Load new Options on a HDI 3000, 3500, 5000

119 views
Skip to first unread message

V T

unread,
Apr 14, 2010, 2:36:10 PM4/14/10
to ultrasoundsolutions


Load new Options on a HDI 3000, 3500, 5000


1. Turn ON the system
2. Press “ Superkey” + “0” to get access to the Service mode. Enter “
csr” in UserID field, and “ user” in the password field
3. Select” Options” on the left pane to go to the Machine Options
screen
4. Select “ Read New Options Settings…” button on upper right corner
of the screen
5. Click “ Ok” on a next warning dialog. Then insert a MO disk with a
new options settings, then click OK to load the new settings.
6. If the new options settings are correct, press OK to accept. A
dialog will display “ Settings Successful”, click OK again , the MO
disk will be ejected
7. Restart the system

анатолий грушецкий

unread,
May 8, 2022, 11:53:55 AM5/8/22
to ultrasoundsolutions

To the community. I will explain how to change the options, and turn on the ones needed on the Philips ATL HDI 5000 ultrasonic machine, maybe someone will then help me get a service key for the 512 redwood.

To change the options you need to have a magneto-optical disk. Insert it into the device and format it. Then press the superkey and the “0” key and the C text at the same time.

The time on the device will stop, 10-15 minutes and the device will write the information to the magneto-optical disk, and throw it out about.

The MO disk must be inserted into external media, such as a computer, and opened. Copy file MACHOPTS.DAT to your computer.

Open this file with notepad ++

You see functions and options with changes 0 and 1.

Respectively 1- works, 0-no

The number of options working in the device is constant. But you need to decide what you need to change - language, sensor, cardio and the like.

In order to enable what you want, you need to disable what you consider unnecessary.

To do this, right-click in the program notepad ++

Perform the action - cut, paste.

https://drive.google.com/file/d/15Nh8VchnaNWwB9q7JRBCnES1Yh-Xz_6r/view?usp=sharing

To the comma, before 0 or 1, cut out the whole inscription, and rearrange it to the desired option, and the one that was there to cut and paste in place of the previous one.

Save changes.

Format MO drive. Create the PEG_HD folder on the computer, the SITE folder in it, the MACHCFG folder in it, and insert the edited MACHOPTS.DAT file there.

 Burn this folder to the MO drive.

 Then insert the MO disk into the ultrasonic machine, press the superkey and key "0", enter the login "csr" and password "user" select options, and read new options, confirm actions, and reboot the device - and you need the options included.