Unable to load image cone beam CT from Carestream

1,766 views
Skip to first unread message

Roberto Fichera

unread,
May 14, 2018, 5:42:16 AM5/14/18
to Orthanc Users
Hi Guys,

I've just got this problem when trying to upload Cone Beam images from a Carestream device, below the log.

Please let me know what I can do to get Orthanc v1.3.2 to accept such images.

Thanks in advance,
Roberto Fichera.

May 14 09:10:56 localhost Orthanc[649]: W: Found element (0009,11a0) with VR UN and undefined length, reading a sequence with transfer syntax LittleEndianImplicit (CP-246)
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Length of element (0009,11a1) is odd
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Dataset not in ascending tag order, at element (0000,0000)
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Element (0000,0000) found twice in one data set or item, ignoring second entry
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Element (0000,0000) found twice in one data set or item, ignoring second entry
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Element (0000,0000) found twice in one data set or item, ignoring second entry
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Element (0000,0000) found twice in one data set or item, ignoring second entry
May 14 09:10:56 localhost Orthanc[649]: W: DcmItem: Length of element (028f,028b) is odd
May 14 09:11:04 localhost Orthanc[649]: E: DcmElement: Unknown Tag & Data (028f,028b) larger (39846519) than remaining bytes (563228) in file, premature end of stream
May 14 09:11:04 localhost Orthanc[649]: W: DIMSE Warning: (KPServer,ORTHANC): DIMSE receiveDataSetInMemory: dset->read() Failed (Invalid stream)
May 14 09:11:04 localhost Orthanc[649]: E0514 09:11:04.644651 StoreScp.cpp:294] Store SCP Failed: DIMSE Failed to receive message

and below you can see the tags details show by python pydicom library

In [1]: from pydicom import dicomio
In [2]: i = dicomio.read_file('export.dcm')
In [3]: i
Out[3]: 
(0008, 0005) Specific Character Set              CS: 'ISO_IR 192'
(0008, 0008) Image Type                          CS: ['DERIVED', 'PRIMARY', 'VOLUME', 'RESAMPLED']
(0008, 0012) Instance Creation Date              DA: '20180514'
(0008, 0013) Instance Creation Time              TM: '111411'
(0008, 0016) SOP Class UID                       UI: X-Ray 3D Craniofacial Image Storage
(0008, 0018) SOP Instance UID                    UI: 1.2.250.1.90.2.772081700.20180514111411.5796.12204
(0008, 0020) Study Date                          DA: '20180406'
(0008, 0021) Series Date                         DA: '20180514'
(0008, 0022) Acquisition Date                    DA: '20180406'
(0008, 0023) Content Date                        DA: '20180514'
(0008, 0030) Study Time                          TM: '180549'
(0008, 0031) Series Time                         TM: '111411'
(0008, 0032) Acquisition Time                    TM: '180741'
(0008, 0033) Content Time                        TM: '111411'
(0008, 0050) Accession Number                    SH: ''
(0008, 0060) Modality                            CS: 'DX'
(0008, 0070) Manufacturer                        LO: 'Carestream Health'
(0008, 0080) Institution Name                    LO: 'XXXXXXX'
(0008, 0090) Referring Physician's Name          PN: 'XXXXXXX'
(0008, 1030) Study Description                   LO: ''
(0008, 103e) Series Description                  LO: 'Exported Volume'
(0008, 1040) Institutional Department Name       LO: 'RADIOLOGY'
(0008, 1090) Manufacturer's Model Name           LO: 'CS 9000'
(0008, 9206) Volumetric Properties               CS: 'VOLUME'
(0008, 9207) Volume Based Calculation Technique  CS: 'MPR'
(0009, 0011) Private Creator                     LO: 'TROPHY'
(0009, 11a0)  Private tag data   4 item(s) ---- 
   (0009, 11a1) Private tag data                    LO: '3DACQ_FUNCTION_MODE'
   (0009, 11a2) Private tag data                    UL: 6
   (0009, 11a3) Private tag data                    UN: 'Normal'
   ---------
   (0009, 11a1) Private tag data                    LO: '3DACQ_NB_VOLUME_TO_STITCH'
   (0009, 11a2) Private tag data                    UL: 1
   (0009, 11a3) Private tag data                    UN: '1\x00'
   ---------
   (0009, 11a1) Private tag data                    LO: '3DACQ_SCAN_MODE'
   (0009, 11a2) Private tag data                    UL: 8
   (0009, 11a3) Private tag data                    UN: 'FullScan'
   ---------
   (0009, 11a1) Private tag data                    LO: '3DACQ_SCAN_OBJECT'
   (0009, 11a2) Private tag data                    UL: 7
   (0009, 11a3) Private tag data                    UN: 'Patient\x00'
   ---------
(0010, 0010) Patient's Name                      PN: 'XXXXX^YYYYY'
(0010, 0020) Patient ID                          LO: '7884839267658269838482696577926866926546828671926548484848494856'
(0010, 0030) Patient's Birth Date                DA: 'XXXXMMDD'
(0010, 0040) Patient's Sex                       CS: 'O'
(0018, 0015) Body Part Examined                  CS: 'JAW'
(0018, 1000) Device Serial Number                LO: 'BEBH009'
(0018, 1020) Software Version(s)                 LO: ['AcqPanoramic.dll##5.4.4.32##01/07/14', 'Pano.dll##5.4.4.32##01/07/14', 'Odt2000.dll##5.4.4.32##01/07/14', 'TPjbus.dll##5.4.4.32##01/07/14', 'Trajectory.dll##5.4.4.32##01/07/14', 'TPita.dll##5.4.4.32##01/07/14', 'TaskVolumizer.dll##3.1.12.0##22/05/2014##W32', 'CS Volume Converter##1.0.9.0##29/02/12']
(0018, 9507)  X-Ray 3D Acquisition Sequence   1 item(s) ---- 
   (0018, 0060) KVP                                 DS: "70"
   (0018, 115e) Image and Fluoroscopy Area Dose Pro DS: "2.6127864"
   (0018, 7004) Detector Type                       CS: 'SCINTILLATOR'
   (0018, 700a) Detector ID                         SH: 'BDTD044/VBAJ11'
   (0018, 702a) Detector Manufacturer Name          LO: 'Carestream Health'
   (0018, 702b) Detector Manufacturer's Model Name  LO: '3D_V1'
   (0018, 9328) Exposure Time in ms                 FD: 10800.0
   (0018, 9330) X-Ray Tube Current in mA            FD: 12.0
   (0018, 9516) Start Acquisition DateTime          DT: '20180406180741'
   (0020, 0012) Acquisition Number                  IS: '1'
   ---------
(0018, 9530)  X-Ray 3D Reconstruction Sequence   1 item(s) ---- 
   (0018, 9524) Application Name                    LO: 'CS Volume Converter'
   (0018, 9525) Application Version                 LO: '1.0.9.0'
   (0018, 9526) Application Manufacturer            LO: 'Carestream'
   ---------
(0020, 000d) Study Instance UID                  UI: 1.2.250.1.90.4.772081700.20180406180549.2564.16
(0020, 000e) Series Instance UID                 UI: 1.2.250.1.90.3.772081700.20180514111411.5796.12202
(0020, 0010) Study ID                            SH: '1.2.250.1.90.1.772081700.1523030489.683'
(0020, 0011) Series Number                       IS: '1'
(0020, 0013) Instance Number                     IS: '1'
(0020, 0020) Patient Orientation                 CS: ['L', 'P']
(0020, 0052) Frame of Reference UID              UI: 1.2.250.1.90.5.772081700.20180406180813.6140.33
(0020, 1040) Position Reference Indicator        LO: ''
(0020, 4000) Image Comments                      LT: '3D acquisition mode: Simple volume'
(0028, 0002) Samples per Pixel                   US: 1
(0028, 0004) Photometric Interpretation          CS: 'MONOCHROME2'
(0028, 0008) Number of Frames                    IS: '357'
(0028, 0010) Rows                                US: 541
(0028, 0011) Columns                             US: 541
(0028, 0100) Bits Allocated                      US: 16
(0028, 0101) Bits Stored                         US: 16
(0028, 0102) High Bit                            US: 15
(0028, 0103) Pixel Representation                US: 0
(0028, 2110) Lossy Image Compression             CS: '00'
(0028, 9520) Image to Equipment Mapping Matrix   DS: ['1.00000000E+000', '0.00000000E+000', '0.00000000E+000', '1.70000000E+001', '0.00000000E+000', '1.00000000E+000', '0.00000000E+000', '4.00000000E+000', '0.00000000E+000', '0.00000000E+000', '1.00000000E+000', '0.00000000E+000', '0.00000000E+000', '0.00000000E+000', '0.00000000E+000', '1.00000000E+000']
(0028, 9537) Equipment Coordinate System Identif CS: 'ISOCENTER'
(0040, 0555)  Acquisition Context Sequence   0 item(s) ---- 
(2050, 0020) Presentation LUT Shape              CS: 'IDENTITY'
(5200, 9229)  Shared Functional Groups Sequence   1 item(s) ---- 
   (0018, 9504)  X-Ray 3D Frame Type Sequence   1 item(s) ---- 
      (0008, 9007) Frame Type                          CS: ['DERIVED', 'PRIMARY', 'VOLUME', 'RESAMPLED']
      (0008, 9206) Volumetric Properties               CS: 'VOLUME'
      (0008, 9207) Volume Based Calculation Technique  CS: 'MPR'
      (0020, 9536) Reconstruction Index                US: 1
      ---------
   (0020, 9071)  Frame Anatomy Sequence   1 item(s) ---- 
      (0008, 2218)  Anatomic Region Sequence   1 item(s) ---- 
         (0008, 0100) Code Value                          SH: 'T-D1213'
         (0008, 0102) Coding Scheme Designator            SH: 'SRT'
         (0008, 0104) Code Meaning                        LO: 'Jaw region'
         ---------
      (0020, 9072) Frame Laterality                    CS: 'U'
      ---------
   (0020, 9113)  Plane Position Sequence   0 item(s) ---- 
   (0020, 9116)  Plane Orientation Sequence   1 item(s) ---- 
      (0020, 0037) Image Orientation (Patient)         DS: ['1.00000000E+000', '0.00000000E+000', '0.00000000E+000', '0.00000000E+000', '1.00000000E+000', '0.00000000E+000']
      ---------
   (0028, 9110)  Pixel Measures Sequence   1 item(s) ---- 
      (0018, 0050) Slice Thickness                     DS: "0.1"
      (0028, 0030) Pixel Spacing                       DS: ['1.00000000E-001', '1.00000000E-001']
      ---------
   (0028, 9132)  Frame VOI LUT Sequence   1 item(s) ---- 
      (0028, 1050) Window Center                       DS: "1.04800000E+003"
      (0028, 1051) Window Width                        DS: "4.09600000E+003"
      ---------
   (0028, 9145)  Pixel Value Transformation Sequence   1 item(s) ---- 
      (0028, 1052) Rescale Intercept                   DS: "-1000"
      (0028, 1053) Rescale Slope                       DS: "1"
      (0028, 1054) Rescale Type                        LO: 'HU'
      ---------
   ---------
(5200, 9230)  Per-frame Functional Groups Sequence   357 item(s) ---- 
   (0020, 9111)  Frame Content Sequence   0 item(s) ---- 
   (0020, 9113)  Plane Position Sequence   1 item(s) ---- 
      (0020, 0032) Image Position (Patient)            DS: ['-2.70000000E+001', '-2.70000000E+001', '-1.78000000E+001']
      ---------
   (0020, 9116)  Plane Orientation Sequence   0 item(s) ---- 
   ---------
   (0020, 9111)  Frame Content Sequence   0 item(s) ---- 
   (0020, 9113)  Plane Position Sequence   1 item(s) ---- 
      (0020, 0032) Image Position (Patient)            DS: ['-2.70000000E+001', '-2.70000000E+001', '-1.77000000E+001']
      ---------
   (0020, 9116)  Plane Orientation Sequence   0 item(s) ---- 
   ---------
[..... other 356 blocks ....]
   ---------
(7fe0, 0010) Pixel Data                          OB: Array of 90218756 bytes

Sébastien Jodogne

unread,
May 14, 2018, 6:27:29 AM5/14/18
to Orthanc Users
Hello,

According to the Orthanc log, your DICOM file looks corrupted.

You can use the reference tool "dciodvfy" from David Clunie's dicom3tools in order to check the validity of the file wrt. the DICOM standard:

Please provide a sample (possibly anonymized) DICOM image if you want further support from the Orthanc community.

HTH,
Sébastien-

Roberto Fichera

unread,
May 14, 2018, 6:48:10 AM5/14/18
to Orthanc Users
Hi Sébastien,

I think the problem is related to not standard tags present in the image

dciodvfy /TempData/export.dcm

(0x0009,0x11a0)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a1)  ?  - Warning - Private tag without owner - use owner of parent sequence "TROPHY"
(0x0009,0x11a1)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a2)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a3)  ?  - Warning - Unrecognized tag - explicit value representation is UN
(0x0009,0x11a1)  ?  - Warning - Private tag without owner - use owner of parent sequence "TROPHY"
(0x0009,0x11a1)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a2)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a3)  ?  - Warning - Unrecognized tag - explicit value representation is UN
(0x0009,0x11a1)  ?  - Warning - Private tag without owner - use owner of parent sequence "TROPHY"
(0x0009,0x11a1)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a2)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a3)  ?  - Warning - Unrecognized tag - explicit value representation is UN
(0x0009,0x11a1)  ?  - Warning - Private tag without owner - use owner of parent sequence "TROPHY"
(0x0009,0x11a1)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a2)  ?  - Warning - Unrecognized tag - assuming explicit value representation OK
(0x0009,0x11a3)  ?  - Warning - Unrecognized tag - explicit value representation is UN

Error - Functional Group Sequence already used in Shared Functional Groups Sequence - (0x0020,0x9113) Plane Position Sequence - in Per-frame Functional Groups Sequence Item #1
Error - Functional Group Sequence already used in Shared Functional Groups Sequence - (0x0020,0x9116) Plane Orientation Sequence - in Per-frame Functional Groups Sequence Item #1
Error - Functional Group Sequence already used in Shared Functional Groups Sequence - (0x0020,0x9113) Plane Position Sequence - in Per-frame Functional Groups Sequence Item #2
...
Error - Functional Group Sequence already used in Shared Functional Groups Sequence - (0x0020,0x9116) Plane Orientation Sequence - in Per-frame Functional Groups Sequence Item #356
Error - Functional Group Sequence already used in Shared Functional Groups Sequence - (0x0020,0x9113) Plane Position Sequence - in Per-frame Functional Groups Sequence Item #357
Error - Functional Group Sequence already used in Shared Functional Groups Sequence - (0x0020,0x9116) Plane Orientation Sequence - in Per-frame Functional Groups Sequence Item #357
Warning - Value dubious for this VR - (0x0008,0x0090) PN Referring Physician's Name  PN [0] = <XXXXXX> - Retired Person Name form
Error - Value invalid for this VR - (0x0020,0x0010) SH Study ID  SH [0] = <1.2.250.1.90.1.772081700.1523030489.683> - Length invalid for this VR = 39, expected <= 16
Error - Dicom dataset contains invalid data values for Value Representations
XRay3DCraniofacialImage
Error - Unrecognized enumerated value <RESAMPLED> for value 4 of attribute <Frame Type>
Error - Missing attribute Type 1 Required Element=<PixelPresentation> Module=<CommonCTMRImageDescriptionFrameLevelMacroForXRay3DImage>
....
and other stuff.

I will anonymize the image and will post the link to it.

Thanks.
Roberto Fichera.

Roberto Fichera

unread,
May 14, 2018, 7:48:30 AM5/14/18
to Orthanc Users
Hi Sébastien,


Orthanc is currently configured with LoadPrivateDictionary = false because I wasn't able to import certain images due to Endian and also some private tags.
I've also tried to register via Dictionary the above private (0009,11a0) tags, but at moment I wasn't able to make it correctly.

Sébastien Jodogne

unread,
May 14, 2018, 9:03:28 AM5/14/18
to Orthanc Users
Hello,

Your DICOM file has "SOPClassUID" set to "1.2.840.10008.5.1.4.1.1.13.1.2", which is a proprietary, non-standard DICOM class by Carestream:

In order to make Orthanc accept such files, you'll have to set its "UnknownSopClassAccepted" configuration option to "true":

HTH,
Sébastien-


PS: For reference, here are the instructions to allow DCMTK's storescu to send such files to Orthanc on Ubuntu 16.04:

# cd /tmp/
# sed 's/^\(PresentationContext125\).*$/\1 = 1.2.840.10008.5.1.4.1.1.13.1.2\\JPEGLossless/' /etc/dcmtk/storescu.cfg > storescu.cfg
# storescu -xf storescu.cfg Default localhost 4242 ./DX000000.dcm

Check out the following page for more information:

Roberto Fichera

unread,
May 25, 2018, 1:27:21 PM5/25/18
to Orthanc Users
Hi Sèbastien,

I've tried to set UnknownSopClassAccepted = true, but I still get the same problem

May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.398946 main.cpp:1298] Orthanc version: 1.3.2
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.399935 OrthancInitialization.cpp:168] Scanning folder "/etc/orthanc/" for configuration files
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.399976 OrthancInitialization.cpp:120] Reading the configuration from: "/etc/orthanc/orthanc.json"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.400169 OrthancInitialization.cpp:120] Reading the configuration from: "/etc/orthanc/worklists.json"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.400215 OrthancInitialization.cpp:120] Reading the configuration from: "/etc/orthanc/serve-folders.json"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.407317 FromDcmtkBridge.cpp:164] Loading the external DICOM dictionary "/usr/share/dcmtk/dicom.dic"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.413632 FromDcmtkBridge.cpp:164] Loading the external DICOM dictionary "/usr/share/dcmtk/private.dic"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.423836 main.cpp:671] Loading plugin(s) from: /usr/share/orthanc/plugins/
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424013 PluginsManager.cpp:269] Registering plugin 'worklists' (version 1.3.2)
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424025 PluginsManager.cpp:168] Sample worklist plugin is initializing
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424095 PluginsManager.cpp:168] Worklist server is disabled by the configuration file
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424211 PluginsManager.cpp:269] Registering plugin 'serve-folders' (version 1.3.2)
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424350 PluginsManager.cpp:168] ServeFolders: Empty configuration file: No additional folder will be served!
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424363 OrthancInitialization.cpp:1004] SQLite index directory: "/var/lib/orthanc/db-v6"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424460 OrthancInitialization.cpp:1074] Storage directory: "/var/lib/orthanc/db-v6"
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.424704 HttpClient.cpp:691] HTTPS will use the CA certificates from this file: /etc/orthanc/
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.425108 ServerScheduler.cpp:135] The server scheduler has started
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.425244 LuaContext.cpp:103] Lua says: Lua toolbox installed
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.425275 ServerContext.cpp:182] Disk compression is disabled
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.425282 ServerIndex.cpp:1403] No limit on the number of stored patients
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.425291 ServerIndex.cpp:1420] No limit on the size of the storage area
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.425297 main.cpp:651] Installing the Lua scripts from: /etc/orthanc/send_image.lua
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.426632 main.cpp:863] DICOM server listening with AET ORTHANC on port: 8000
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.426669 MongooseServer.cpp:1075] HTTP compression is enabled
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.427107 main.cpp:796] HTTP server listening on port: 8001
May 25 19:20:49 puma Orthanc[218672]: W0525 19:20:49.427122 main.cpp:683] Orthanc has started
May 25 19:21:04 puma Orthanc[218672]: W: Found element (0009,11a0) with VR UN and undefined length, reading a sequence with transfer syntax LittleEndianImplicit (CP-246)
May 25 19:21:04 puma Orthanc[218672]: W: DcmItem: Length of element (0009,11a1) is odd
May 25 19:21:06 puma Orthanc[218672]: E: DcmElement: PrivateCreator (0dff,00e0) larger (4261412864) than remaining bytes (209010253) in file, premature end of stream
May 25 19:21:06 puma Orthanc[218672]: W: DIMSE Warning: (KPServer,ORTHANC): DIMSE receiveDataSetInMemory: dset->read() Failed (Invalid stream)
May 25 19:21:06 puma Orthanc[218672]: E0525 19:21:06.282193 StoreScp.cpp:294] Store SCP Failed: DIMSE Failed to receive message

Sébastien Jodogne

unread,
May 26, 2018, 4:32:12 AM5/26/18
to Orthanc Users
Hello,

I have checked again, and your file works properly.

Here are the full steps to reproduce using Docker. Open a command-line terminal 1:

# cd /tmp/
# sudo docker run -p 4242:4242 -p 8042:8042 --rm jodogne/orthanc:1.3.2

Open a command-line terminal 2:

# cd /tmp/
# sed 's/^\(PresentationContext125\).*$/\1 = 1.2.840.10008.5.1.4.1.1.13.1.2\\JPEGLossless/' /etc/dcmtk/storescu.cfg > storescu.cfg
# storescu -xf storescu.cfg Default localhost 4242 ./DX000000.dcm
I: checking input files ...
I: Requesting Association
I: Association Accepted (Max Send PDV: 16372)
I: Sending file: ./DX000000.dcm
[...]
I: Received Store Response (Success)
I: Releasing Association

The issue is thus to be searched in your DICOM SCU client.

Sébastien-
Reply all
Reply to author
Forward
0 new messages