I made a few small changes:
The PBS should be filtered even if it is freshly made.
I reduced the collection volume to 160 uL instead of 175 uL. I double checked, and the dead volume on the Attune is 30 uL, so acquiring 160 is close to the max – and that leaves a bit in the well to make sure we don’t pull in bubbles.
If everyone else agrees with these changes, we are ready to perform.
Also, next Monday, I will be at a conference, so I will miss the phone Google Hangout-
Thanks-
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/D62098E3-911B-4C26-BFDC-6D7BC395CEDA%40stanford.edu.
I made a few small changes:
The PBS should be filtered even if it is freshly made.
I reduced the collection volume to 160 uL instead of 175 uL. I double checked, and the dead volume on the Attune is 30 uL, so acquiring 160 is close to the max – and that leaves a bit in the well to make sure we don’t pull in bubbles.
If everyone else agrees with these changes, we are ready to perform.
Also, next Monday, I will be at a conference, so I will miss the phone Google Hangout-
Thanks-
From: sbsc-flow-cytometry@googlegroups.com [mailto:sbsc-flow-cytometry@googlegroups.com] On Behalf Of Ariel Haim Hecht
Sent: Thursday, February 02, 2017 2:03 PM
To: sbsc-flow-cytometry@googlegroups.com
Subject: Re: [sbsc-fc] Follow-on from today's meeting
I am ready to perform.
-Ariel
On Jan 30, 2017, at 2:27 PM, Jake Beal <jake...@gmail.com> wrote:
Everybody who will be involved in the experiment, please review it in detail. When you are satisfied, write and say you are ready to perform. When everybody confirms, we will proceed to coordinate calendars for shipments.
Thanks,
-Jake
--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytometry+unsub...@googlegroups.com.
To post to this group, send email to sbsc-flow-cytometry@googlegroups.com.
Visit this group at https://groups.google.com/group/sbsc-flow-cytometry.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/CALR99W2GoHo%2BRf8W5_0yHN2h3RimZO_tuW6gK6HNyZkRhEnbeg%40mail.gmail.com.
For more options, visit https://groups.google.com/d/optout.
--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytometry+unsub...@googlegroups.com.
To post to this group, send email to sbsc-flow-cytometry@googlegroups.com.
Visit this group at https://groups.google.com/group/sbsc-flow-cytometry.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/D62098E3-911B-4C26-BFDC-6D7BC395CEDA%40stanford.edu.
For more options, visit https://groups.google.com/d/optout.
--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytometry+unsub...@googlegroups.com.
To post to this group, send email to sbsc-flow-cytometry@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/BY1PR09MB04532D69C8D229545C06AFAFE74C0%40BY1PR09MB0453.namprd09.prod.outlook.com.
My understanding of the protocol is that we're supposed to do one technical replicate for each voltage, and then repeat the whole process 3 different times (e.g. on three different days). So, the three wells of 200 uL each are one well for each of the three voltages. If that is not clear, then we should edit the document so that it is.
Also, what is the measured carry-over rate that you get with zero focusing fluid blanks?
Thanks-
Hi Jake.I don't see the reason for the technical replicates at each voltage here. It's not testing much of anything having 3 wells of 200 uL of the exact same sample, much like you wouldn't test anything more having 6 wells of 100 uL each would. Collecting at different voltages seems more interesting, especially since the math doesn't really work out if you want three technical replicates of three different voltages. This would be 200x9 = 1800 uL of cells. The protocol only gives you 1000 uL of cells. If I had to choose one set of three I'd much rather then different voltages, but I'm happy to know thoughts on this.
As far as what strains I'll be sending, I don't have a good two-colour. Getting rid of positional or resource or proximity or context effects is not something that's been overcome, and the sample I sent around last time is too strong an expressor and it results in a negative population we've seen. Instead I'll send a Strong and a Medium for GFP and a Strong and a Medium for mCherry and an Empty Vector. These 5 should be comparable and interesting and more well-behaved than my "what do I have lying around" batch last time. I will include the maps in a folder in the Interlab and a key in the experimental workflow (I promise to try my best not to mislabel one like last time).
Also as far as the analysis goes, we all do it a bit differently with our respective programs that do roughly the same things, just wanna make sure that's clear there; for example I don't use a Gaussian mixture model but rather a combination of SSC-A, FSC-A, and FSC-H, as well as the other steps with the Cytoflow package. So the analysis is somewhat a different beast than the data collection.
For shooting blanks between samples, this section should be variable. I need 0. Some people from my understanding need more than 1. I prefer the wording "Place as many focusing/sheath fluid blanks your machine requires before each sample. If you don’t know what this number is, alert the group ASAP and we will help you figure it out". I distinctly remember some cytometers being discussed requiring 4 or 5?Shoot for as close as you can to 0.5 uL/second flow rate (record actual)
--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytom...@googlegroups.com.
To post to this group, send email to sbsc-flow...@googlegroups.com.
Visit this group at https://groups.google.com/group/sbsc-flow-cytometry.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/CALR99W0RbJse1%3DjRpN5XtdEfoYKKwSq%3DzsTJtyN_9egET-6z4A%40mail.gmail.com.
I changed some things in the Experimental Design section in an attempt to organize/clarify it (in a manner which hopefully didn't significantly change the meanings of previous statements). I also added a bunch of comments (apologies ahead of time for being so nit-picky...). Once those are addressed, we (Rice / John+Sebastian) are on board. (I have not had a chance to review the analysis section yet, though)
Yes. We confirm. That week works for NIST Gaithersburg.
Thanks-
Yes. We confirm. That week works for NIST Gaithersburg.
Thanks-
From: sbsc-flow-cytometry@googlegroups.com <sbsc-flow-cytometry@googlegroups.com> on behalf of Jake Beal <jake...@gmail.com>
Sent: Thursday, February 9, 2017 6:30 AM
Subject: Re: [sbsc-fc] Follow-on from today's meeting
Also, since I haven't seen email go out about it yet:
In our meeting on Monday, we proposed for the experiment to be run on the week of February 20 - 24, starting Feb 21 or 22 (depending on whether Nicholas can ship on President's day or not). These dates are confirmed to work for MIT, NIST Stanford, and Rice. NIST Gaithersburg, can you please confirm as well?
Also, if any lurkers on the list wish to participate, speak up and you will be welcome to join. :-)
Thanks,-Jake
On Thu, Feb 9, 2017 at 5:27 AM, Jake Beal <jake...@gmail.com> wrote:
On Thu, Feb 9, 2017 at 4:02 AM, John Sexton <john.t...@rice.edu> wrote:
I changed some things in the Experimental Design section in an attempt to organize/clarify it (in a manner which hopefully didn't significantly change the meanings of previous statements). I also added a bunch of comments (apologies ahead of time for being so nit-picky...). Once those are addressed, we (Rice / John+Sebastian) are on board. (I have not had a chance to review the analysis section yet, though)
Please do *not* apologize, John. The whole point is to make sure everybody is acting as equivalently as possible, and queries like you're making are absolutely required to ensure that. I've resolved some and added responses to some, for your perusal. Others require input from Nicholas or other lab folks.
Thanks,-Jake
--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytometry+unsub...@googlegroups.com.
To post to this group, send email to sbsc-flow-cytometry@googlegroups.com.
Visit this group at https://groups.google.com/group/sbsc-flow-cytometry.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/CALR99W0xazZBNtX%2BO-N8QKk82qCegMA5KsDPVAF5oOu57ScOdw%40mail.gmail.com.
For more options, visit https://groups.google.com/d/optout.
--
You received this message because you are subscribed to the Google Groups "SBSC Flow Cytometry Working Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sbsc-flow-cytometry+unsub...@googlegroups.com.
To post to this group, send email to sbsc-flow-cytometry@googlegroups.com.
Visit this group at https://groups.google.com/group/sbsc-flow-cytometry.
To view this discussion on the web visit https://groups.google.com/d/msgid/sbsc-flow-cytometry/BY1PR09MB045301E84BD8185ACF9487DEE7450%40BY1PR09MB0453.namprd09.prod.outlook.com.