So for a formal followup, the 2016 interlab folder has been reorganized a bit, but has all of the data we have collected to date. I have duplicated the spherotech spectra csv and added it as an xlsx sheet with some manually scaled plots.
In followup to our conference call—and some more careful thinking about the data we have and what we're trying to do—I'd like to post the following corrections/hypotheses:
- We can use the spectra from the beads and the fluors—for any available excitation wavelength—to convert measured a.u. signal values to "bead units" for all band-passes for that excitation line. This would put all measurements for a given laser onto the same scale, albeit it still arbitrary and unit-less.
- Because the beads are excitable at all wavelengths, and emit across 'all' wavelengths—and also because the emission profile does not appear to change significantly as excitation wavelength increases—we can(?) assign MEFL units to the measured bead peaks in all channels, for all laser lines.
- With this, it's a simple transformation to take all channels for all lasers and place measured fluors onto the rainbow bead MEFL scale.
I *currently think* this works for the same reason that spillover is calculated as a linear contribution to the target channel... but I'm not too confident in this idea yet. My current big question is whether we can ignore the differences in "excitability" for the reference particles and the fluors for all excitation wavelengths.
Peter
Why? Well, glad you asked. I believe that these emission spectra allow us to establish the relationship between the fluors and the beads through different band-passes.