OrthancException.h:86] Bad type for a parameter

33 views
Skip to first unread message

Esa

unread,
Jun 3, 2020, 3:58:10 PM6/3/20
to Orthanc Users
Hello,

I am querying Orthanc archive using dicom interface and I get following error message and the query fails:

I0604 05:40:27.687058 OrthancFindRequestHandler.cpp:603]   (0008,0052)  QueryRetrieveLevel = IMAGE
I0604 05:40:27.687079 OrthancFindRequestHandler.cpp:603]   (0010,0020)  PatientID = test2195
I0604 05:40:27.687099 OrthancFindRequestHandler.cpp:612]   (0008,114a)  ReferencedInstanceSequence : sequence tag whose content will be copied
I0604 05:40:27.687114 OrthancFindRequestHandler.cpp:612]   (0040,0275)  RequestAttributesSequence : sequence tag whose content will be copied
I0604 05:40:27.687131 OrthancFindRequestHandler.cpp:612]   (0407,10a1)  Unknown Tag & Data : sequence tag whose content will be copied
I0604 05:40:27.687501 ServerContext.cpp:938] Number of candidate resources after fast DB filtering on main DICOM tags: 1
I0604 05:40:27.687692 FilesystemStorage.cpp:155] Reading attachment "a232244d-7020-4d78-b7e0-d5ad930b25ff" of "JSON summary of DICOM" content type
E0604 05:40:27.699394 OrthancException.h:86] Bad type for a parameter: Bad Parameter type for tag 0407,10a1
E0604 05:40:27.700718 FindScp.cpp:298] C-FIND request handler has failed: Bad type for a parameter

Problem is probably caused because of the private tag 0407,10a1.
Do I need to add private dictionary to support this query or what might be the problem?

I am using the latest version 1.7.1.

Thanks,

Alain Mazy

unread,
Jun 6, 2020, 9:18:24 AM6/6/20
to Esa, Orthanc Users
Hi,

This issue has been investigated and fixed in the scope of a dedicated support contract but I'm also informing the community:

- The fix has been implemented in this commit: https://hg.orthanc-server.com/orthanc/rev/884b55ce01f6
- and the private sequence tags had to be declared in the configuration file:
  "Dictionary" : {
    "0407,10a1" : ["SQ", "OCT Cube Sequence", 1, 1, "99CZM_CapeCod_OctGeneral" ],
    "0407,1005" : ["SQ", "OCT Cube Sequence2", 1, 1, "99CZM_CapeCod_OctGeneral" ],
    "0407,0010" : ["LO", "OCT Cube Sequence3", 1, 1, "99CZM_CapeCod_OctGeneral" ]

Br,

Alain

--
You received this message because you are subscribed to the Google Groups "Orthanc Users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to orthanc-user...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/orthanc-users/4725b79c-0b39-443d-b76c-3fa99b5f2bbe%40googlegroups.com.


--

Alain Mazy / Orthanc-Studio Manager & Software Developer
a...@osimis.io / +32 494 31 67 27

Osimis

OSIMIS S.A. 
Quai Banning 6BE-4000 Liège 
www.osimis.io

Twitter LinkedIn