Hi Adams,
You don’t really have to harvest murine lung.
You could look at the inlet of a chip and count how many cells are entering the chip per unit of time (e.g. per 1 min).
Then you should count how many beads are entering the chip per same unit of time.
If you know the frequency (or volume) of your droplets you could calculate "productive droplets" using Poisson equation.
My suspicion is that you may have very very low amount of productive droplets, which will lead you to very low amounts of cDNA.
We also see debris (depending on a sample). Sometimes we spin cells in Lympholyte or Ficol or Optiprep. Typically we use Optiprep, but if the amount of cells is very low (e.g. <20.000 cells in total), then we just encapsulate cells without gradient centrifugation. As you can expect, debris increases the background read noise quite a lot.
Do you have any advice on how to to mitigate ambient RNA accumulation?
It depends on your experimental set-up and lysis conditions. If you are not performing cDNA synthesis inside the droplets you might add 10-100 times diluted RNAse into your cell suspension. The idea would be that RNAse would degrade ambient RNA during encapsulation, but once cells will enter the droplet the lysis reagents will immediately destroy RNAse. I have never tried this approach though.
Let me know if you have other comments/thoughts,
Linas