I am currently trying to implement a pgFocus-based focus stabilization system on a TIRF setup with a 60x 1.49 oil immersion objective.
I use a Thorlabs CPS808S collimated laser diode with an iris to clean up the beam profile. A 400mm focal length lens should focus the beam onto the BFP and a neutral density filter attenuates the beam.
It travels through a beam splitter and is reflected using a shortpass dichro onto the TIRF dichro cube inside the microscope body, parallel to the excitation laser beam which is focused onto the BFP using a 200mm lens.
On the way back, it is reflected again by the shortpass dichro, and the portion which is reflected inside of the beamsplitter falls onto a linear CCD array.
I was wondering what kind of optics other people use to get a nice beam shape, and to focus the infrared laser onto the BFP and where they had put that lens. Placing it upstream of the translatable mirror forces me to use a very long focal length lens and to focus through the beamsplitter. Also, I think that closing the iris too far to improve the beam shape messes with the collimation of the beam.
Any input is welcome,
Chris

--
You received this message because you are subscribed to the Google Groups "pgFocus" group.
To unsubscribe from this group and stop receiving emails from it, send an email to pgfocus+u...@googlegroups.com.
To post to this group, send email to pgf...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/pgfocus/2ab70e16-cb10-4bc4-8ed6-dab537764ae4%40googlegroups.com.
For more options, visit https://groups.google.com/d/optout.
Hi,
thanks for your input. I think I could improve the centering of the beam on the iris. The board is a ST microelectronics board (stm discovery kit) we had around.
About the tube lens: I am not sure if I get you correctly. I do have a tube lens to focus my fluorescence emission onto the chip of a sCMOS camera. The pgFocus pathway however is coupled into the excitation pathway and since it should be collimated in the sample plane, it has to be focused onto the back focal plane of the objective (just like the TIR beam used for fluorescence excitation) - which I try to do with the 400mm lens in the pgFocus pathway.
I know that there are focus locking solutions using a dichroic mirror in the emission pathway, but I didn't want to risk loosing any fluorescence photons in the emission pathway.
Chris
--
You received this message because you are subscribed to the Google Groups "pgFocus" group.
To unsubscribe from this group and stop receiving emails from it, send an email to pgfocus+u...@googlegroups.com.
To post to this group, send email to pgf...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/pgfocus/9ee490f7-4aa1-4442-b277-8d2b7c62a9e8%40googlegroups.com.
For more options, visit https://groups.google.com/d/optout.
To unsubscribe from this group and stop receiving emails from it, send an email to pgf...@googlegroups.com.
To post to this group, send email to pgf...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/pgfocus/9ee490f7-4aa1-4442-b277-8d2b7c62a9e8%40googlegroups.com.
For more options, visit https://groups.google.com/d/optout.