About a watt, peak. Don't want to fry the baby's eyeballs if it's
presenting face forward, or burn the mum's skin.
>
> You could almost do time resolution of depth. Use a gain-switched
> laser and a fast receiver.
You need a lot of detector area, which makes fast receivers hard unless
you use something like a microchannel plate PMT. The light is diffuse,
so you can't usefully play concentrator tricks to reduce the area.
> Maybe average one photon per shot and use
> an avalanche-type detector. That might be worth thinking about. I
> guess the velocity of light drops in a scattering medium like a baby.
It's very much analogous to carriers propagating by diffusion in
semiconductors--on the order of c*(mean free path/propagation
distance)**2, with a factor of order unity to take account of 3D
propagation.
> Does pulse ox use multiple wavelengths?
Usually two, often 600- and 900-nm-ish, depending on available LED
wavelengths. You can see the Hb/Hb02 absorption spectra at
<
https://medicine.uiowa.edu/iowaprotocols/pulse-oximetry-basic-principles-and-interpretation>.
It also helps a lot with rejecting room lights and stuff. You
typically phase-lock to the pulse rate for noise rejection purposes.
(This requires some smarts because the jitter is very large.)
The blood spot detector we did some years ago also sensed haemoglobin,
but needed to reject a huge overlapping absorption from the brown-egg
pigment protoporphyrin--like 10,000x stronger than the detection
threshold, which is 80 dB electrical. (That's looking for a 1%
absorption signal in the presence of a two-AU background, i.e. 99% peak
absorption from the protoporphyrin.)
That one worked by picking another wavelength on the opposite shoulder
of the protoporphyrin peak, so that the two went up and down together as
a function of egg colour, and then take one minus the ratio. We didn't
invent the scheme, but we did come up with some new calibration and
post-processing wrinkles that allowed the wavelength tolerance of the
filters to be relaxed quite a lot, which helped reduce costs. (Just
getting rid of the usual flashlamps and PMTs was the main cost-reduction
freight.)
The challenge of the present project is more in the SNR and background
rejection areas.
>
> It's nice to do stuff that helps people.
Yup. Especially mums and babies. My younger daughter is planning to
get married in the fall, so grandchildren are a strong possibility in
the next few years.
Cheers
Phil Hobbs