I think that any circuit with a voltage source and some resistance/
impedance in series with the voltage source before it is accesible to
the outside would be current limited. For instance, a 5V source with
a 1000 ohm resistor in series with the lead going outside the cabinet
would be current-limited to 5 milliamperes.
So according to your explanation, shall we only make an examination of
limited current circuits if access is possible.
For example an EMI filter or a switching power supply , they have
special case and access
to the inner case is not possible by the operator , in this situation
there is no need to make a limited current circuit measurement, is it
right.
there is no requirement for an LCC for EMI filters in a
Class I construction, assuming all other construction
requirements are met and no component bridges the basic
insulation between input and protective earth.