Comment fields on NM/PET page

21 views
Skip to first unread message

Dave Ashworth

unread,
Oct 11, 2025, 7:09:21 AMOct 11
to OpenREM
Hi, 

Which DICOM tags do the Comment fields in the details list and the radiopharmaceutical table on the NM/PET exam page use?

I've been trying to populate this with the bed speed / time per bed information extracted from ActualFrameDuration but they're always showing as blank or None. Have tried all the comments tags I could find...

(0010,4000) Patient Comments
(0020,4000) Image Comments
(0020,9158) Frame Comments
(0024,0044) Comments on Patient's Performance of Visual Field
(0028,7007) System Status Comment
(0028,702A) Test Result Comment
(0032,4000) Study Comments
(0038,4000) Visit Comments
(0040,0280) Comments on the Performed Procedure Step
(0040,0310) Comments on Radiation Dose
(0040,0400) Comments on the Scheduled Procedure Step
(0040,1400) Requested Procedure Comments
(0040,2400) Imaging Service Request Comments
(0066,0004) Surface Comments
(0100,0424) SOP Authorization Comment
(3004,0006) Dose Comment
(3006,00C8) Frame of Reference Transformation Comment
(3008,0202) Treatment Status Comment
(300A,0402) Setup Image Comment

Thanks

Ed McDonagh

unread,
Oct 11, 2025, 4:28:41 PMOct 11
to OpenREM
Hi Dave

I think the comment field is from the Radiopharmaceutical Radiation Dose Structured Report (RRDSR): https://bitbucket.org/openrem/openrem/src/develop/openrem/remapp/extractors/rrdsr_methods.py

I don't know if this is something you are able to add information to?

Kind regards

Ed 


--
You received this message because you are subscribed to the Google Groups "OpenREM" group.
To unsubscribe from this group and stop receiving emails from it, send an email to openrem+u...@googlegroups.com.
To view this discussion, visit https://groups.google.com/d/msgid/openrem/8f2e5152-16aa-4610-9c1c-d02485af72d5n%40googlegroups.com.

Dave Ashworth

unread,
Oct 13, 2025, 2:43:41 AMOct 13
to OpenREM
Hi Ed, 

I'm having to use dose info from NM/PT image files rather than RRDSRs as I'm pulling all my data in from Sectra PACS rather than sending direct from the scanners and haven't been able to get the RRDSRs onto Sectra so that'll be why none of that worked. And also one of our gamma cameras doesn't produce RRDSRs. 

Think I'll write the bed speed information to the Operator's Name field instead as we currently anonymise that anyway. Not as clean as having it in the comment field but just want it to be something we can see on the web interface and is in the export data so that'll work fine. 

Thanks for your help.

Dave

Ed McDonagh

unread,
Oct 13, 2025, 8:58:27 AMOct 13
to OpenREM
Looking at the code again, the nm_image routines don't attempt to populate the comment field, only the rrdsr_methods routines do.

Therefore to capture anything from an NM comment field you'd need to modify the code. 

The comment field in the initial block actually appears to come from the associated CT side, but I haven't followed it through to make sure. The comment field in the table is from the RRDSR.