Hi,
I have an Omicron room temperature STM-1 (horizontal tip, sample biased, tip grounded).
When I use the auto approach, the tip always ends up as if it is crashed (fully retracted "in the red", and the scope tells me the current is maximum).
Going one (!) coarse step backward, the tip can not see the surface anymore (fully extended "in the yellow" and the scope does not see a tunneling current).
So the tip never ends up in tunneling contact with a moderately extended piezo. It's either crashed or can't find the surface.
After going one (largest) coarse step backwards, I can manually approaching by taking smaller coarse steps one by one, and checking (going "forward") whether I can find the surface.
Most of the times this results in the exact same thing, even with the smallest coarse step size that my dial allows me to use: the tip never finds the surface, and at some point after one smallest coarse step size it's crashed.
I had a very useful extensive discussion with Omicron.
Based on that I tested boards. No luck.
I exchanged the scala controller with another controller from colleagues. No luck.
I checked the voltages going to the z-piezo - they are okay (+ 138 and -127 V). I checked the piezo+cable capacitance: 3 or 4 nF. That's normal.
During a very slow auto approach I checked whether the voltage to the piezo changes as it should be (whether ramps). It did.
In vacuum everything seems normal, outside of vacuum everything seems normal, but it does not work.
The only thing I can not exchange is the pre-amp cable as Omicron does not have one available to loan me.
The symptoms are consistent with the piezo not changing length despite the correct piezo voltages being sent to them. However, the measured capacitance is correct.
According to the manual, "the auto approach runs 50-200 nm per step". I presume that means that a coarse step is 50-200 nm. This makes me thing that the change in length of the piezo will be about 500 nm max. That means it's optically not possible to see this (if a microscope could be set up close enough to have a look).
I would welcome any helpful suggestions as to what could be the cause, or how to find out whether the tip actually moves upon application of piezo voltages.
THanks,
Paul