I know that the conductivity of 0.05 M KCl is 6660 uS and 0.01 M KCl is 1414
uS
I recently bought a Corning CD30 conductivity meter for use in my
greenhouse. I cost about $60. It seems to have the proper conductivity range
for you. It came with calibration solutions. With it, you should be able to
make candidate solutions and measure them. Then calculate how much salt or
water you need to get your desired conductivity. Do that and recheck your
conductivity.
Bill
Krassi
K+: 73.50 Scm^2/mol
Na+: 50.10 Scm^2/mol
Cl-: 76.35 Scm^2/mol
For example KCl:
250 uS/cm = [KCl]*(73.50 + 76.35) Scm^2/mol
[KCl] = 1.668e-6 mol/cm^3 = 1.668 mM = 0.124 g/l
The solution is so dilute that the molar ionic conductivities at inifinite
dilution can be used without error being too large.
Kohlrausch law is said to be valid for strong electrolytes, like KCl or
NaCl, i.e. the molar conductivity of salt would decrease as a fuction of the
square root of the concentration:
L = L0 - K*sqrt(C)
but actually such a plot is not a very good straight line. (I took the
conductance data from CRC Handbook of Chemistry and Physics).
--
Lasse Murtomäki
Helsinki University of Technology
Lab. Phys. Chem. Electrochem.
lasse.m...@hut.fi