Hi, Melchor!
First of all, thank you very much for this amazing unit and incredible support. SARK-110 really adds significant capabilities to electronics lab, especially amateur.
Question #1
The question is rather strange, I know. The absolute of reflection can not be higher than 1, yet it possible that due to maybe not perfect match load calibration standard,
the measured reflection can actually be higher than 1 (noise and some small matching error). So as I can get this far - the SARK replaces such S11>0dB values with
exact 0 dB (well 10^-15 or something in PLOTS, for us - exact zero).
I noticed, that when I perform an OSL calibration, the Short Load measure bars appear in red and reflection coefficient shows something like |Rho| = 0.92 and Phase(Rho) ~ -175.5 (Should be 1 and -180).
OSL Open load calibration values seems to be correct: 1.0 and -0.1. Match load values are also looks fine: 0.05 < -56 (yes, ideal is 0 and undetermined, bet hey, these values looks OK for an experiment)
So, the question is this due to SARK measuring algorithm (force values >0 to be exact zero) or maybe it is due to improper detector calibration (signal goes over the range of ADC maybe)?
Detector calibration values are the following (the path is SARK MCX -> little SARK default cable -> BNC -> Adapter to SMA -> calibration standards):
Short Load:
Voltage ~61000
Current ~116650
Open Load:
Voltage ~97000
Current ~115
Match Load:
Voltage ~75170
Current ~73050
The results of measured standards after the OSL calibration lools like this:
Short load (in the range <70 MHz the values are zeroed, phase looks OK though):

Open load (looks reasonable):

Nothing connected (a.k.a. open load with different offset length, hence non zero phase) in the region >120 MHz again these zeroed values:

Match load (looks fine to me):

Question #2
The SARK OSL calibration is performed with 400 measurement point, but then in SARK plots it is possible to set arbitrary number of points, even significantly more than there was during calivration.
How does SARK determines the values of calibration coefficients for frequency points where there are no cal data (between calibrated frequency points), interpolation?
Question #3
During OSL calibration, does SARK assume that the calibration standards are ideal and with zero offset? In other words if I have a SHORT load with non zero length from the measurement plane
(where the phase is supposed to be -180 for all frequencies) to the actual short contact, that means that SARK doesn't take into account such offset?
Thank you!
Dimitri.
P.S. Can not wait to see two-ports SARK with extended frequency range and SMA connectors! :)