Device synchronization

545 views
Skip to first unread message

Anan moran

unread,
Jan 26, 2021, 3:54:54 PM1/26/21
to Open Ephys
Hi,
I am using the GUI with a Neuropixels electrode. The computer is connected to a NI chassis in which the imec card and a NI-6341 daq reside. My current problem is with synchronizing the digital signals with the recorded channels of the NP. Following the guidelines I have merged the neuropixels-pxi and DAQmx sources into a Recording-node, set the  neuropixels-pxi as the Master, and the DAQ receives the sync signal on the 1st digital channel (I also disconnected this line, but with similar results). When I start acquiring, the orange change to green on the imec board, but the daq remains orange. I can see the 1Hz of the DIO in the LFP viewer of the neuropixels-pxi, however in the LFP viewer of the daq I see a constant ON. I confirmed that the SMA gives constant 4.9V out. I also confirmed that there is nothing wrong with the card by using the same hardware and connection with SpikeGLX, which produced alternating 1Hz signal and passed sync test. 
I use the latest version of the GUI and plugins.

Any ideas?

Thanks for your help
Anan

Pavel Kulik

unread,
Jan 26, 2021, 4:36:24 PM1/26/21
to Open Ephys
Hi Anan,

I just tested my NI-6341 card and syncing with Neuropixels is working properly in OpenEphys using a 1Hz sync.

Since your LFP Viewer shows constant ON for the NIDAQ module, that means the sync signal is not detected on the NIDAQ's digital lines. The output of the SMA should be physically connected to one of the P0 digital inputs on the terminal block (not the PFI ones). 

Also, if you have analog inputs available on your terminal block, you can try connecting the output to the SMA sync to any of the analog inputs and see if the sync shows up in the LFP Viewer as a continuous channel. 

Best,

Pavel

Anan moran

unread,
Jan 26, 2021, 5:01:34 PM1/26/21
to Open Ephys
Hi Pavel,
Thanks for the prompt response. 
I believe it is not the matter of the NIDAQ not detecting the signal, but the signal is not created properly by the imec card. I have tested the SMA output with a voltmeter, and it is a constant 4.9V. By the way, the SMA is indeed physically connected to P0. The imec card produce the sync signal internally to the imec card, as I can see it in the LFP viewer of the electrodes, but no such signal comes out of the SMA. Oddly, SpikeGLX seems to make the SMA output that 1Hz signal.   

Anan

Pavel Kulik

unread,
Jan 26, 2021, 5:22:16 PM1/26/21
to Open Ephys
Ah I see, so the Neuropixels-PXI module sync settings are set to 'OUTPUT' and '1 Hz', and the sync signal is present on NP LFPViewer but the SMA output stays high (4.9V)?  Are you using the latest version of the Neuropixels-PXI plugin (0.3.0) available via the PluginInstaller?

Pavel

Anan moran

unread,
Jan 26, 2021, 5:36:33 PM1/26/21
to Open Ephys
Exactly as you described, and yes, my Neuropixels-PXI plugin is of the latest version. 
Anan

Anan moran

unread,
Jan 27, 2021, 3:41:32 AM1/27/21
to Open Ephys
Maybe the debug log will help. Sorry for the long dump.

Debug console...

Created processor graph.

Got audio device.
Audio device name: Speakers (Realtek(R) Audio)
Audio device type: Windows Audio
Audio output channels: 2
Audio device sample rate: 44100
Audio device buffer size: 1024
Created audio component.
Created message center.
Created info label.
Created graph viewer.
Adding tab with index 1
Adding tab with index 2
Created data viewport.
Created editor viewport.
Created control panel.
Created filter list.
Created plugin manager
Loading Plugin: ArduinoOutput...
Loading Plugin: BasicSpikeDisplay...
Loading Plugin: CAR...
Loading Plugin: ChannelMappingNode...
Loading Plugin: EvntTrigAvg...
Loading Plugin: FilterNode...
Loading Plugin: IntanRecordingController...
Loading Plugin: LfpDisplayNode...
Loading Plugin: LfpDisplayNodeBeta...
Loading Plugin: PhaseDetector...
Loading Plugin: PulsePalOutput...
Loading Plugin: RecordControl...
Loading Plugin: Rectifier...
Loading Plugin: RhythmNode...
Loading Plugin: SerialInput...
Loading Plugin: SpikeSorter...
Loading Plugin: KWIKFormat...
D:\a\plugin-GUI\plugin-GUI\Source\Processors\PluginManager\PluginManager.cpp:219: Failed to load plugin DLL: DLL Error 0x7e
 DLL Load FAILED
Loading Plugin: neuropixels-pxi...
Loading Plugin: nidaq-plugin...
Loading Plugin: NWBFormat...
D:\a\plugin-GUI\plugin-GUI\Source\Processors\PluginManager\PluginManager.cpp:219: Failed to load plugin DLL: DLL Error 0x7e
 DLL Load FAILED
Loading Plugin: probe-viewer...
Num built in engines: 2
Adding engine: Binary
Adding engine: Open Ephys
Num plugin engines: 0

Loading window bounds.

Num engines: 2
Got creator
Loading processor graph.
Creating from plugin info...
Neuropix-PXI(1)::Neuropix-PXI
[Neuropixels-PXI] Scanning for devices...
[Neuropixels-PXI]   Found 1 device.
[Neuropixels-PXI]   Opened BS on slot 4
[Neuropixels-PXI]     Searching for probes...
[Neuropixels-PXI] ***detectHeadstage failed w/ error code: 33
[Neuropixels-PXI] ***detectHeadstage failed w/ error code: 33
[Neuropixels-PXI] ***detectHeadstage failed w/ error code: 33
[Neuropixels-PXI] ***detectHeadstage failed w/ error code: 33
[Neuropixels-PXI]     Found 0 probes.
[Neuropixels-PXI] Checking for V1 basestations...
[Neuropixels-PXI]   Found V1 Basestation
[Neuropixels-PXI]   Opened BSv1 on slot ♦
[Neuropixels-PXI] openProbe: Port: ☺ errorCode: 0
configure serdes failed with NO_LOCK after 5 retries
[Neuropixels-PXI] openProbe: Port: ☻ errorCode: 24
[Neuropixels-PXI] Check if cable is connected properly!
configure serdes failed with NO_LOCK after 5 retries
[Neuropixels-PXI] openProbe: Port: ♥ errorCode: 24
[Neuropixels-PXI] Check if cable is connected properly!
configure serdes failed with NO_LOCK after 5 retries
[Neuropixels-PXI] openProbe: Port: ♦ errorCode: 24
[Neuropixels-PXI] Check if cable is connected properly!
[Neuropixels-PXI] Found 1 probe.
[Neuropixels-PXI] Probe (slot=4, port=1) CH=384 SR=30000 Hz
Creating from plugin info...
(-1)::Merger
[Neuropixels-PXI] Configuring probe...
Setting source node A.
Creating from plugin info...
(-1)::Record Node
Got creator
Creating from plugin info...
LFP viewer(1)::LFP Viewer
Creating from plugin info...
LFP viewer(1)::LFP Viewer
Creating from plugin info...
LFP viewer(1)::LFP Viewer
Creating from plugin info...
NI-DAQmx(1)::NI-DAQmx
Device Category: 15858
Product Name: PXIe-6341
Product Num: 1948959940
Serial Num: 32545325
PXI1Slot2/ai0 - Terminal Config: 7
PXI1Slot2/ai1 - Terminal Config: 7
PXI1Slot2/ai2 - Terminal Config: 7
PXI1Slot2/ai3 - Terminal Config: 7
PXI1Slot2/ai4 - Terminal Config: 7
PXI1Slot2/ai5 - Terminal Config: 7
PXI1Slot2/ai6 - Terminal Config: 7
PXI1Slot2/ai7 - Terminal Config: 7
Found digital inputs:
PXI1Slot2/port0/line0
PXI1Slot2/port0/line1
PXI1Slot2/port0/line2
PXI1Slot2/port0/line3
PXI1Slot2/port0/line4
PXI1Slot2/port0/line5
PXI1Slot2/port0/line6
PXI1Slot2/port0/line7
Min sample rate: 0.02
Max single channel sample rate: 500000.00
Max multi channel sample rate: 62500.00
Got source type for input 0: 0
Got source type for input 1: 0
Got source type for input 2: 0
Got source type for input 3: 0
Got source type for input 4: 0
Got source type for input 5: 0
Got source type for input 6: 0
Got source type for input 7: 0
Restoring parameters for each processor...
Setting Merger source A to 100
Setting source node A.
Setting Merger source B to 102
Setting source node B.
Neuropix-PXI updating settings.
Neuropix-PXI setting num outputs to 768
Loading parameters for Neuropix-PXI
Neuropix-PXI updating settings.
Neuropix-PXI setting num outputs to 768
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 768
Loading parameters for Merger
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 768
Record Node updating settings.
Loading parameters for Record Node
Record Node updating settings.
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 768
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Loading parameters for LFP Viewer
TOTAL HEIGHT = 15360
LfpDisplay found 0
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 768
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 768
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Loading parameters for LFP Viewer
TOTAL HEIGHT = 15360
LfpDisplay found 1
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 768
[Neuropixels-PXI] init: slot: ♦ port: ☺ errorCode: 0
[Neuropixels-PXI] setOPMODE: slot: ♦ port: ☺ errorCode: 0
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
[Neuropixels-PXI] setHSLed: slot: ♦ port: ☺ errorCode: 0
[Neuropixels-PXI] ADC file: C:\Program Files\Open Ephys 0.5.3\CalibrationInfo\19108317172\19108317172_ADCCalibration.csv
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 768
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
[Neuropixels-PXI] Successful ADC calibration.
[Neuropixels-PXI] Gain file: C:\Program Files\Open Ephys 0.5.3\CalibrationInfo\19108317172\19108317172_gainCalValues.csv
Loading parameters for LFP Viewer
TOTAL HEIGHT = 15360
[Neuropixels-PXI] Successful gain calibration.
LfpDisplay found 2
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 768
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
NI-DAQmx updating settings.
NI-DAQmx setting num outputs to 8
Loading parameters for NI-DAQmx
Device Category: 15858
Product Name: PXIe-6341
Product Num: 1948959940
Serial Num: 32545325
PXI1Slot2/ai0 - Terminal Config: 7
PXI1Slot2/ai1 - Terminal Config: 7
PXI1Slot2/ai2 - Terminal Config: 7
PXI1Slot2/ai3 - Terminal Config: 7
PXI1Slot2/ai4 - Terminal Config: 7
PXI1Slot2/ai5 - Terminal Config: 7
PXI1Slot2/ai6 - Terminal Config: 7
PXI1Slot2/ai7 - Terminal Config: 7
Found digital inputs:
PXI1Slot2/port0/line0
PXI1Slot2/port0/line1
PXI1Slot2/port0/line2
PXI1Slot2/port0/line3
PXI1Slot2/port0/line4
PXI1Slot2/port0/line5
PXI1Slot2/port0/line6
PXI1Slot2/port0/line7
Min sample rate: 0.02
Max single channel sample rate: 500000.00
Max multi channel sample rate: 62500.00
Device Category: 15858
Product Name: PXIe-6341
Product Num: 1948959940
Serial Num: 32545325
PXI1Slot2/ai0 - Terminal Config: 7
PXI1Slot2/ai1 - Terminal Config: 7
PXI1Slot2/ai2 - Terminal Config: 7
PXI1Slot2/ai3 - Terminal Config: 7
PXI1Slot2/ai4 - Terminal Config: 7
PXI1Slot2/ai5 - Terminal Config: 7
PXI1Slot2/ai6 - Terminal Config: 7
PXI1Slot2/ai7 - Terminal Config: 7
Found digital inputs:
PXI1Slot2/port0/line0
PXI1Slot2/port0/line1
PXI1Slot2/port0/line2
PXI1Slot2/port0/line3
PXI1Slot2/port0/line4
PXI1Slot2/port0/line5
PXI1Slot2/port0/line6
PXI1Slot2/port0/line7
Min sample rate: 0.02
Max single channel sample rate: 500000.00
Max multi channel sample rate: 62500.00
Device Category: 15858
Product Name: PXIe-6341
Product Num: 1948959940
Serial Num: 32545325
PXI1Slot2/ai0 - Terminal Config: 7
PXI1Slot2/ai1 - Terminal Config: 7
PXI1Slot2/ai2 - Terminal Config: 7
PXI1Slot2/ai3 - Terminal Config: 7
PXI1Slot2/ai4 - Terminal Config: 7
PXI1Slot2/ai5 - Terminal Config: 7
PXI1Slot2/ai6 - Terminal Config: 7
PXI1Slot2/ai7 - Terminal Config: 7
Found digital inputs:
PXI1Slot2/port0/line0
PXI1Slot2/port0/line1
PXI1Slot2/port0/line2
PXI1Slot2/port0/line3
PXI1Slot2/port0/line4
PXI1Slot2/port0/line5
PXI1Slot2/port0/line6
PXI1Slot2/port0/line7
Min sample rate: 0.02
Max single channel sample rate: 500000.00
Max multi channel sample rate: 62500.00
Got source type for input 0: 0
Got source type for input 1: 0
Got source type for input 2: 0
Got source type for input 3: 0
Got source type for input 4: 0
Got source type for input 5: 0
Got source type for input 6: 0
Got source type for input 7: 0
NI-DAQmx updating settings.
NI-DAQmx setting num outputs to 8
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 776
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 776
Record Node updating settings.
Record Node updating settings.
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
[Neuropixels-PXI] Successfully wrote probe config
invalid channel: 191. selection of reference electrode 191/575/959 not allowed
[Neuropixels-PXI] Updating electrode settings for slot: 4 port: 1 dock: 0
[Neuropixels-PXI] ************WRITE PROBE CONFIGURATION****************
[Neuropixels-PXI] AP Gain: 500
[Neuropixels-PXI] LFP Gain: 250
[Neuropixels-PXI] REF: Ext
[Neuropixels-PXI] FILTER: 1
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Got creator
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Setting saturation warning to to 6389
Trigger source: -1
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Setting saturation warning to to 6389
Trigger source: -1
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Setting saturation warning to to 6389
Trigger source: -1
NI-DAQmx updating settings.
NI-DAQmx setting num outputs to 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 776
Saving lfp display params
Saving lfp display params
Saving lfp display params
Record Node updating settings.
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
NI-DAQmx updating settings.
NI-DAQmx setting num outputs to 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 776
Saving lfp display params
Saving lfp display params
Saving lfp display params
Record Node updating settings.
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 776
Re-setting num inputs on LfpDisplayNode to 384
Sample rate = 30000
Adding channel 384 for event source node 6553600
Adding channel 385 for event source node 6553601
Adding channel 386 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 384
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 384
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Tab button clicked: 0
Setting subprocessor to 2
Sample Rate: 2500
LfpDisplayNode setting subprocessor to 6553601
Setting subprocessor to 3
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6684672
TOTAL HEIGHT = 320
Tab button clicked: 1
Tab button clicked: 0
Set master: {100,0}
[Neuropixels-PXI] Setting slot ♦ sync frequency to 1 Hz...
[Neuropixels-PXI] Setting sendSyncAsContinuousChannel to: 1
[Neuropixels-PXI] Probe (slot=4, port=1) CH=385 SR=30000 Hz
Neuropix-PXI updating settings.
Neuropix-PXI setting num outputs to 770
Saving lfp display params
Saving lfp display params
Saving lfp display params
Merger updating settings.
   Merger source A found.
   Merger source B found.
Number of merger outputs: 778
Saving lfp display params
Saving lfp display params
Saving lfp display params
Record Node updating settings.
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 778
Re-setting num inputs on LfpDisplayNode to 385
Sample rate = 30000
Adding channel 385 for event source node 6553600
Adding channel 386 for event source node 6553601
Adding channel 387 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 385
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 385
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
TOTAL HEIGHT = 15400
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 778
Re-setting num inputs on LfpDisplayNode to 385
Sample rate = 2500
Adding channel 385 for event source node 6553600
Adding channel 386 for event source node 6553601
Adding channel 387 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 385
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 385
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
TOTAL HEIGHT = 15400
Saving lfp display params
Saving lfp display params
Saving lfp display params
LFP Viewer updating settings.
Setting num inputs on LfpDisplayNode to 778
Re-setting num inputs on LfpDisplayNode to 8
Sample rate = 30000
Adding channel 8 for event source node 6553600
Adding channel 9 for event source node 6553601
Adding channel 10 for event source node 6684672
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 385
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 385
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
Saving lfp display params
Saving lfp display params
Saving lfp display params
Setting subprocessor to 1
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6553600
Setting subprocessor to 2
Sample Rate: 2500
LfpDisplayNode setting subprocessor to 6553601
Setting subprocessor to 3
Sample Rate: 30000
LfpDisplayNode setting subprocessor to 6684672
Signal chain: 0

Source node: Neuropix-PXI.

Source node: Record Node.

Source node: LFP Viewer.

Source node: LFP Viewer.

Source node: LFP Viewer.
     No dest node.

Signal chain: 1

Source node: NI-DAQmx.

Record Node 103 has already been connected.
     Connecting Neuropix-PXI 100
              to Record Node 103
     Connecting NI-DAQmx 102
              to Record Node 103
     Connecting LFP Viewer 104
              to LFP Viewer 105
     Connecting Record Node 103
              to LFP Viewer 104
     Connecting LFP Viewer 105
              to LFP Viewer 106
Enabling processors...
Source node received enable signal
Record node 103 connected to Message Center
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 385
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 385
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
   Enabling VisualizerEditor
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 385
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 385
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
   Enabling VisualizerEditor
Resizing buffer for Subprocessor 6553600. Samples: 300000, Inputs: 385
Resizing buffer for Subprocessor 6553601. Samples: 25000, Inputs: 385
Resizing buffer for Subprocessor 6684672. Samples: 300000, Inputs: 8
   Enabling VisualizerEditor
Source node received enable signal
Adding audio callback.
[Neuropixels-PXI]   Starting thread.


Pavel Kulik

unread,
Jan 27, 2021, 2:04:18 PM1/27/21
to Open Ephys
Hi Anan,

I've just released a new version of the Neuropixels-PXI plugin (0.3.1) that fixes the SMA output setting for V1 basestations. 
Can you try installing this version via PluginInstaller and see if that fixes the issue?

Thanks,

Pavel

Anan moran

unread,
Jan 28, 2021, 5:08:36 PM1/28/21
to Open Ephys
Hi Pavel,
Your fix did the trick. It is working correctly.
Thanks a lot for the quick fix. Much appreciated .