Hello!
I’ve conducted several LFP recordings from the rat hippocampus and some recordings show a rhythmic artifact around every 400 milliseconds (the distance between artifacts appears to be very stable). The artifact resembles a large negative amplitude deflection, which can also be seen in the spike viewer within the OpenEphys GUI, with a characteristic waveform. The artifact repeatedly occurs for several minutes apparently randomly, i.e irrespective of the animals movement (during sleep and exploration), and in all four LFP channels but in some animals also in EEG channels, though sometimes channels vary in amplitude.
I’ve attached screenshots of the artifact and a power-spectrum over one entire raw recording (duration: 4 hours) between 0 and 300 Hz as well as a power-spectrum over NREM sleep epochs between 0 and 35 Hz (corrected for dF/F).
Animals are connected to one RHD 32-channel headstage, connected to a standard SPI interface cable. The SPI cable is connected to a custom-made commutator (30 $ conductive slip ring, MST005-22-12A), which is connected to a 1.8m standard SPI cable, which is connected to the Open Ephys Acquisition Board. The overall cable, thus, runs ~2.8 m. I’ve used Version 0.5.5 of the OpenEphys to conduct the recordings.
1. My best guess is that the artifact might be caused by the commutator (weird signal degradation?), or relatedly, that data transmission between the Headstage and Acquisition Board is faultily, leading to the rhythmic artifact. Has anybody experienced such an artifact or has any idea what the issue might be?
2. We’ve used the custom-made commutator, because it has a hole in the center, through which optic fibers can be run for optogenetic stimulation (I’ve attached a optical rotary joint above the commutator to release the tension of all cables). Has anybody good suggestions for a combined optical + ephys commutator, that we could try alternatively?
Thanks,
Max
--
You received this message because you are subscribed to the Google Groups "Open Ephys" group.
To unsubscribe from this group and stop receiving emails from it, send an email to open-ephys+...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/open-ephys/5edb8ff8-1831-4f0d-b3c6-1ebcffc47d0an%40googlegroups.com.
Hi Jonathan,
Thanks for your message!
Indeed the power supply of the laser is incredibly noisy. I’ve already spent quite some time to ground the laser and the box in which the animals is placed, which improved the signal a lot (but you can still see the power spikes in 50 Hz + harmonics in the power spectra). The reason why I think that it is not an electrical artifact from the room, though, is that I previously recorded animals in the same room configuration with a Neuralynx Digital Lynx SX Amplifier + analogue MillMax Headstage and did not see any similar artifacts. I found this thread (https://open-ephys.atlassian.net/wiki/spaces/OEW/pages/491644/Commutators), in which Rodrigo reported on weird LFP step–like chopped signals, which were solved by reducing the cabling the Acquisition Board and commutator. Rodrigos' artifacts were not rhythmic, however, and the signal looked less prototypical than the artifacts I’m experiencing. I’m wondering whether the grounding or power supply of the headstage is somehow interrupted periodically, leading to the large negative deflection and subsequent slower return to baseline of the signal… I’m not sure whether this is possible, though. I couldn’t find any information on the transmission frequency between the headstage and the Acquisition Board, which I could compare to the rhythm in which the artifact occurs. What do you think?
However, I will try your suggestions and let you know, if I can get rid of the artifact.
Best,
Max
To view this discussion on the web visit https://groups.google.com/d/msgid/open-ephys/fc5d81f0-5589-486f-a196-cf50a4eabac7n%40googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/open-ephys/a3fd2ee3-3dd0-4ba6-8e9b-d2ca87b6b352n%40googlegroups.com.