Hello Vanessa,
I am not sure if you were able to find your answer yet or not, but I may be able to help with at least part of your question!
On the instrument itself, I do not believe it will show you the information you are looking for, more so if it is a calibration pre-installed on the device. One workaround is to take a .PDZ file completed in that calibration and open it in the BIT (Bruker Instrument Tools) software on you PC. When viewing the Spectra, you can then select Options in the upper left of the application, and click Spectrum Info. This will give you all of the information available for that scan. It is important to note that in most scans, they are dual-phase calibrations, so you will need to view both scans PDZs (PDZ:1 & PDZ:2 are how it shows up in BIT) that are listed under that PDZ. I will attach screenshots for clarity!
Note: This was a scan using the 5G and the PlantF1 calibration, but the same process will get you the information you will need for MUDROCK or any other calibration.
Hope this helps a little,
Chris