We used to use T3 (~4” long) bulbs bought at the local hardware store in vacuum for heating. I quit only because students would sometimes run them too hot, and parts near the lamp without good thermal conductivity to the chamber walls would get way too hot.
We just wiped them down with 95% ethanol.
Scott
Scott L. Anderson
Dist. Prof. and Henry Eyring Endowed Chair
Chemistry Dept./University of Utah
315 S 1400 E, Salt Lake City, UT 84112
--
You received this message because you are subscribed to the Google Groups "Surface Science" group.
To unsubscribe from this group and stop receiving emails from it, send an email to
surfacescienc...@googlegroups.com.
To view this discussion on the web visit
https://groups.google.com/d/msgid/surfacescience/12fe23cc-988b-4aac-af7d-fba29c6b0220n%40googlegroups.com.
Hi Brandon,
I agree with Scott Anderson. I’ve used the commercial G8 and GY8 style of quartz halogen bulbs from the hardware store in UHV for years. I’ve run them under full power for hundreds of hours. They give off a tremendous amount of heat if you put the full 110V on them so put a Variac® power regulator on them.
Of course, never touch quartz with your fingers. You can clean them with standard solvents before putting them in vacuum.
I have use IR heaters (Huntington Labs) and UV lamps (RBD Instruments) for bake-out but not the SpinLamp-300 halogen light sources.
Best,
Gary
═══════════════════════════════════
═══════════════════════════════════
Gary Navrotski, P.E., Ph.D. Navr...@anl.gov
X-ray Science Division - Beamline Instrumentation Group
Advanced Photon Source Office: 401-B3200
Argonne National Laboratory calendar link
9700 South Cass Avenue Ph: (630)252-7576
Lemont, IL 60439-4801 Fax: (630)252-9350
═══════════════════════════════════
To view this discussion on the web visit https://groups.google.com/d/msgid/surfacescience/06262e036bac4520811e0c4e7c864ffd%40chem.utah.edu.
--
Dear Prabhakar,
I also use halogen bulbs to heat crystals of my X-ray monochromator. I found that the nominal lifetime of this halogen bulb is only 50 h mentioned in its manual. I wonder why you can use so long time, for us it was broken within one year.
Do you always keep the lamp on all time or just turn on it before running XPS? If it was turned on before XPS measurement how long is the warming up time?
Best regards,
Jisheng
To view this discussion on the web visit https://groups.google.com/d/msgid/surfacescience/CAHKHx2UGnM%2Bmpta59e0gYR%3Dn%3DO8_8UHOPJgMAacjfOm58Z2GZw%40mail.gmail.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/surfacescience/SG2PR01MB2143BA877F4E665FD43080FBB4540%40SG2PR01MB2143.apcprd01.prod.exchangelabs.com.