Hi Listers,

For Varian TrueBeam machines, Philips recommends a lower value than the Varian specification for the Maximum MLC Leaf Speed parameter, 2.25 cm/s versus 2.50 cm/s. The rationale is that this will help ensure VMAT plan deliverability. We feel that by doing so, one is limiting the optimizer solution space, i.e. fewer segments are available.

Has anyone used a value of 2.50 cm/s with success?

How about other conservative recommendations?

Maximum Gantry Angle Speed: 5.5 vs. 6.0 degrees/sec

Maximum Jaw Speed: 2.25 vs. 2.50 cm/s

Thanks in advance,

Sean Frigo

John Shakeshaft

unread,
Sep 27, 2016, 12:54:01 AM9/27/16
to pinnacl...@googlegroups.com

Hi Sean,

 

I am not familiar with the particular example of the TrueBeam, so the comments below are a general point.

 

In coming to a decision as to how to proceed, you need to consider carefully the information that is sent from Pinnacle to the linac, and how the linac then interprets or implements the request. In the DICOM RTPLAN (exported from any TPS) the information that is sent is the start-of-segment and end-of-segment positions of all components and the number of MU to deliver during the segment. The speed that Pinnacle predicts that the linac will use for a motion of a component (eg leaf speed/gantry speed) is *NOT* sent within the DICOM RTPLAN, and so may not be used in reality. On receiving the RTPLAN, the linac control software decides on to deliver the requested segment. There are obviously a number of solutions to the problem given gantry speed, leaf speed, dose rate and possibly jaw speed (if these move) can all be varied. In the past (it’s a while since I have done any detailed work on this), one solution that was adopted by some linacs was to deliver at the maximum gantry speed possible given dose-rate constraints. (Therefore the gantry was only slowed if it was not possible (because max dose rate was insufficient) to deliver the required MU within the segment at the maximum gantry speed.) In this case setting the *ratio* of max gantry speed and max leaf speed incorrectly in Pinnacle could lead to plans that were not deliverable (or at least generated a warning) on the linac due to leaf speed violations.

 

I suspect because of the many possible ways (with variable gantry speed/leaf speed/dose rate) of delivering X MU between point A and point B, you are not effectively limiting your solution space by using the Pinnacle recommended parameters (unless you have binned dose rate, which you should not).

 

Regards

 

John

 

Dr John Shakeshaft
Director - Therapeutic Physics | Radiation Oncology
Princess Alexandra Hospital |
Metro South Health
199 Ipswich Rd Woolloongabba QLD 4102 Australia

--
--
You received this message because you are subscribed to the Google
Groups "pinnacle3-users" group.
To post to this group, send email to pinnacl...@googlegroups.com
To subscribe to this group, send email to
pinnacle3-us...@googlegroups.com
To unsubscribe from this group, send email to
pinnacle3-use...@googlegroups.com
For more options, visit this group at
http://groups.google.com/group/pinnacle3-users?hl=en

---
You received this message because you are subscribed to the Google Groups "pinnacle3-users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to pinnacle3-use...@googlegroups.com.
For more options, visit https://groups.google.com/d/optout.

********************************************************************************

This email, including any attachments sent with it, is confidential and for the sole use of the intended recipient(s). This confidentiality is not waived or lost, if you receive it and you are not the intended recipient(s), or if it is transmitted/received in error.

Any unauthorised use, alteration, disclosure, distribution or review of this email is strictly prohibited. The information contained in this email, including any attachment sent with it, may be subject to a statutory duty of confidentiality if it relates to health service matters.

If you are not the intended recipient(s), or if you have received this email in error, you are asked to immediately notify the sender by telephone collect on Australia +61 1800 198 175 or by return email. You should also delete this email, and any copies, from your computer system network and destroy any hard copies produced.

If not an intended recipient of this email, you must not copy, distribute or take any action(s) that relies on it; any form of disclosure, modification, distribution and/or publication of this email is also prohibited.

Although Queensland Health takes all reasonable steps to ensure this email does not contain malicious software, Queensland Health does not accept responsibility for the consequences if any person's computer inadvertently suffers any disruption to services, loss of information, harm or is infected with a virus, other malicious computer programme or code that may occur as a consequence of receiving this email.

Unless stated otherwise, this email represents only the views of the sender and not the views of the Queensland Government.

**********************************************************************************

Reply all
Reply to author
Forward
0 new messages