It depends on your input material. In the dataset from human
reference RNAs (in the paper) across the range from 10 pg to 100 ng
input, which we sequenced together with negative controls in one
batch on the NextSeq, we got 75% PF, 79% Q30, and the first base
composition graph below. There are several reasons for the
non-uniform base composition:
- built-in GGG in positions 6-8
- high contamination by adapter dimers in the lowest-input
samples
- contamination by poly(A) sequence increasing toward the end
- spurious G's detected at the end after sequencing through the
adapter dimer (on a NextSeq no signal looks the same as G)
Point 3 is exacerbated when the input is degraded RNA, such as the
FFPE LCM experiment in the paper (bulk tissue and single cells in
the same batch), which had 72% PF, 61% Q30, and the second base
composition graph below; those results are typical or maybe even a
little better than usual for FFPE LCM libraries.