In our case, these would be the scan coils, they diverge the
beam on the 1 mil titanium sheet window that keeps the vacuum in
check but is thin enough for the electrons to pass/displace.
>
> focusing coil:
> ebay search: "EM-PM FOCUS COILS RLF-028 & RLF-029"
The focus coils sit above the scan horn and scan coil section.
They are responsible for confining the beam to a focal point on
the target, the titanium window in this case. THe focus is calibrated
to not produce the smallest point, otherwise, we'd be creating holes
and enjoy nice implosions from the near perfect vacuum.
> also a thermionic source from an electron beam welder that might be
> good for EBM:
> ebay search: "Hairpin EB Welder Filaments HSD W-2 Leybold 12 Mil"
>
> Sorry for not putting the ebay url's but I got an error from the url's
> being too long.
>
> cheers,
> Jamie
Our system uses 100-200 hz of electron beam scanning. Steering coils
are in place on some of them to avoid structures that tend to miss align
the beam that is not part of the scanning. Also, These are employed with
what is called wobble coils to oscillate the beam using 60 hz. This
prevents burning of the window material.
At the tips of each scan, on the outer edges of the window (horn),
Trapezoid peeks are added in the scan signal to help start the beam
back on its return path sooner. The theory here is, the beam spends more
time on the edges as it gets ready to scan on the return and thus may
heat that section of the window more. So, this wave form was
added to help this out however, from what I've seen over all, it does
not seem to do much for it. The signal at the peeks are getting lost in
the induction of the scan coils.
If you need more specifics, I can supply more but this was just a
general break down. As for cooling, we use chilled water that needs to
be circulated around the drift tube, aperture and scan horn sections.
With out this, thermo distortion in the structure would cause steering
problems and heat would cause vacuum leaks.
Jamie