Hi all,
I
noticed something important recently about the alignment of the optics
with regards to the pgFocus performance. I thought I would share what I
learned on the forum.
In previous posts I noted that the pgFocus
performance can be hindered if cells on the coverslip scatter the laser
beam and create a messy beam profile. This essentially turns a nice,
Gaussian beam into a beam with a lot of fringes, which is difficult for
the pgFocus to lock onto. What I didn't realize at the time--though now
it seems obvious--is that the strength of the scattering by the sample
varies with the penetration depth of the evanescent field into the
sample space. This penetration depth is essentially controlled by the location of
the finite-sized, focused beam is in back exit pupil of the
objective. I attached an illustration below to highlight this.
To
get optimal alignment for pgFocus operation, I start with the IR beam
in widefield, i.e. it goes straight up from the objective (be careful
about looking into the objective when you do this). I then turn the
screw on the mirror that translates the beam across the exit pupil of
the objective. This translates the focused beam across the exit pupil,
as seen in the left-most part of the figure. Once the beam enters the
annular region where TIR occurs, and you will see a faint spot on the IR
viewer if everything is well aligned. Translating slightly further, the
beam goes fully into TIR, producing a bright spot on the IR viewer.
This is not optimal, however, because the evanescent light is not
strongly confined to the coverslip.
The ideal location is to have
a very small amount of light from the focused beam passing near the
edge of the aperture, with the rest of the light being cutoff from the
finited-sized aperture. This ensures that most of the light is hitting
the coverslip at the maximum angles allowed by the objective, which
means the evanescent field penetration is small. This reduces scattering
of the beam and should give a nice, clean profile. (Note that I also
use a zero-aperture iris immediately in front of the laser to clean up
the beam before it enters the microscope.)
By consciously placing
most of the beam outside the aperture, I have managed to get ~2 to 3 nm
std. deviation locking with an NA 1.4 objective and the sample in
water.
I hope this helps everyone with their alignments!
Kyle