On Tue, 18 Oct 2011 11:18:58 -0500 Andy Cuffe <
acu...@gmail.com> wrote in
Message id: <
4h9r9793ivn7c2fmb...@4ax.com>:
I never trust a 2 wire DMM when reading resistances less than an ohm or
so. Contact and test lead resistance always seem to be an issue. I would
expect that you'll find some around 5 ohms or more, though.
>The other two seemed fine, but I think I'll
>replace them all so I don't have to go in again.
Don't bother with the thin blue one, I've yet to see one of those go
bad...
>The bad shutter has an area the size of a quarter which appears green
>when the scope is off (as opposed to the normal dark magenta). When
>it's operating, the color in this area appears much less saturated. I
>would say it's either between the LCD layers, or a failure of one of
>the LCDs.
Once again, I'd bet that it's the optical gel.
>It sounds like it's a delicate operation. I might attempt swapping
>the good shutter onto a new CRT if I find one, but otherwise I'll
>leave it alone for now. I don't want to risk destroying both CRTs.
>At least the weak one responded well to my CRT rejuvenator. How hard
>does the optical gel make it to remove the shutter?
It's not the gel that makes it difficult, it's the RTV that seals the
shutter to the CRT.
>Does you have to
>cut it away from the face of the CRT, or can you just pull it off
>after separating the RTV form the sides of the CRT? Does it still
>look good without the gel?
You have to cut the RTV off from the side of the CRT, then cut inward
between the face of the CRT and the shutter. It still takes quite a bit of
strength to pull the shutter free. If it doesn't come free then cut some
more, lather, rinse and repeat. It looks fine without the gel, but it's a
gooey mess cleaning it all off the CRT and shutter. I use a soft plastic
scraper to scoop most of it up, then a lot of rags and isopropyl alcohol
to finish the job. Wish I knew of something that would attack the stuff
better.
>I am using a fan when running it with the cover off. You're right
>that it gets hot quickly. I'm surprised the service manual doesn't
>have a warning about that.
>
>I did email the TDS500 software to several people, so you probably got
>it from me. Originally, I was looking for a way to back up the NVRAM,
>but in the end I was able to back it up using my EPROM programmer, so
>I never had to use it.
>
>Now I would like to find the cal software for the 744. It's nice to
>have, and it would answer my question about the TV trigger. I'll let
>you know if I find a copy.
There's a guy named Hakan on the Yahoo Tektronix group that probably has
the TDS744 cal software. He also has all the schematics for TDS500, 600
and 700 series of scopes, but is unwilling to scan them for some reason...
http://tech.groups.yahoo.com/group/TekScopes/
>I will post back with the results.
Looking forward to it.
>Thanks,
>
>Andy
You're welcome - good luck!