Thanks
John
On my Thurlby's (considerably older though), OFL means you have the
current limit control set fully clockwise equating to the 'no current
limit' setting (let me smoke & catch fire first....). This when the
power out switch is off.
Try turning the control down? Could be the control itself is faulty if
no change.
--
Adrian C
Thanks. However, current limit control setting makes no difference. Always
shows OFL. Thought it was unlikely to be the control but I will try swapping
it for a different one to see if it makes any difference.
John
"Adrian C" <em...@here.invalid> wrote in message
news:6k57lbF...@mid.individual.net...
That happened to mine (Thandar TS3021S) There was a diode soldered
directly across the output terminals which was shorted - I suspect
someone had use it to charge a gelcell and connected it up the wrong
polarity.
Ron(UK)
John
"Ron(UK)" <r...@lunevalleyaudio.com> wrote in message
news:OdCdnZU_A6JViUPV...@bt.com...
OK. If your assertion that the OFL display means overload is correct,
I'd take that to mean that its an indication from the display modules
built-in ADC that there is excess voltage at it's input compared with
it's specifications for full scale 'deflection'.
Can you see what part they used, find a datasheet and trace the input
pin(s). Then stick a meter there, attach a load and see if on varying
the output voltage if there is anything sensible changing for these inputs.
I'm guessing all of the circuitry still is plain discrete analogue
driven (compared with my old PL154) and you might find the culprit
through simple fault finding. If it is a silicon jungle chip maybe not.
Though a manual has to be found from somewhere ...
--
Adrian C
If your PSU senses the current via a low ohms series resistor, then an
open resistor may produce the observed symptom. In any case it may be
a good place to start troubleshooting.
- Franc Zabkar
--
Please remove one 'i' from my address when replying by email.