Notes on RFI results via FlowCal Excel UI
=========================================
* NIST Gaithersburg and NIST Stanford need to employ better SSC thresholds
* event lists are dominated by non-cell events
* center beads in FSCvSSC
* FSC and SSC channels are independent of fluorescence channels; we can
change the voltage as needed to make beads and cells appear cleanly in FSC
and SSC while not affecting fluorescence channels (which need to be
consistently measured at the same voltages between beads and samples)
* NIST Stanford beads appear to be saturated or near saturated in SSC
* DeLateur beads appear to be saturated or near saturated in FSC
* area ("-A") vs. height ("-H") for event quantification
* "-A" works "better" for DeLateur beads
* -H works "better" for NIST Gaithersburg and NIST Stanford beads
* harder to say for samples. Comparing Arithmetic Mean of samples measured
via "-A" and "-H" resulted in *very* similar values, though.
* looks like NIST Stanford doesn't have BL2 (LP0049) as originally thought
* cautiously optimistic / excited because I think there's signal in all the
places where we wanted signal. Going to take some gating magic to clean
everything up, but I think it can be done.
* DeLateur pL2f1091 (green) FlowCal Excel UI Sample IDs:
* DL007 - DL009
* DL019 - DL021
* DL031 - DL033
* DL043 - DL045
* DL055 - DL057
* DL067 - DL069
* DL079 - DL081
* ^^^ all show decent, clean, HIGH signal
* NIST Gaithersburg pL2f1091 (green) FlowCal Excel UI Sample IDs:
* -A
* plate 1
* NG012 - don't see anything
* NG046 - don't see anything
* NG080 - don't see anything
* plate 2
* NG029 - minor secondary HIGH population on BL1-A
* NG063 - *very* minor secondary HIGH population on BL1-A. Can't tell
about BL2-A.
* NG097 - can't tell on BL1-A. Minor secondary peark on BL2-A.
* -H
* plate 1
* NG114 - see *very* minor secondary HIGH population in BL1-H
* NG148 - see *very* minor secondary HIGH population in BL1-H and BL2-H.
BL2 is lower than BL1.
* NG182 - see *very minor secondary HIGH population on BL2-H.
* plate 2
* NG131 - see clear minor secondary HIGH population in BL1-H.
* NG165 - see clear minor secondary HIGH populations in BL1-H and BL2-H.
* NG199 - see clear minor secondary HIGH populations in BL1-H and BL2-H.
* NIST Stanford pL2f1091 (green) FlowCal Excel UI Sample IDs:
* -A
* NS_013, NS_017, NS_021
* NS_037, NS_041, NS_045
* NS_061, NS_065, NS_069
* -H
* NS_085, NS_089, NS_093
* NS_109, NS_113, NS_117
* NS_133, NS_137, NS_141
* ^^^ all show minor secondary LOW peak which is likely non-cell events
and primary HIGH peak which is expected GFP signal
* density gating works better on -H FSCvSSC--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytometry+unsub...@googlegroups.com.
To post to this group, send email to sbsc-flow-cytometry@googlegroups.com.
Visit this group at https://groups.google.com/group/sbsc-flow-cytometry.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/CACfa_%3DbZi9tsoirRJRr8MW3Eac3rYFe%2BEWgMNvnKdaps2Xqe2A%40mail.gmail.com.
For more options, visit https://groups.google.com/d/optout.