Error message

300 views
Skip to first unread message

Nina Owczarek

unread,
Oct 26, 2012, 11:51:26 AM10/26/12
to px...@googlegroups.com
Dear All,

Today I received an error message that interrupted my collection and shut down the S1PXRF software. It read: Run time error overflow 6
Can anyone explain this error to me?

I am using a Bruker SDIII. It was set in "lab rat" mode (i.e. 40kV, 13mA, no filter). I was looking at blue pigment on a thick wooden substrate.

Thanks so much for any thoughts on this.
Best,
Nina

Vanessa Muros

unread,
Oct 26, 2012, 1:17:08 PM10/26/12
to px...@googlegroups.com
Hi Nina,

I've been getting this error recently as well, but I thought it might haveto do with software trouble I've been having with a new laptop we are using with our XRF (the Tracer III-V+).  Now that I see you are having the same error message, I'm curious to see what is causing it.

Vanessa


--
--
This message comes from the pXRF for Cultural Heritage group.
 
To read discussions non-members can join the group at:
http://groups.google.com/group/pxrf
 
All group activities can be accessed by members through its website:
https://sites.google.com/site/pxrfgroup/
 
 

Dennis Piechota

unread,
Oct 26, 2012, 1:45:35 PM10/26/12
to px...@googlegroups.com
I too get the overflow error repeatedly and I don't know what causes it (looking up the error definition is no help unless you live to program) but in looking back it seems to me that it occurs most often:
1- when I try to use the software to do two things at once e.g. doing an element ID while a reading is being taken;
2- when I do NOT religiously follow the hardware/software start up instructions.

Dennis

Dennis Piechota
Archaeological Conservator
Fiske Center for Archaeological Research
UMass Boston
Office: 617-287-6829
RTI Group Admin

pant...@aol.com

unread,
Oct 29, 2012, 2:26:19 AM10/29/12
to px...@googlegroups.com
That error occurs occasionally when Microsoft begins an update or save operation in the middle of data acquistion.
Be sure to not have any other operations going on during data acqusition
You just restart the software and repeat the acquisiion
Dr Bruce Kaiser
A brilliant old fellow seeking the Truth in every thing
Call him any time to discuss 509 528 7608
Or visit him in Salt Lake UT

 

Dennis Piechota

unread,
Oct 29, 2012, 11:47:57 AM10/29/12
to px...@googlegroups.com
Hi Bruce,
To be clear are you saying that common background operations like automated virus database updates of software like AVG or Windows update checks could trigger the overflow error if they occur during a reading and need to be disabled or set to perform manual update checks only?

And within S1PXRF can we expect to be able to begin doing element IDs during a read without triggering the overflow?

Dennis

Dennis Piechota
Archaeological Conservator
Fiske Center for Archaeological Research
UMass Boston
Office: 617-287-6829
RTI Group Admin



Bruce Kaiser

unread,
Oct 29, 2012, 2:51:21 PM10/29/12
to px...@googlegroups.com, px...@googlegroups.com
Dennis
These items do not in general cause the   Error
It is just when a very random set of them come together when you are doing certain pxrf operations 
Interesting to note when is using Apple it never crashes :)

We are investigating windows 8
It appears Microsoft has fixed a lot of issues with that platform 
Time will tell
Thanks
Dr Bruce Kaiser
--
-----
------------
+ This message comes from the pXRF for Cultural Heritage group.
+ To read its discussions non-members can join the group at:
http://groups.google.com/group/pxrf/
+ Members can post to the group by sending to px...@googlegroups.com
+ To unsubscribe send an email to pxrf+uns...@googlegroups.com
+ All group activities can be accessed by members through its website:
https://sites.google.com/site/pxrfgroup/
------------
-----
 
 

MacKenzie, Mark, DCA

unread,
Oct 30, 2012, 9:48:35 AM10/30/12
to px...@googlegroups.com
Good day all.  Just a small note on this thread and one question.

1.  After losing a MuSIS system due to an IT background "system update" requiring a trip to Holland to get both a hardware update and a software system fix, I do not allow our PC computers which control analytical equipment to be part of our state IT network.  So far, no further problems.

2.  Referring to the above, I believe I have been able to do element IDs within S1PXRF on our factory supplied Bruker system laptop without every causing a crash or lock up.  That laptop is never connected to our IT network or at least not when being used to control the Tracer unit.

3.  Now for my question.  Bruce, you tantalized me with your reference to Apple.  Do you have a version of S1PXRF which runs on my MacBookPro's or are you using a Windows system on the Macs to do this?

Regards from Santa Fe.

Mark MacKenzie, Chief Conservator
Dir. Museums of New Mexico Conservation
Santa Fe, NM
505-476-5081

From: px...@googlegroups.com [px...@googlegroups.com] on behalf of Bruce Kaiser [pant...@aol.com]
Sent: Monday, October 29, 2012 12:51 PM
To: px...@googlegroups.com
Cc: px...@googlegroups.com

Subject: Re: [pxrf] Error message

Dennis Piechota

unread,
Nov 1, 2012, 11:03:18 AM11/1/12
to Dennis Torres, px...@googlegroups.com, B Kaiser, Bruce Kaiser, HH Service
Thanks Dennis T.,
I checked my version and it was already uptodate (3.8.0.30). BTW I'm using the S1PXRF installation on the HP laptop supplied with my Tracer IIISD. So I deleted the .Mca file and restarted Windows XP Pro and restarted the analyzer and software per the instructions (Quick Startup List for Tracer IIISD). It worked well for several hours with me taking computer controlled timed readings while simultaneously doing Elem IDs on the developing spectra. Then, as though some necessary file became corrupted the overflow errors (Runtime error 6 overflow) began again. I was IDing fewer than 10 elements per spectrum when the overflow errors started again.

I repeated the recommended process (deleting .Mca, restarting etc.) and had the same experience again.

My most reliable performance from S1PXRF comes when I dedicate the software to just taking the readings and do not attempt element IDs until the readings are complete.

Dennis P.

Dennis Torres

unread,
Nov 6, 2012, 12:22:50 PM11/6/12
to Dennis Piechota, px...@googlegroups.com, B Kaiser, Bruce Kaiser, HH Service

Hi Dennis P.,

 

Thanks for this information you have provided.  This is a trend we have seen as well.  Can you tell me what system you are using (Windows 64 bit? Windows 7?).  Thanks.

 

Best Regards,

 

 

 

Dennis Torres
Customer Support Specialist, Handheld XRF

 

Bruker Elemental

415 N. Quay Street

Kennewick, WA 99336

USA

Phone:    +1 509 783-9850 x268

Fax:         +1 509 735-9696

dennis...@bruker-elemental.net
www.handheldxrf.com

 


 

 

 

From: deep...@gmail.com [mailto:deep...@gmail.com] On Behalf Of Dennis Piechota
Sent: Thursday, November 01, 2012 8:03 AM
To: Dennis Torres; px...@googlegroups.com
Cc: B Kaiser; Bruce Kaiser; HH Service
Subject: Re: [pxrf] Error message

 

Thanks Dennis T.,

IMPORTANT NOTICE: This e-mail message and/or attachments to this e-mail message may contain information that is confidential and/or subject to legal privileges. IF THIS E-MAIL WAS SENT TO YOU IN ERROR: Please immediately inform the sender by sending a reply e-mail, and then immediately delete this e-mail and any attachments to this e-mail. Please do not read, print, copy, save, disclose or distribute this e-mail or any attachments. Any review, printing, copying, retention, disclosure, distribution or use of this e-mail or any attachments is strictly prohibited.

 

IMPORTANT NOTICE: This e-mail message and/or attachments to this e-mail message may contain information that is confidential and/or subject to legal privileges. IF THIS E-MAIL WAS SENT TO YOU IN ERROR: Please immediately inform the sender by sending a reply e-mail, and then immediately delete this e-mail and any attachments to this e-mail. Please do not read, print, copy, save, disclose or distribute this e-mail or any attachments. Any review, printing, copying, retention, disclosure, distribution or use of this e-mail or any attachments is strictly prohibited.


David Dempsey

unread,
Nov 6, 2012, 1:18:48 PM11/6/12
to px...@googlegroups.com
Hi Dennis,
We are still on Windows XP.
I wondered whether that might be an issue.
Thanks,
David

--
-----
------------
+ This message comes from the pXRF for Cultural Heritage group.
+ To read its discussions non-members can join the group at:
http://groups.google.com/group/pxrf/
+ Members can post to the group by sending to px...@googlegroups.com
+ To unsubscribe send an email to pxrf+uns...@googlegroups.com
+ All group activities can be accessed by members through its website:
https://sites.google.com/site/pxrfgroup/
------------
-----
 
 



--

David Dempsey
Associate Director for Museum Services
Smith College Museum of Art
Northampton, MA 01063 USA
ddem...@smith.edu
phone: 413-585-2766
fax: 413-585-2782

Bruce Kaiser

unread,
Nov 9, 2012, 4:23:58 PM11/9/12
to px...@googlegroups.com, px...@googlegroups.com
The software only runs on windows
So yes you must be running windows on the Mac


Thanks
Dr Bruce Kaiser

Dennis Piechota

unread,
Nov 12, 2012, 11:24:23 AM11/12/12
to px...@googlegroups.com
Hi Dennis T. (and all),
Sorry for answering so slowly. I missed the question. I run the latest version of S1PXRF under Windows XP Pro v2002 with SP3, 32-bit, on the HP Compaq 6510b laptop that was provided with my Tracer SDIII.

Another question. We've focussed on the need to follow carefully the quick start routine. Is there also a quit procedure that should be followed? Specifically should we clear all spectra before quitting S1PXRF? Or at least make sure the B spectrum is clear? I ask because I've noticed there are times when I launch S1PXRF by clicking on the S1PXRF.exe icon and it opens usually with a previous spectrum already loaded. To follow the quick start procedure completely I clear the spectrum (when one has loaded) by going to PDZ Preview and open an existing spectrum. 

I notice that though I quit with two spectra visible with spectrum overlay set to A/B, S1PXRF.exe re-opens with overlay in A only and with the previous A spectrum visible. If I open spectrum overlay and try to move that spectrum to B I can get the Runtime error 9- subscript out of range error after pressing the OK button.

Dennis Piechota
Archaeological Conservator
Fiske Center for Archaeological Research
UMass Boston
Office: 617-287-6829
RTI Group Admin



Jeff

unread,
Dec 12, 2012, 2:45:45 PM12/12/12
to px...@googlegroups.com
I had this error at the end of October/early November. I tried running the XRF on several of my computers and consistently got the same error--very frustrating! I then removed the USB to serial adapter from the cable and connected the instrument directly to the serial port (I have an older laptop). I haven’t had any errors since (fingers crossed).

Orion Analytical (gmail)

unread,
Apr 17, 2013, 9:08:08 AM4/17/13
to px...@googlegroups.com
Hi All,

I have recently encountered zirconium (Zr) in pXRF analyses of a group of paintings attributed to different artists.

I would be grateful to know whether -- and in what contexts -- other members of this group have found Zr in XRF or other elemental analyses of paintings or painted surfaces.

All the best,

Jamie

James Martin
Orion Analytical, LLC - A Materials Analysis and Consulting Firm
www.orionanalytical.com
mar...@orionanalytical.com

(sent using our Gmail alternate account)





Glenn Gates

unread,
Apr 17, 2013, 10:09:06 AM4/17/13
to px...@googlegroups.com, Cathy Selvius DeRoo
Chris at MoMA put Zr paints in tests in the Round Robin of paint samples circa 2008ish
I'd direct your query there.
As I recall, he sees it lots
--
-----
------------
+ This message comes from the pXRF for Cultural Heritage group.
+ To read its discussions non-members can join the group at:
http://groups.google.com/group/pxrf/
+ Members can post to the group by sending to px...@googlegroups.com To
+ unsubscribe send an email to pxrf+uns...@googlegroups.com All
+ group activities can be accessed by members through its website:
https://sites.google.com/site/pxrfgroup/
------------
-----
---
You received this message because you are subscribed to the Google Groups "pXRF for Cultural Heritage" group.
To unsubscribe from this group and stop receiving emails from it, send an email to pxrf+uns...@googlegroups.com.
For more options, visit https://groups.google.com/groups/opt_out.


Aaron Shugar

unread,
Apr 17, 2013, 12:31:29 PM4/17/13
to px...@googlegroups.com, Cathy Selvius DeRoo
Ok,

This is a real problem that I have encountered and I think I have it figured out. There are two potential interference that can make one think there is Zr present.  1st is the Strontium Kb peak that is at 15.84 keV (Zr is at 15.78).  The second is more plausible with paintings and it deals with higher concentrations of Pb.  The L-1 absorption edge is at 15.86keV and is not labeled as Pb in the Artax software (for those using Bruker systems). See  http://skuld.bmsc.washington.edu/scatter/AS_periodic.html and http://www.kayelaby.npl.co.uk/atomic_and_nuclear_physics/4_2/4_2_1.html for examples).

Remember, there should be a Zr Kb peak to confirmation.

Hope that helps

Aaron

--
Aaron Shugar
aaron....@gmail.com
shug...@buffalostate.edu


Inline image 1

Shugar, A. and Simmons S. (2013)

Archaeometallurgy in Mesoamerica: Current Approaches and New Perspectives

http://www.upcolorado.com/book/New_Titles/Archaeometallurgy_in_Mesoamerica_cloth

 

Shugar, A. and Mass J. (2012)

Handheld XRF in Art and Archaeology

http://upers.kuleuven.be/en/titel/9789058679079

Orion Analytical (gmail)

unread,
Apr 17, 2013, 1:44:24 PM4/17/13
to px...@googlegroups.com
Thank you Aaron.

In the various paint spectra, I have a Zr Ka peak at 15.74 Kev and Zr Kb at 17.67 Kev.

In these spectra, other detectable elements vary by element and abundance -- frequently significant amounts of zinc (Zn) -- including some spectra that show no detectable lead (Pb) or strontium (Sr).  In spectra that contain Pb or Sr peaks, Zr Ka and Kb peaks are evident.

I understand that Zr can occur as an oxide as a white pigment, as a coating on certain pigments, and as a paint drier.

--
-----
------------
+ This message comes from the pXRF for Cultural Heritage group.
+ To read its discussions non-members can join the group at:
http://groups.google.com/group/pxrf/
+ Members can post to the group by sending to px...@googlegroups.com
+ To unsubscribe send an email to pxrf+uns...@googlegroups.com
+ All group activities can be accessed by members through its website:

https://sites.google.com/site/pxrfgroup/
------------
-----
---
You received this message because you are subscribed to the Google Groups "pXRF for Cultural Heritage" group.
To unsubscribe from this group and stop receiving emails from it, send an email to pxrf+uns...@googlegroups.com.
For more options, visit https://groups.google.com/groups/opt_out.
 
 

Jillian Huntley

unread,
Apr 22, 2013, 9:22:01 PM4/22/13
to pXRF for Cultural Heritage
Hi Jamie - I'm not surprised. I see Zr in a lot of mineral pigments
(ochres) used in rock art.
In a few recent cases I have actually seen the zerconium crystals in
the ochre during SEM work.

Jillian Huntley
Archaeologist
PhD Candidate, Archaeology and Palaeoanthropology, University of New
England
e: jfo...@une.edu.au


On Apr 17, 11:08 pm, "Orion Analytical (gmail)"
<orionanalyti...@gmail.com> wrote:
> Hi All,
>
> I have recently encountered zirconium (Zr) in pXRF analyses of a group of paintings attributed to different artists.
>
> I would be grateful to know whether -- and in what contexts -- other members of this group have foundZrin XRF or other elemental analyses of paintings or painted surfaces.

Orion Analytical

unread,
Apr 23, 2013, 9:43:24 AM4/23/13
to px...@googlegroups.com
Thank you Jillian.

While I am detecting zirconium (Zr) in paints that contain iron oxides, I also am detecting Zr in other paints that (a) do not contain iron oxides and (b) do not lay over paints that contain iron oxides. For example, I am finding Zr in metallic paints and white paints, in some cases with detectable cobalt (Co).

The Zr (Ka and Kb) occurs in original paints from a group of works that purport to have been created by different artists in the late 1940s and 1950s. All the works are suspect and share a common source.

My working hypothesis is that the Zr stems from drier added to accelerate hardening of the paints.

I would be grateful to hear from anyone who has found Zr in documented paintings from the 1940s, 1950s, or 1960s.

Thank you.

All the best,

Jamie
> --
> This message comes from the pXRF for Cultural Heritage group.
> To read its discussions non-members can join the group at:
> http://groups.google.com/group/pxrf/
> Members can post to the group by sending to px...@googlegroups.com
> All group activities can be accessed by members through its website:
> https://sites.google.com/site/pxrfgroup/
> ---