My hamfest special Fluke 80 something series something 20 year old meter
started to act weird so it was time to upgrade.

Got a spiffy 87 series V meter from Specialized.

It's much larger than the old meter which is sort of a bummer, but it has
a nice display and came with a cheapo temp probe.

I tried to measure the temp of the shower water I like and few minutes
later the meter started to freak out with "L38d" errors. It would not shut
the hell up unless it was in the mA or uA range.

So apparently the humidity condensed inside the lead sockets and shorted
out the other side of the connector shell causing the meter to think I was
stuck in a current range of some sort.

a dry qtip wasn't able to dry the sockets out. Neither was pure alcohol,
or a fan. I was about to get a RAM on the piece of shit, but finally,
after about 20 minutes of being set in front of a fan, the thiung dried up
enough to work again.

I checked the service manual and found a strange section suggesting your
spray a swab with WD-40 and work it around inside the lead sockets to
prevent this problem.

WTF.

Is there some some secret settings I can enter on this meter to supress
the input warnings, or is it time to unsolder half the connectors?

Since the entire meter is semi sealed and had gaskets and orings all over,
it seems stupid that a miniscule amout of moisture or a drop of water will
completely disable the meter.

I'm calling shit design on this one.




Charles

unread,
Sep 3, 2012, 6:55:26 PM9/3/12
to


"Cydrome Leader" wrote in message news:k232om$nqs$1...@reader1.panix.com...
I have been using a CEN-TECH P98674 DMM with an external temperature sensor.
It is kind of flimsy (the sensor) so I have been careful when using it. I
will not stick it into the shower!

I own 2 Fluke DMMs and they are both flaky.

Good luck!



Jamie

unread,
Sep 3, 2012, 7:31:33 PM9/3/12
to
Fluke does not make meters like they used to any more. Actually, I don't
think they make them in the US any more. The last I heard, they are all
made in china now..

For a while, they were having the low ends made in china and kept the
upper models where ever they were doing it the last time. Then, that
changed and all of it is now chinese made.. That's what the sales rep
told us.

We have a Fluke 289 that was not made in china. Couple of the guys
wanted a 289 so when theirs came in, those were made in china and you
can see the difference in the case and switch action, they are not the
same quality..

It does not surprise me that you have found that design flaw.. You
most likely got a chinese designed version in a case that mimics the
original fluke and approved by fluke, which is nothing more than an
office and web sight, as far as I can see.

Jamie

ED

unread,
Sep 3, 2012, 11:01:45 PM9/3/12
to

"Cydrome Leader" <pres...@MUNGEpanix.com> wrote in message
news:k232om$nqs$1...@reader1.panix.com...
NOT per http://www.youtube.com/watch?v=CMV7E186_Nw&feature=plcp


Jeff Liebermann

unread,
Sep 3, 2012, 11:07:39 PM9/3/12
to
On Mon, 3 Sep 2012 20:08:54 +0000 (UTC), Cydrome Leader
<pres...@MUNGEpanix.com> wrote:

>Got a spiffy 87 series V meter from Specialized.

Oh-oh. If you have one of the older units, you may have RF
susceptibility problems, which includes clearing the flash and
bricking the meter (15min):
<http://www.eevblog.com/2010/09/18/eevblog-112-gsm-vs-the-fluke-87v-multimeter/>
and the fixed version (18min):
<http://www.eevblog.com/2011/07/11/eevblog-185-fluke-87v-multimeter-gsm-fix/>
Old and new versions comparison:
<http://www.flickr.com/photos/eevblog/sets/72157627120630430/>
or just listen to the beep. The old model has a rather loud beep. The
new version has a much quieter beep. Also, the new one comes with
shrouded probe tips.


--
Jeff Liebermann je...@cruzio.com
150 Felker St #D http://www.LearnByDestroying.com
Santa Cruz CA 95060 http://802.11junk.com
Skype: JeffLiebermann AE6KS 831-336-2558

Cydrome Leader

unread,
Sep 4, 2012, 2:13:55 AM9/4/12
to
From what little experience I have with the new meter vs the one from
decaces ago, I'm going to agree. This one is assembled in the US, like my
old one though.

> For a while, they were having the low ends made in china and kept the
> upper models where ever they were doing it the last time. Then, that
> changed and all of it is now chinese made.. That's what the sales rep
> told us.
>
> We have a Fluke 289 that was not made in china. Couple of the guys
> wanted a 289 so when theirs came in, those were made in china and you
> can see the difference in the case and switch action, they are not the
> same quality..
>
> It does not surprise me that you have found that design flaw.. You
> most likely got a chinese designed version in a case that mimics the
> original fluke and approved by fluke, which is nothing more than an
> office and web sight, as far as I can see.

It's a legit Fluke meter bought in the last 6 months from a valid
distributor. It's still preposterous that the slightest amount of moisture
causes it to think the leads are in the wrong sockets or that you need to
spray WD-40 all over the fucking thing.


Cydrome Leader

unread,
Sep 4, 2012, 2:33:49 AM9/4/12
to
Jeff Liebermann <je...@cruzio.com> wrote:
> On Mon, 3 Sep 2012 20:08:54 +0000 (UTC), Cydrome Leader
> <pres...@MUNGEpanix.com> wrote:
>
>>Got a spiffy 87 series V meter from Specialized.
>
> Oh-oh. If you have one of the older units, you may have RF
> susceptibility problems, which includes clearing the flash and
> bricking the meter (15min):
> <http://www.eevblog.com/2010/09/18/eevblog-112-gsm-vs-the-fluke-87v-multimeter/>
> and the fixed version (18min):
> <http://www.eevblog.com/2011/07/11/eevblog-185-fluke-87v-multimeter-gsm-fix/>
> Old and new versions comparison:
> <http://www.flickr.com/photos/eevblog/sets/72157627120630430/>
> or just listen to the beep. The old model has a rather loud beep. The
> new version has a much quieter beep. Also, the new one comes with
> shrouded probe tips.

It's really hard to tell what the heck those photos are supposed to be,
but I do have the one with the metal stub on the mode selection switch,
which would be rev 11 or the new one, I think.

So I just ran a test.

power on meter with no leads. Great.

I then dampened a qtip and squeezed it as dry as it can get as they're too
fat to slide in the jack all the way when fluffed up.

I jammed it into the uA range jack and the meter started the several beeps
er second and LEAd (or whatever it looks like in 7-segment) freak out
input protection mode.

I stuck the other dry end of the qtip in the jack and it still won't
settle down, even after being shut off and on. A second qtip still can't
dry the jack enough. No amount of rotating or just straight up and down
action with any number of qtips seems to dry the thing out enough to work
again.

The only ranges that act normal are mA and uA.

From this photo

http://www.flickr.com/photos/eevblog/5904503364/in/set-72157627120630430/

You can see there are two solder leads per jack. It seems that one side is
the "working" terminal of the jack and the other is just for sensing when
a probe is inserted because it will short out both pins. Inside the red
plastic jack each half terminal is just a nearly semicircle shaped contact
that wraps around almost half the banana-ish plug you's stick in there.

Whatever they're sensing must be in the microvolts or microamps of leakage
between those half terminals. Whatever it is, it's way over sensitive.

Mark Zacharias

unread,
Sep 4, 2012, 6:25:23 AM9/4/12
to
"Cydrome Leader" <pres...@MUNGEpanix.com> wrote in message
news:k232om$nqs$1...@reader1.panix.com...
Fluke will fix it for free or replace it under the warranty. Do not try to
fix it yourself.

I have some experience with Fluke Service. Although nobody is perfect, they
are really the best in the business.

If you needed a waterproof meter you should have bought a 28 series II,
which has identical features.

Mark Z.

Mark Zacharias

unread,
Sep 4, 2012, 6:29:44 AM9/4/12
to
"Jamie" <jamie_ka1lpa_not_v...@charter.net> wrote in message
news:CUa1s.1618$Ok1...@newsfe13.iad...
The Fluke 110 series (116, 117, etc) are Chinese made. The 80 series are all
still U.S. made.

Even the Chinese ones are made to high standards, but only carry a 3 year
warranty, whereas the U.S. made models have the famous "lifetime" warranty.

To my knowledge there are no Chinese - made 289's. Frankly somebody would
have to prove this to me.

I have a brand-new 289 and it is definitely American.

Mark Z.

Mark Zacharias

unread,
Sep 4, 2012, 6:32:33 AM9/4/12
to
"Cydrome Leader" <pres...@MUNGEpanix.com> wrote in message
news:k24673$ckc$1...@reader1.panix.com...
Do not spray WD in there - but since it is necessary to open the unit to
replace a fuse for example, you will not void the warranty simply by opening
it. Once you do, your problem may be obvious. Perhaps there is more water
inside than you realize - in which case you REALLY need to get in there and
clean it out to prevent corrosion.

Mark Z.

Mark Zacharias

unread,
Sep 4, 2012, 6:38:33 AM9/4/12
to
"Cydrome Leader" <pres...@MUNGEpanix.com> wrote in message
news:k247cd$45g$1...@reader1.panix.com...
Sorry if I responded too quickly to your earlier post if you've already
opened it.

Nevertheless, you should get it to Fluke for service. They really will fix
it for free.

I've even heard of many cases where they cover units not 'technically' under
warranty, not original owner, no purchase documentation, etc.

I'm assuming you are in the U.S.

Their turnaround is usually about 2 weeks, including UPS both ways.

1-888-99-FLUKE (1-888-993-5853)

Mark Z.

Jamie

unread,
Sep 4, 2012, 9:19:19 AM9/4/12
to
we have at least 7 people in the shop using them and they're all chinese
made, accept the one I have. It's possible you can still get US
made ones. Our sales rep told us differently..

This reminds me of the Baldor micro AC drives we get, (VS1MD) series.
if you go to their web sight and look at the advertised drive, they show
a membrane keypad with a slightly different art design on the face. when
you actually buy one, you get a chicklet push button keys and the front
is just a plane jane control. It really does not matter but the point
is, they are not shipping you what they advertise.. Also the ones we get
are made in S.Korea.

Jamie


Jeff Liebermann

unread,
Sep 4, 2012, 11:19:01 AM9/4/12
to
On Tue, 4 Sep 2012 06:33:49 +0000 (UTC), Cydrome Leader
<pres...@MUNGEpanix.com> wrote:

>It's really hard to tell what the heck those photos are supposed to be,
>but I do have the one with the metal stub on the mode selection switch,
>which would be rev 11 or the new one, I think.

Comparison of a Rev 9 meter with a prototype Rev 11 meter.

>So I just ran a test.

You did the water test, not the RF susceptibility test. Find a GSM
phone and see what it does. It wasn't my 87-V meter, but one of the
techs on a mountain top radio site announced that his meter was dead
and unusable anywhere near a source of RF. He sent it back to Fluke
for repair. I think (not sure) it was replaced by a later model.

As for waterproof, it appears that the 87-V is NOT waterproof. That's
the newer type 28 meter.

>From this photo
>http://www.flickr.com/photos/eevblog/5904503364/in/set-72157627120630430/
>You can see there are two solder leads per jack. It seems that one side is
>the "working" terminal of the jack and the other is just for sensing when
>a probe is inserted because it will short out both pins. Inside the red
>plastic jack each half terminal is just a nearly semicircle shaped contact
>that wraps around almost half the banana-ish plug you's stick in there.

Reading from left to right on the above photo, that's:
+ terminal - terminal mA/uA Amps
While the mA/uA and Amps terminals have split pin connections, the +
and - terminals do not. I can see a possible problem with leakage
across the split terminals, but not across the + and - which don't
seem to have a sensor. My guess(tm) would be something is leaking
across the PCB, possible from the two Amps terminals to ground or
internal power. It might also be testing for any resistance between
the two Amps terminals, or between the Amp terminals and the +
terminal, which are unusable conditions.

>Whatever they're sensing must be in the microvolts or microamps of leakage
>between those half terminals. Whatever it is, it's way over sensitive.

True. Find a tiny o-ring that fits into the connector and cram it
down to the bottom until it seals the receptacle. That's should help.

Incidentally, I have two Fluke meters (Model 10 and 73). No problems
and well worth the money.

Cydrome Leader

unread,
Sep 4, 2012, 12:23:27 PM9/4/12
to
Sadly there isn't anything else wrong with the meter. The sockets for the
leads are single piece of sealed looking plastic, which is then o-ring
sealed to the front bezel. There was no water splashing around inside
the meter, which is what I expected at first.

I ran another test. Turn on the meter and place is face down an inch or
two over over a cup of warm tap water. Within seconds it sensed a lead
errors. I have to see if my breath can trigger it next.



Cydrome Leader

unread,
Sep 4, 2012, 12:35:00 PM9/4/12
to
I'm doubting the meter is "broken" in that it won't behave like any other
one just like it. I don't need waterproof, just to know that the meter
doesn't have water absorbent plastics in it and it won't completely stop
working if my glasses fog up for a second when going to another room.

The slightest trace of moisture doesn't seem to need to get past the jack
itself, and it's not like I'm pouring water into the jacks. Maybe I need
to make some rubber plugs for the damn thing.

It reminds me of those horrible ceramic thick film resistor looking dew
sensors in VCRs that always failed on their own. Just drawing on them with
a pencil seemed to take care of that (not really sure how though).


John Robertson

unread,
Sep 4, 2012, 2:13:06 PM9/4/12
to
A big NO to WD-40 please!

Instead use some Dow Corning Electrical Grease (#4 is what I like) on
circuits that seem to be moisture sensitive. It is a water displacement
grease, and non-corrosive and won't turn to glue/goo (as a certain
heavily advertised produce already mentioned in this note will) over time.

John :-#)#

--
(Please post followups or tech enquiries to the newsgroup)
John's Jukes Ltd. 2343 Main St., Vancouver, BC, Canada V5T 3C9
Call (604)872-5757 or Fax 872-2010 (Pinballs, Jukes, Video Games)
www.flippers.com
"Old pinballers never die, they just flip out."

Cydrome Leader

unread,
Sep 4, 2012, 3:54:33 PM9/4/12
to
I don't own WD-40 or duct tape for those reasons. I wish I was making up
the WD-40 part, but it's actually in their manual.


John Robertson

unread,
Sep 4, 2012, 7:52:40 PM9/4/12
to
Ah, right, I have to READ all the text <grin>...overlooked that WD-40
was 'recommended' in the manual. A bit scary indeed!

John :-#(#

Mark Zacharias

unread,
Sep 4, 2012, 9:38:54 PM9/4/12
to
"Jeff Liebermann" <je...@cruzio.com> wrote in message
news:ct5c48dkldr9qpi2r...@4ax.com...

<snip>

down to the bottom until it seals the receptacle. That's should help.
>
> Incidentally, I have two Fluke meters (Model 10 and 73). No problems
> and well worth the money.
>
> --
> Jeff Liebermann je...@cruzio.com
> 150 Felker St #D http://www.LearnByDestroying.com
> Santa Cruz CA 95060 http://802.11junk.com
> Skype: JeffLiebermann AE6KS 831-336-2558


I have a 85-III, an 87-V, a 27-II, an 8800A/AF, the new 289 and an 8840A. In
my experience they have been absolutlely reliable, and in measurements
there's not over about 400 microvolts difference on DC volts between them,
even the 8800 produced around (?) 1975 and the 8840A made around 1986.

Yes, I am a "multimeter junkie"...

I have left my 87-V in my refrigerator for days on end getting min-max on
temperature. It would get condensation when moved from the cold to a warm
room, but never malfunctioned.

I have not submerged any of them, but the 27-II is supposed to be
waterproof...

The 87-5 has split terminals for the current jacks - this may explain why
the other fellow was getting the "leads" indication with moisture in there.

Mark Z.

Cydrome Leader

unread,
Sep 5, 2012, 12:35:12 AM9/5/12
to
I just confirmed that just one breath like one would use to clean
eyeglasses is enough to "short out" the small current range jack.

Can anybody else try this?

I'm going to ask fluke about this.

Mark Zacharias

unread,
Sep 5, 2012, 6:22:48 AM9/5/12
to
"Cydrome Leader" <pres...@MUNGEpanix.com> wrote in message
news:k26kq0$mp0$1...@reader1.panix.com...
I'm going to agree with you. I think that is a fault. I'm going to try it
with mine.

Mark Z.

JW

unread,
Sep 5, 2012, 8:03:55 AM9/5/12
to
On Wed, 5 Sep 2012 04:35:12 +0000 (UTC) Cydrome Leader
<pres...@MUNGEpanix.com> wrote in Message id:
<k26kq0$mp0$1...@reader1.panix.com>:

[...]

>I just confirmed that just one breath like one would use to clean
>eyeglasses is enough to "short out" the small current range jack.
>
>Can anybody else try this?
>
>I'm going to ask fluke about this.

Doesn't happen with mine. Serial # is 96800437.

Rod Speed

unread,
Sep 5, 2012, 6:26:34 PM9/5/12
to


"ED" <xxa...@hotmail.com> wrote in message
news:k23quq$ee8$1...@dont-email.me...
Just another example of the sales bullshit not matching reality.

Mark Zacharias

unread,
Sep 5, 2012, 9:12:14 PM9/5/12
to
"JW" <no...@dev.null> wrote in message
news:uqfe48155tia3duo7...@4ax.com...
OK, I checked my 87-5. DOES NOT trigger the lead warning by any amount of
close-up hot breathing into it.

Also checked my 289 again to confirm "Made in America".

It is, "of U.S. and non-U.S. components" - about as I expected.

Mark Z.

Cydrome Leader

unread,
Sep 6, 2012, 1:05:12 AM9/6/12
to
Mark Zacharias <mark_za...@labolgcbs.net> wrote:
> "JW" <no...@dev.null> wrote in message
> news:uqfe48155tia3duo7...@4ax.com...
>> On Wed, 5 Sep 2012 04:35:12 +0000 (UTC) Cydrome Leader
>> <pres...@MUNGEpanix.com> wrote in Message id:
>> <k26kq0$mp0$1...@reader1.panix.com>:
>>
>> [...]
>>
>>>I just confirmed that just one breath like one would use to clean
>>>eyeglasses is enough to "short out" the small current range jack.
>>>
>>>Can anybody else try this?
>>>
>>>I'm going to ask fluke about this.
>>
>> Doesn't happen with mine. Serial # is 96800437.
>
>
> OK, I checked my 87-5. DOES NOT trigger the lead warning by any amount of
> close-up hot breathing into it.

I got an RMA today for mine.

The tech support guy was basically able to cut me off before I described
the problem, and suggested I clean the jack with pure alcohol, which I
tried already. That didn't work so it's going back. It was hinted it would
take years of grime to build up on the jack before it acts weird, which
obviously isn't the case.

> Also checked my 289 again to confirm "Made in America".
>
> It is, "of U.S. and non-U.S. components" - about as I expected.

So it's back to the 73 series II for now I guess.

I'm really curious as to what the resistance is between the "shorted"
contacts, but I can't measure it without desoldering the jack, which would
probably cancel any warranty work on the unit.

Mark Zacharias

unread,
Sep 6, 2012, 9:24:30 PM9/6/12
to
"Cydrome Leader" <pres...@MUNGEpanix.com> wrote in message
news:k29au8$63i$3...@reader1.panix.com...
DC resistance between split halves of current jacks on the 87-5 = 2.4 meg

on the 85-3 it is 30 meg on mA jack, 25 meg on Amp jack
.
The 289 appears to use optical sensing.

Mark Z.

Cydrome Leader

unread,
Sep 7, 2012, 12:57:38 PM9/7/12
to
Interesting- I'll test resistance with HP 34401 (which has been mostly
surprise free except for the really low input impedance of the AC ranges)
before I ship it off.

Yesterday was less humid in the house, and it took an extra puff to
trigger the lead error, but it was still easy. Maybe the shroud of the
jack uses a water absorbing plastic like nylon.




William Sommerwerck

unread,
Sep 7, 2012, 2:10:59 PM9/7/12
to
> Yesterday was less humid in the house, and it took an extra puff to
> trigger the lead error, but it was still easy. Maybe the shroud of the
> jack uses a water absorbing plastic like nylon.

Maybe there's some hygroscopic shmutz around the jacks. Have you thoroughly
cleaned/flushed the area?


Cydrome Leader

unread,
Sep 10, 2012, 12:49:14 AM9/10/12
to
yeah, it's clean, plus it's basically new. I've not been dragging it
around a steel mill or anything like that.

William Sommerwerck

unread,
Sep 10, 2012, 8:32:15 AM9/10/12
to
>>> Yesterday was less humid in the house, and it took an extra puff
>>> to trigger the lead error, but it was still easy. Maybe the shroud
>>> of the jack uses a water-absorbing plastic like nylon.

>> Maybe there's some hygroscopic shmutz around the jacks. Have
>> you thoroughly cleaned/flushed the area?

> Yeah, it's clean, plus it's basically new. I've not been dragging it
> around a steel mill or anything like that.

This has puzzled me ever since you first posted.

CMOS is sensitive to surface contamination -- anything that's conductive,
even only slightly including water.

You might try hitting the area with freeze spray. This ought to reveal two
things... First, is the problem caused simply by blowing on the board (ie,
moisture has nothing to do with it). Second, the low temperature will cause
moisture to condense. Does the problem occur after the spray has evaporated?

If I had access to an ultrasound bath, I'd dunk the board in an inert
solvent and clean it thoroughly. If that didn't clear the problem, I'd be
surprised.


Cydrome Leader

unread,
Sep 11, 2012, 1:55:59 PM9/11/12
to
William Sommerwerck <grizzle...@comcast.net> wrote:
>>>> Yesterday was less humid in the house, and it took an extra puff
>>>> to trigger the lead error, but it was still easy. Maybe the shroud
>>>> of the jack uses a water-absorbing plastic like nylon.
>
>>> Maybe there's some hygroscopic shmutz around the jacks. Have
>>> you thoroughly cleaned/flushed the area?
>
>> Yeah, it's clean, plus it's basically new. I've not been dragging it
>> around a steel mill or anything like that.
>
> This has puzzled me ever since you first posted.
>
> CMOS is sensitive to surface contamination -- anything that's conductive,
> even only slightly including water.

I've decided not to mess with the inside of the meter itself. It's under
warranty, and clearly screwing with it won't help me out in this case. The
board itself looks clean- none of that white dust you see on crappy board
with water based flux or whatever causes that. Just blowing on the jacks
with the meter close will cause the lead error, although today it's harder
as it's less humid out. Maybe there's some factory included contanmination
on the PCB itself, so that the extra gigaohm of my breath is just too much
for it.

I did measure the resistance of the current jack, which was about 1.8M and
it changed as if there was a cap in the sense circuit. There were visible
1M resistors connected to the sense side of the current jacks.

> You might try hitting the area with freeze spray. This ought to reveal two
> things... First, is the problem caused simply by blowing on the board (ie,
> moisture has nothing to do with it). Second, the low temperature will cause
> moisture to condense. Does the problem occur after the spray has evaporated?
>
> If I had access to an ultrasound bath, I'd dunk the board in an inert
> solvent and clean it thoroughly. If that didn't clear the problem, I'd be
> surprised.

Fluke seemed keen on getting in for repair, so we'll see what they do to
it.

William Sommerwerck

unread,
Sep 11, 2012, 4:00:42 PM9/11/12
to
> Fluke seemed keen on getting in for repair, so we'll see
> what they do to it.

Please let us know. I like to hear solutions to odd problems.