The problem lies in exporting the Cone Beam CT to Mosaiq as now the image set does not match the plan. Changing the patient’s MR number is never a good solution to this problem and in many institutions is not possible.
Below is a work around that I’ve put together with the help of Philips. This assumes that a “copy” of the plan is created for CBCT export which has only the relevant structures included.
I hope it helps.
If the planning CT is sent to Pinnacle with an incorrect Patient ID # entered and it is not noticed before planning is completed, the CT images will not transfer to Mosaiq.
The following process was found to be an effective work around to this problem while avoiding the need to re-plan.
1) Ask the RTs to re-send the CT dataset from the scanner with the correct MR#.
2) Create a new plan using the newly imported scan as the primary dataset.
3) In the patient select window, click “edit” under the patient name.
4) Click “revision history”.
5) In the upper right-hand of the window, place the cursor in the “Directory” window.
6) The displayed path will look similar to this: /pinnacle_patient_expansion/NewPatients/Institution_4/Mount_0/Patient_1488
7) Using the arrow keys, move the cursor all of the way to the right.
8) Record the Institution number, Mount number, and Patient number.
9) Now open the new plan created in step 2.
10) Go to the contours tab>Options>ROI Spreadsheet
11) In ROI Spreadsheet go to File>Import ROI
12) Place the cursor in the “Directory” window. Select the displayed path and delete.
13) Enter the following path: /usr/local/adacnew/Patients/Institution_4/Mount_0/Patient_1488 (substituting the Institution, Mount, and Patient numbers recorded in step # 8)
14) Select the appropriate plan (plans will be labeled as “Plan_0”, “Plan_1” etc… so select the plan that corresponds to the “CBCT” plan that you’ve created for this patient. ie. The second, or third plan in the list)
15) In the left-side pane select “plan.roi”
16) The new plan will now populate with the “CBCT roi’s” and will give an error message for each. This is normal so click through these.
17) Now open the “Expansion/Contraction” window and expand each roi by “0”. To do this select each structure in turn, select “Create new ROI” and type the name of the structure you’re about to expand but include an “x” on the end of it. This will effectively “fool” Pinnacle into believing that these structures were created on this dataset.
18) Once all structures have been “expanded by 0” into new ROIs, delete the original ROIs and remove the “x” from the end of the names for the new ROIs.
19) Reset the colors for each structure.
20) Add a beam to the plan. This will also generate a POI, and a prescription. These are needed in order to export.
21) Export as usual.
--
You received this message because you are subscribed to the Google
Groups "pinnacle3-users" group.
To post to this group, send email to pinnacl...@googlegroups.com
To subscribe to this group, send email to
pinnacle3-us...@googlegroups.com
To unsubscribe from this group, send email to
pinnacle3-use...@googlegroups.com
For more options, visit this group at
http://groups.google.com/group/pinnacle3-users?hl=en
Have.ct.resent.with.correct medical record.number.so.it.overwrites data in.pinnacle. Then.it.will.go.to.Mosaiq
--
Thanks. How can I orverwrite the old CT image after I load to Pinnacle?
Xiliang
On Mon, Apr 9, 2012 at 8:38 AM, Jackie Rothenbuehler <jrot...@gmail.com> wrote:
Have.ct.resent.with.correct medical record.number.so.it.overwrites data in.pinnacle. Then.it.will.go.to.Mosaiq
On Apr 9, 2012 10:31 AM, "Xiliang Nie" <xilia...@gmail.com> wrote:
Dear List.
Our CT Tech accidentally put the wrong Medical Record # on the CT image. I changed it after I loaded it to Pinnacle. But I still can't export CT image to Mozaiq from Pinnacle. I had to change the MR# in Mozaiq so that I can export. My understanding is that I can export to Mozaiq after I changed the MR# in Pinnacle even the MR# from CT is not right. Anyone experienced same problem? Is that R&V related. I used Lantis before. Looks like I had no this problem if I export to Lantis. of course I didn't export the CT image to R&V since Siemens machine doesn't relay on R&V for the CT image. Thanks. I can export the plan to Mozaiq even with wrong MR#.
Xiliang Nie