Removing flowcell between two reads

18 views
Skip to first unread message

Sivakumar Gowrisankar

unread,
Nov 16, 2010, 12:34:26 PM11/16/10
to sol...@googlegroups.com
Hi

Though it is not advisable, can we remove a flowcell and place it back between two reads of a run (say for paired-end run)?

Thanks
Siva

Lee Timms

unread,
Nov 16, 2010, 12:45:42 PM11/16/10
to sol...@googlegroups.com
You can remove the flowcell between reads 1 and 2 however you should not treat the run as paired end - you should analyze both reads as single read runs.

Lee


--
You received this message because you are subscribed to the Google Groups "solexa" group.
To post to this group, send email to sol...@googlegroups.com.
To unsubscribe from this group, send email to solexa+un...@googlegroups.com.
For more options, visit this group at http://groups.google.com/group/solexa?hl=en.

Yunchen Gong

unread,
Nov 16, 2010, 12:48:25 PM11/16/10
to sol...@googlegroups.com
I agree, and that's because you introduce an offset too big for the pipeline to correct.

Yunchen

Kevin M. Carr

unread,
Nov 16, 2010, 12:51:30 PM11/16/10
to sol...@googlegroups.com
No, you absolutely cannot remove the flowcell mid run. The cluster template is determined only once, at the start of read 1. Unless you have some sort of super-power there is no way you could reposition the flowcell within the tolerances required.

Kevin M. Carr

*************************************
Bioinformatics Specialist
Research Technology
Support Facility
S-20 A Plant Biology Lab
Michigan State University
East Lansing, MI. 48824

Ph: (517) 355-6759 x102
Fax: (517) 355-6758
************************************

Sivakumar Gowrisankar

unread,
Nov 16, 2010, 12:54:18 PM11/16/10
to sol...@googlegroups.com
Thanks everyone! This is useful information.

Lee Timms

unread,
Nov 16, 2010, 12:58:26 PM11/16/10
to sol...@googlegroups.com
You should start a new run after first base incorporation from R2.  

Once first base has completed, stop your run and open a new recipe.  Recalibrate, find edge etc in the new recipe.

J

unread,
Nov 16, 2010, 5:40:42 PM11/16/10
to solexa
On a GA, no. It's much more likely to work on a HiSeq, though as you
say, it's not a great general practice.
Reply all
Reply to author
Forward
0 new messages