Please can anyone tell me the 'calibration procedure' for this meter?
TIA - Mark.
More likely a repair job, 0.3 to 0.4 percent out for a calibration issue,
yours is 2 orders of magnitude over that.
The first thing to check is consistency, or not, of the wrong readings for
known test items
--
Diverse Devices, Southampton, England
electronic hints and repair briefs , schematics/manuals list on
http://home.graffiti.net/diverse:graffiti.net/
>Hi.
>I have a second-hand LCR-meter, Tenma model 72-6634, which seems to be
*seems to be???* So you are not sure?
>reading 30-40% high on the L/C ranges. (R seems accurate, though!)
>
>Please can anyone tell me the 'calibration procedure' for this meter?
You can't determine whether it is reading high or not (you do say
"Seems to be" but you want to calibrate it?
Maybe try taking it to someone who knows how to evaluate it for
accuracy (has the requisit standards, for example) then think about
having someone calibrate it.
>
>TIA - Mark.
> MAC <crump...@googlemail.com> wrote in message
>
news:2ebf32bf-bd18-430a...@y9g2000yqg.googlegroups.com...
>> Hi.
>> I have a second-hand LCR-meter, Tenma model 72-6634, which seems to
>> be reading 30-40% high on the L/C ranges. (R seems accurate, though!)
>>
>> Please can anyone tell me the 'calibration procedure' for this meter?
>>
>> TIA - Mark.
I've got one of those ! Put in a good battery and recheck it. It eats
batteries, at least mine does, and the readings vary greatly with
voltage.
>
> More likely a repair job, 0.3 to 0.4 percent out for a calibration
> issue, yours is 2 orders of magnitude over that.
> The first thing to check is consistency, or not, of the wrong readings
> for known test items
>
>
--
Best Regards:
Baron.