- To avoid occlusion of the left subclavian artery and visceral and renal
arteries, its tip should be slightly cephalad to the adjacent carina (2
nd -3 rd intercostal space). The balloon should
not occlude more than 85-90% of the aortic diameter. Balloon rupture
with air embolization and septicemia are rare potential complications. - Case:After
uneventful insertion of IABP (intra-aortic balloon pump) - you have
been informed that, it will take a while before CXR can confirm proper
location of tip of IABP. What is one easy method to determine that you
have not 'over-shooted' the tip of IABP?
Answer: Check the left radial pulse
Ideally,
the tip of the balloon should be positioned 2–3 cm distal to the origin
of the left subclavian artery (LSCA). This position results in maximum
augmentation of coronary artery flow although minimizing the risk of
embolization to the cerebral vessels and occlusion of the LSCA. If you
have good left radial artery pulsation, probably your tip of IABP is
distal to LSCA.
Carina as a Radiographic Landmark for Positioning the IABP There are reports in literature that targeting tip of
Intraaortic Balloon Pump (IABP) just below the aortic knob on CXR
radiologically may still cause occlusion of left subclavian artery (upto
7 - 16%). 1,2.
See this interesting another approach to target carina as a
Useful Radiographic Landmark for Positioning the Intraaortic Balloon
Pump.
METHODS: The distance from the
top of the distal aortic arch (aortic knob) to the left subclavian
artery (LSCA) on three-dimensional computed tomography angiography in
100 patients, was measured. The distance from the level of the LSCA
origin to the level of the carina was also measured using
three-dimensional computed tomography in 150 additional patients.
RESULTS: In 16% of the aortic
knob study population, the LSCA to aortic knob distance was <0 cm or 0
cm. The median distance from the LSCA to the carina was 42 mm (range:
30–63 mm). In the carina study population, the origin of the LSCA was
35–55 mm above the carina in 95.3% of patients.
CONCLUSION:
In 16% of patients, the IABP was too close to the LSCA origin when
it was placed at the aortic knob, whereas Positioning the IABP at 2 cm above the carina provided an adequate position for the IABP tip (1.5–3.5 cm distal to the origin of the LSCA) in 95.3% of patients.
The carina may be a more reliable landmark for positioning the IABP than the aortic knob.