I'm using the CATphan on the CT and some months it "sees" the couch and other months it does not.  Is there a remove couch option?

I included some screen shots from the report.  you can see the couch included and hot it affects the uniformity graph.

Thanks,

Steve


image003.png
image001.png
Unknown.png
image002.png

Stephen Amadon

unread,
Nov 20, 2025, 2:04:20 PM11/20/25
to Pylinac
I've played around and changed my FOV of the scan and it adjusted the results to remove the couch.  Interstling, the MTF for line pairs increases from .25 to .35.  This difference is what made us look into why the results looked differently.

Attached are the 2 different reports to show the difference.

Thanks,

Steve
catphan_CT.pdf
catphan_CT_FOV.pdf

tbe...@gmail.com

unread,
Feb 5, 2026, 2:56:34 AMFeb 5
to Pylinac
Hi Stephen,

MTF has to change if you change the FOV, since the numer of pixels for each slice is always 512 or 1024 pixels (or whatever your settings are) for each dimension. If you half the FOV, the resolution doubles.