I concur with Jac's advise that PVC tubing can work well, provided the meter is
calibrated for the specific diameter and thickness of tubing to be used.
Jac's advice on switching tubes in the same hole is good, HOWEVER, it is important
that the outside diameters of the tubes are the same. Otherwise, there will be an air
gap outside the PVC tube that will affect the calibration. One can get around the tube
size problem by starting with which ever tube is smallest and then reauger the hole to
fit the larger tube.
If they are useful, there are PDF files for nine proceedings papers on calibration of
neutron meters and intercomparison of bulk density sampling methods that were
presented at a 1993 ASCE conference in Park City, Utah (seven of the papers) and at
a 1990 ASCE conference in Durango, Colorado. The url for these papers is:
http://www.kimberly.uidaho.edu/water/papers/index.html
One of the 1990 papers by Allen and Segura compares calibrations for aluminum vs.
PVC and shows the impact of air gap and tube diameter on the calibration. That
paper is found specifically at:
http://www.kimberly.uidaho.edu/water/papers/neutron_meters/Allen_Segura_1990%20
ASCE%20Access_Tube_Char.pdf
(all one line)
One of the 1993 papers by Dickey et al. also shows comparisons among PVC and
aluminum tubes as well as comparisons of tube type / meter type interactions for two
common meter types (CPN and Troxler). This paper's URL is:
http://www.nwisrl.ars.usda.gov/pubs/811.pdf
Also, remember that the shield count, if taken each day of operation, must be made
with the meter at least 1.2 m or more above the ground and away from moisture,
including that in the human body. The ASCE Task Committee that created the 1993
sessions on neutron meters strongly recommended positioning the meter on top of a
tall access tube that was at least 1.2 m high (above the ground) and made of the
same material as used for access tubes.
Regards
Rick Allen
Univ. Idaho