Dear Pedro,
in principle you are right: a change of the density table should not change the dose calculation with density-override. However, in reality there are some numerical problems:
1) Depending on the size of the ROI with density-override, you may observe minor changes in the source-surface distance. (This is also sensitive to the chosen OutsideAirThreshold.)
2) At least in earlier versions, the "density override" function worked very indirectly. For a voxel with "density override", first Pinnacle calculates the corresponding HU from the density table and sets this voxel to the override-HU. Then Pinnacle calculates the dose using the the original HU outside the override region and the override-HU in the override region. All these HU are converted to density by the density table; i.e., the "density override" voxels are converted from the override density to override-HU in a first step and from override-HU to density in a second step. This back-and-forth procedure depends on the interpolation between the calibration points in the chosen density table. (Tou may check this behavior by setting a density outside your density table - this should lead to an error message because the above back-and-forth procedure cannot work.)
I have found very minor changes for the calculated dose (<< 1%) as long as the density override occurs for densities close to 1.0 (where many well-defined points are found in my density tables), but somewhat larger changes (but still below 1%) for low densities (e.g., 0.5) or high densities (e.g., 2.5) where my density tables have only few points.
From my point of view, the 0.5% change in your case seems to indicate that at least one of your density tables has very few points (maybe, with some scatter) around your override density.
Best regards,
Peter
Dr. Peter Mohr
Strahlenschutzbevollmächtigter
Leitender Medizinphysiker
Diakonie-Klinikum Schwäbisch Hall gGmbH
Diakoniestraße 10
D-74523 Schwäbisch Hall
Tel. 0791 753 4845
Fax 0791 753 4911
Email:
Peter...@dasdiak.de<mailto:
Peter...@dasdiak.de>
www.dasdiak.de<
http://www.dasdiak.de>
________________________________
From:
pinnacl...@googlegroups.com <
pinnacl...@googlegroups.com> on behalf of Pedro <
palme...@gmail.com>
Sent: Tuesday, March 19, 2019 12:50 PM
To: pinnacle3-users
Subject: [p3rtp] CT-density table affecting the calculation when all the densities are overriden
--
--
You received this message because you are subscribed to the Google
Groups "pinnacle3-users" group.
To post to this group, send email to
pinnacl...@googlegroups.com
To subscribe to this group, send email to
pinnacle3-us...@googlegroups.com
To unsubscribe from this group, send email to
pinnacle3-use...@googlegroups.com
For more options, visit this group at
http://groups.google.com/group/pinnacle3-users?hl=en
---
You received this message because you are subscribed to the Google Groups "pinnacle3-users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to
pinnacle3-use...@googlegroups.com<mailto:
pinnacle3-use...@googlegroups.com>.
For more options, visit
https://groups.google.com/d/optout.
Diakonie-Klinikum Schwäbisch Hall gGmbH
Geschäftsführung: Michael Kilb
Registergericht: Amtsgericht Stuttgart, HRB 571831
Ein Unternehmen des Evangelischen Diakoniewerkes Schwäbisch Hall e.V.
Das Diakonie-Klinikum Schwäbisch Hall unterhält eine kooperative Partnerschaft mit dem Landkreis-Klinikum Crailsheim.
Diese Information ist ausschließlich für den Adressaten bestimmt und kann vertraulich oder gesetzlich geschützte Informationen enthalten. Wenn Sie nicht der bestimmungsgemäße Adressat sind, unterrichten Sie bitte den Absender und vernichten Sie diese Mail. Anderen als dem bestimmungsgemäßen Adressaten ist es untersagt, diese E-Mail zu lesen, zu speichern, weiterzuleiten oder ihren Inhalt auf welche Weise auch immer zu verwenden. Wir verwenden aktuelle Virenschutzprogramme. Für Schäden, die dem Empfänger gleichwohl durch von uns zugesandte mit Viren befallene E-Mails entstehen, schließen wir jede Haftung aus.