I assume you are in 120V land!
A simple calculation (P=(V^2)/R) shows the resistor probably dissipates
72 watts which would correspond neatly with your observed 78W draw.
However the originals are only rated for 20W - no wonder they fail!
You do not know what extra testing has been done by "Helen of Troy" to
qualify these resistors for use at 3.6 times their rating, nor do you
know if a third party's 20W 200 ohm resistor will fail safe.
If you had appropriate professional liability insurance for small
medical appliance repair you would run away screaming from a customer
who asked you to fix a device that overran a heater by that high a
factor using unapproved parts.
The fact that you belive the GFI can be relied on to backstop your
possible shoddy workmanship means I can be certain that you are not a
fit and competent person to repair these appliances.
Basically you are putting all your assets on the line. Your home, your
wife's business etc. and because she let you work on them, she's
negligent as well if they are ever put back into service.
Find a company willing to repair them (if you can) and make sure they
have appropriate insurance or forget about it. --