Hi there,
 
The best process for fusion I have found in the case where the PET CT is to be fused with a planning CT is this:
 
-Fuse CT to CT automatically then tweak in region of interest if applicable (using like window levels and in some cases countoured anatomy (bony easiest but not always best));
 
-copy the same translation/rotation parameters to the PET from the PET/CT.  This always creates a wild result at first butby tweaking the y translation by 20cm it lines up perfectly (assuming rotational degrees are minor) ... usually.
 
-Can someone help me understand why sometimes this number is 15 or 25cm? and what it represents (different origin of CT and PET images?)
-it may be as simple as different scanners...
 
Thanks,
Kevin


--
The pessimist complains about the wind.
The optimist expects it to change.
The leader adjusts the sails.


~ John Maxwell
 

John Shakeshaft

unread,
Feb 29, 2012, 7:40:44 PM2/29/12
to pinnacl...@googlegroups.com

Hi Kevin,

 

This process needs quite careful QA for each scanner that you use (if you are going to use this registered image data for target volume creation). Remember that if the registration is wrong, this will potentially create a systematic error which propagated all through treatment. You need to know the size of your potential systematic errors when defining your margins. So depending on your on which formalism you use, a systematic registration error of 1mm should typically add 2-2.5mm to your margins. Sub-mm registration accuracy given deformations in the patient between image sets, and in the case of MR given distortions in the image can be quite a challenge.

 

Some questions:

Are the datasets (PET and CT) the same length with the same field of view? This should be checked for each dataset. They usually are but sometimes data can be lost along the way. This will cause errors in Pinnacle if you use the method you describe.

Are the coordinate systems the same in both image sets? This can easily be checked by finding an anatomical point which can be identified in both datasets, placing a point on it and copying it between the datasets and checking that both copies have the same coordinates.

Rotations: If you are seeing these sorts of problems you need to be VERY careful to make sure that rotations are expressed about the same point. Small errors in rotations can create large registration errors.

 

The robust way to copy registrations is to use the DICOM registration object, which has been supported in Pinnacle since version 8.0. However it is not supported in all systems.  It would take some pages to give a full tutorial on this here.

 

Best regards

 

John

 


--
You received this message because you are subscribed to the Google
Groups "pinnacle3-users" group.
To post to this group, send email to pinnacl...@googlegroups.com
To subscribe to this group, send email to
pinnacle3-us...@googlegroups.com
To unsubscribe from this group, send email to
pinnacle3-use...@googlegroups.com
For more options, visit this group at
http://groups.google.com/group/pinnacle3-users?hl=en

Pfund,John

unread,
Feb 29, 2012, 8:49:13 PM2/29/12
to pinnacl...@googlegroups.com

The y translation is related to the differences in field of view.  The PET is typically 70cm while it is typically 50cm for your CT.  The way P3 interprets these FOV’s results in the 20cm offset.  A good way to get around this is to have your PET/CT techs reconstruct the PET FOV to match the CT before they send them to P3.

 

John Pfund, MS

Sanford Health

Roger Maris Cancer Center

Fargo, ND

 

From: pinnacl...@googlegroups.com [mailto:pinnacl...@googlegroups.com] On Behalf Of Kevin Riddell
Sent: Wednesday, February 29, 2012 6:17 PM
To: pinnacl...@googlegroups.com; meddos
Subject: [p3rtp] PET CT fusion on Pinnacle

 

Hi there,

--

You received this message because you are subscribed to the Google
Groups "pinnacle3-users" group.
To post to this group, send email to pinnacl...@googlegroups.com
To subscribe to this group, send email to
pinnacle3-us...@googlegroups.com
To unsubscribe from this group, send email to
pinnacle3-use...@googlegroups.com
For more options, visit this group at
http://groups.google.com/group/pinnacle3-users?hl=en



-----------------------------------------------------------------------
Confidentiality Notice: This e-mail message, including any attachments,
is for the sole use of the intended recipient(s) and may contain
privileged and confidential information. Any unauthorized review, use,
disclosure or distribution is prohibited. If you are not the intended
recipient, please contact the sender by reply e-mail and destroy
all copies of the original message.

Bert Goudry

unread,
Mar 1, 2012, 3:17:03 AM3/1/12
to pinnacl...@googlegroups.com

Hello,

 

We have seen the same problem a few years ago.  And after some trial and error we came up with an answer. In fact a collegue (Steven Staelens) of the UZA (Universitair Ziekenhuis Antwerpen) came up with a solution. It seemed that it was a “small bug” in Pinnacle that causes these problems.

 

I will add the solution as an attachment.

 

The way we work is by fusing the CT-images from the PET-CT. When the fusion of the CT is done adjust the position of the PET-images by the attached protocol.

 

Another way of solving this problem is as mentioned before by John Pfund: Equal FOV reconstruction for CT and PET part.

 

 

Hope this will solve the problem,

 

 

Bert

 

____________________________________________________________________________

Bert Goudry

Dosimetrist

Universitaire Radiotherapie Antwerpen
ZNA Middelheim
Lindendreef 1
2020 Antwerpen
Belgium

 

+32 3 2802188

bert.goudry@zna.be

____________________________________________________________________________

 

PET_CT_shift_RT.pdf

Bryan Murray

unread,
Mar 1, 2012, 11:40:58 AM3/1/12
to pinnacl...@googlegroups.com

I will throw one more step in the process that helps me sleep at night…

 

After fusing the pet ct to the planning ct, add an roi that is attached to the pet ct.  Auto-contour the external surface of the pet ct in the region of the planning ct (the secondary dataset will need to be set to “transverse”.  Then, transfer that contour to the planning ct.  Switch the secondary dataset to the pet and window/level to where you can see the outer contour of the pet.  When copying the translation from the ct to the pet, the outer contour should “fit”.  I like to look at this in a coronal and sagittal view as well setting the “pet ct external” to poly.

 

Bryan Murray, CMD

Dallas, TX

 


Reply all
Reply to author
Forward
0 new messages