release both keys.
5. User Maintenance Tools will be display on the screen.
6. Select Data Backup (first icon on the screen)
7. Click Format MO
8. Click OK to format MO disk
9. Click OK again to confirm
10. Format complete press OK to continue
11. Select Acuson FAS Backups
System Data Key Stroke Log Auto Doppler
Calc Data Exam Preset System Configuration
Software maintenance Prog Annotation Feature Control

Select all to backup to MO disk

12. Type in backup name in the box: Sequoia 951 ( any name or leave it
blank )
13. Click OK
14. Click OK to confirm Backup – The System will backup 9 selections
15. Wait for a while for the System to Backup, When the System Backup
complete
16. Click OK to continue , It will bring you back to User Maintenance
Tools
17. Select Network Setup and print out all Network configuration, ie.
Hostname Address Type
Sequoia 127.0.0.1 Network Host
SEQ61199 157.266.12.101 Hostname
SIMPLEDICOM 157.266.12.125 Dicom Server
( The Hostname, Address, and Type parameters depend on whatever you
setup for your systems to communicate with the Dicom Server )
18. Select the 2nd line of your Network setup ie. In this case is
SEQ61199
19. Click Edit
20. Click OK
21. Press Print key to print out the Network Information screen
22. Do the same for the 3rd line and print out the Network Information
screen
23. Click Prior
24. Click Eject MO
25. Click Quit
26. Click OK to Reboot

B. Upgrade Acuson Sequoia C512 from v9.51 to v12.2
1. Insert the upgrade MO disk v12.2 into MO drive
2. Power on the system
3. Flip down the Gate key while hold down the U key for about 5
seconds
4. The system start reading the upgrade MO disk and starting Software
installer; On the screen will show
Sequoia Software Installer 12.2
Sequoia Software Installation
Sequoia Hard Disk contains release: 9.51
Install Media Contains release : 12.2
5. Verbose Output, Select Yes
6. Select Initial Install
7. Click OK tab
8. Click OK to confirm ( it will format the Hard Disk and Loading new
version 12.2 software to Hard Disk) Wait until it finished loading
the new version 12.2 software ( This process will take awhile, The
system will stop several times, Just let it run until it finished
upgrade )
9. Click OK to reboot the system


C. Restore the Preset Systems from MO disk
1. When system reboot, flip down MultiHz key while hold down S key for
about 5 seconds and release
2. on the screen will display User Maintenance Tools
3. Select Data Restore icon
4. Insert Backup MO disk
5. Click OK to begin
6. Select each file to restore ( Restore System, Calc, AutoD, Config,
ExamPr, FrCtrl, ProAnn )
7. Eject MO disk after DONE restore
8. Restart the System

EOF

UltraSound

unread,
Apr 20, 2010, 12:00:54 PM4/20/10
to ultrasoundsolutions

Note: To Copy User Presets
Press Setup key  Select Service UI  Click Data Backup  Select
Acuson FAS Backups  Select Exam Preset Copy  Click OK  OK  inside
System Disk box pane all of the User Presets are listed  Highlight
User Presets to copy  Click Right Arrow button – this will copy User
Presets from system to MO disk  Click OK to make a copy
--
Subscription settings: http://groups.google.com/group/ultrasoundsolutions/subscribe?hl=en

UltraSound

unread,
May 4, 2010, 6:57:48 PM5/4/10
to ultrasoundsolutions
5/4/2010

How To Clone a Sequoia 512 system from v9.51 to v12.221

1)Make sure to print out all Features of the unit to be cloned from.
And make all FAS system backup to the MO disk.
2) Take out the original Hard Disk and put aside. Put in a New HD on
the cloned unit. Plug in all cables to HDD assy. Turn on the system
and hold S+Gate toggle switch up and release as soon as the screen
flashes
3) Perform Initial Install. Select Yes to Verbose Output. Wait until
it finished loading the new version 12.221 software (This process will
take awhile, The system will stop several times, Just let it run until
it finished upgrade). Click OK to reboot the system.
4) Now, restore all original Features on the new Hard Disk. Make sure
everything is working… (features, color, calcs …)

Alternative ( Step 1-4 ) – Restore image of Sequoia 512 v12.221
(Sequoia 512 v12.221 MachineID_64696.tib) to New Hard Disk ( 50 pin
SCSI 18GB) and put the restored Hard Disk to Sequoia

5) Turn it OFF. Put in the RDP5 board from the cloning unit into this
cloned unit. Turn it ON and let it reprogramming the new serial number
onto this board until it fully booted up ( keep pressing Set up key
to make sure). Replace the RDP5 board to their original location.
6) Put in the PIC2 board from the cloning unit into this cloned unit.
And turn the system ON and let it re-programming the PIC2 completely
as the RDP5 board. Replace the PIC2 board to their original location.

Now , you should turn ON the cloning unit and check to see everything
is working as the cloned unit ( sn#, features…). Run & compare to see
anything works the same in between the 2 Sequoia units.

Side notes: - The cloning involves 3 parts: HD, RDP5 and PIC2. If you
have an extra set of these 3 parts. You can clone a full featured
system and keep the set to be used on a next cloning system.

- DON’T try to put both RDP5 and PIC2 board in and power up at the
Same time ???


Note: Hard Disk locate next to the MO Drive, Unscrew 2 screws on the
back and lift the back cover up. PIC2 board – Unplug all of the cable
on the back of the Sequoia and unscrew all screw to remove the PIC2
board.
Test the feature, image, color, … each time after cloned the RDP5 and
PIC2, and go inside the Sequoia system to do the extensive board test
by using laptop to connect to Sequoia.

EOF

=============================================================================================

On Apr 14, 11:46 am, UltraSound <ultrasound2...@gmail.com> wrote:

UltraSound

unread,
May 4, 2010, 7:01:38 PM5/4/10
to ultrasoundsolutions

===============================================================================

Machine ID: 64696 Premium Maintenance Toos 04 May 2010
version 12.221 03:40:04 pm

Hardware Diagnostic Detail Log Review

SLOT NAME PART NAME REVISION PART NUMBER SERIAL NUM

4DA empty empty empty empty
AAM aam2 44772
AUX aux xe 41382 4397
BDM bdm xk 32062 1830
BF1 bf4 b 52282 21544
BF2 bf4 xb 52282 0772
CN cn3 a 39522 21255
CSD csd2 a 41462 21294
FIZ fiz xl 35992 25675
FPP fpp xc 31642 empty
IOE empty empty empty empty
IOV iov2 a 41482 20728
MDI empty empty empty empty
MX mx2r xm 36262 1640
PIC pic2 a 43242 21053
PPC empty empty empty empty
PPS pps3 empty 42526 empty
PVI empty empty empty empty
RDP rdp5 a 53552 21151
RX rx4r a 51562 28430
TX1 tx3 xb 39142 1440
TX2 tx3 xb 39142 1492


EOF --- Boards installed inside the Acuson Sequoia 512 Ultrasound
System Machine ID: 64696 Software Version 12.221

===============================================================================