Hi Luis!
Thank you so much for the reply!
I have now changed the settings in tandem_params.xml to the following:
<note type="input" label="residue, modification mass">45.987721@C,
144.102063@[,144.102063@K</note>
<note type="input" label="residue, potential modification
mass">15.994915@M,144.102063@Y</note>
<note type="input" label="spectrum, minimum fragment mz">100.0</note>
Additionlay, I am not sure how to set the mass tolerance. I am using a
High resolution Synapt QTof that should handle at least 50 ppm as
precursor and 0.05 Da as fragments.
This is how param file look like:
<note> PRECURSOR MASS TOLERANCES. In the example below, a -2.0 Da to
4.0 Da (monoisotopic mass) window is searched for peptide candidates.
Since this is monoisotopic mass, so for non-accurate-mass instruments,
for which the precursor is often taken nearer to the isotopically
averaged mass, an asymmetric tolerance (-2.0 Da to 4.0 Da) is
preferable. This somewhat imitates a (-3.0 Da to 3.0 Da) window for
averaged mass (but not exactly)</note>
<note type="input" label="spectrum, parent monoisotopic mass error
minus">2.0</note>
<note type="input" label="spectrum, parent monoisotopic mass error
plus">4.0</note>
<note type="input" label="spectrum, parent monoisotopic mass error
units">Daltons</note>
<note>The value for this parameter may be 'Daltons' or 'ppm': all
other values are ignored</note>
<note type="input" label="spectrum, parent monoisotopic mass isotope
error">no</note>
<note>This allows peptide candidates in windows around -1 Da and -2
Da from the acquired mass to be considered. Only applicable when the
minus/plus window above is set to less than 0.5 Da. Good for accurate-
mass instruments for which the reported precursor mass is not
corrected to the monoisotopic mass. </note>
Should these changes fullfill my requirements or should I just leave
it? :
<note> PRECURSOR MASS TOLERANCES. In the example below, a -2.0 Da to
4.0 Da (monoisotopic mass) window is searched for peptide candidates.
Since this is monoisotopic mass, so for non-accurate-mass instruments,
for which the precursor is often taken nearer to the isotopically
averaged mass, an asymmetric tolerance (-2.0 Da to 4.0 Da) is
preferable. This somewhat imitates a (-3.0 Da to 3.0 Da) window for
averaged mass (but not exactly)</note>
<note type="input" label="spectrum, parent monoisotopic mass error
minus">25</note>
<note type="input" label="spectrum, parent monoisotopic mass error
plus">25</note>
<note type="input" label="spectrum, parent monoisotopic mass error
units">ppm</note>
<note>The value for this parameter may be 'Daltons' or 'ppm': all
other values are ignored</note>
<note type="input" label="spectrum, parent monoisotopic mass isotope
error">yes</note>
<note>This allows peptide candidates in windows around -1 Da and -2
Da from the acquired mass to be considered. Only applicable when the
minus/plus window above is set to less than 0.5 Da. Good for accurate-
mass instruments for which the reported precursor mass is not
corrected to the monoisotopic mass. </note>
In the param file it is also possible to define, taxonomy, database
and file locations etc? Do I need to pay any attention to this or may
I just select the mzXML file, the param file, and fasta database
through Petunia and hit the search button? :)
/Sincerely Marcus
On 11 Okt, 20:30, Luis Mendoza <
Luis.Mend...@systemsbiology.org>
wrote:
> Hello Marcus,
> As luck would have it, I just had to conduct analysis of an iTRAQ(8) sample
> using X!Tandem.
>
> First, there is unfortunately no GUI to set the parameters for the search,
> at least within the TPP.
>
> Do make sure that you specify the appropriate modifications based on the
> chemistry you used to prepare the peptides. Other than alkylation and MMTS,
> you should add the isobaric mass tag to all peptides -- typically at the
> N-term as well as Lysines, but this might depend on the exact protocol you
> followed. Add the following to the "residue, modification mass" tag:
> 144.102063@[,144.102063@K . A less abundant potential modification on Y
> might be useful to add, under "residue, potential modification mass",
> include 144.102063@Y , as well as any other potential mods (e.g. Oxidized M,
> etc).
>
> Also, note that the iTRAQ8-plex reagent has a very different mass
> (304.205360), should you ever use that label.
>
> The standard xtandem/k-score parameters that we ship with the TPP already
> exclude fragment ions lower than 125.0 Daltons; you can change this
> parameter by adding the following line to your params file, just replace the
> value with your desired minimum m/z:
>
> <note type="input" label="spectrum, minimum fragment mz">125.0</note>
>
> Hope this helps with your searches,
> --Luis
>