Hi Arjun,
Apologies for the delay getting back to you - your message ended up in my spam folder for some reason and I only saw it when visited the google group page.
I assume your talking about using the PMTs with pyPhotometry rather than pyControl. The factors that will determine whether a photodetector is suitable for use with pyPhotometry are:
1. Sensitivity - can it detect the low light levels of photometry signals at the wavlengths you will be working with. I guess the answer is yes for the PMTs are they are normally very sensitive and the spec sheet says they work from 185-900nM wavelength.
2. Bandwidth - does the output signal respond fast enough to changes in light intensity to work with the time division illumination modes. The bandwidth of the newport photodetectors in DC mode is ~0-750Hz so bandwith this wide or more will be fine. The PMT and amplifier have much wider bandwidth (0-60MHz) so that is fine.
3. Output voltage - the pyPhotometry board can acquire signals from 0-3.3V, and although they have protection diodes to protect from damage against overvoltage inputs, a signal far outside this voltage range might risk damaging something. The PMT amplifier looks fine here because although the default output voltage is -1.5-1.5V, there is a DC bias control that you can use to shift it to 0-3V which is idea.
Overall they look like they should work, though I have never tried this type of photodetector myself.
best,
Thomas