SOLT Isolation Calibration - What

19 views
Skip to first unread message

just Random

unread,
Oct 19, 2021, 7:05:36 AMOct 19
to scikit-rf
Is it possible to use the isolation parameter in the SOLT routine to remove test set capacitances? 

The isolation coefficients should represent crosstalk between the channels?

I put together a small setup with simulated ideals / measurements in awr.
The Coff capacitances are inherit in my switches...

Setup, code and results:

The isolation is the 2-port network form the "real" load measurement. It does not seem to do much...

Thx in advance :)




Henrik Forstén

unread,
Oct 19, 2021, 12:43:56 PMOct 19
to scik...@googlegroups.com

Isolation calibration just subtracts the isolation measurement S21 and S12 from the measurement S21 and S12. It's a very simple calibration and can only correct for the leakage outside the calibration error model. Your measurement has more complicated leakage and it can't be applied.

--
You received this message because you are subscribed to the Google Groups "scikit-rf" group.
To unsubscribe from this group and stop receiving emails from it, send an email to scikit-rf+...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/scikit-rf/ff62c819-4ed2-4d44-ab30-c1ca26f55c76n%40googlegroups.com.

just Random

unread,
Oct 20, 2021, 3:33:48 AMOct 20
to scikit-rf
I noticed that yesterday too  while reading the source for the algorithm given in the source code:
"Handbook of Microwave Component Measurements: with Advanced VNA Techniques"

Oh well, makes sense. Thank you for your answer!