Does anybody have a user manual for the Hewlett Packard
5302A Frequency Counter?
Does anybody know when this unit was sold or what the
specified accuracy is?
Gerard Tel
URL: http://www.cs.uu.nl/~gerard/
The 5302A was still for sale in 1983 (my catalog date). Basic Specs:
5300A and 5300B Measurement System Mainframe
===================================
Time Base
Standard Crystal Frequency: 10 MHz
Stability
Aging rate: <3 parts in 10**7/mo
Temperature: < +- 1 part in 10**6, 0 deg to 50 deg C
Typically: <+- 2 parts in 10**6, 15 deg to 40 deg C
Line Voltage:
5302A 50 MHz Universal Counter Module
===================================
Range : Channel A 10 Hz to 50 MHz; Channel B 10 Hz to 10 MHz
Sensitivity (Min): 25 mV rms sine wave 50 Hz to 1 MHz, 50 mV rms sine wave
10 Hz to 10 MHz. 100 mV rms sine wave at at 50 MHz. 150 mV p-p pulse at
minimum pulse widith, 50 ns.
Trigger level: selectable positive, negative, or zero volts.
Hope this helps,
Rob
Gerard Tel <ger...@cs.uu.nl> wrote in message news:376A55...@cs.uu.nl...
>Hello,
>
>Does anybody have a user manual for the Hewlett Packard
>5302A Frequency Counter?
>Does anybody know when this unit was sold or what the
>specified accuracy is?
I have a manual for this, but it is not for sale. If you have specific
questions I can probably answer them.
After calibration, these are usually within 0.4 ppm after 30 minutes
warmup. The accuracy is a function of the crystal clock in the 5300A/B
display and these displays drift up (age) 0.1-0.2 ppm/year, making
their readings go low at about that rate. These are rather old now, so
the aging rate has significantly slowed down. They are nice counters.
I have several of this series (5300-5312) for sale.
I believe they were sold from the early 70s to the early 80s.
Jim
-
-----------------------------------------------