Hi, Everyone,Did you ever meet the problem that some channels are not working? In my situation, the channel 2 is strange. When the power is on, the noise of the channel 2 can be reduced. When I input sine signal from functional generator, the channel 2 shows nothing. However, the problem will disappear sometimes. I checked the connection of the cables, nothing found. Is there anything I can try to solve the problem?
--
You received this message because you are subscribed to the Google Groups "NeuroRighter Users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to neurorighter-us...@googlegroups.com.
To post to this group, send email to neurorigh...@googlegroups.com.
Visit this group at http://groups.google.com/group/neurorighter-users.
For more options, visit https://groups.google.com/d/optout.
This is very common and is likely due to a poor contact between the amplifier's gold pin and the MEA contact pad. Two things to try:1. Clean the MEA contact pads using a small amount of isopropyl alcohol with a cotton swab.2. When the MEA is in the amplifier, gently nudge the contact pad laterally (not up and down) using the wooden end of the cotton swab.
On Sat, Mar 21, 2015 at 11:28 PM, Reinhard <laos...@gmail.com> wrote:
Hi, Everyone,Did you ever meet the problem that some channels are not working? In my situation, the channel 2 is strange. When the power is on, the noise of the channel 2 can be reduced. When I input sine signal from functional generator, the channel 2 shows nothing. However, the problem will disappear sometimes. I checked the connection of the cables, nothing found. Is there anything I can try to solve the problem?
--
You received this message because you are subscribed to the Google Groups "NeuroRighter Users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to neurorighter-users+unsub...@googlegroups.com.
Maybe so. Are you storing your amplifier in an incubator? If so, it must not be at 100% relative humidity or you can start to see electrical anomalies like this. The incubator should be kept down around 65% relative humidity. Additionally, it could be caused at the interface between the amplified signal and the NI breakout boards.
On Fri, Mar 27, 2015 at 4:50 AM, Reinhard <laos...@gmail.com> wrote:
Thank you, Jon. I cleaned the pad and pins, it is better now. However, some certain channels will still be abnormal sometimes even I changed the MEA, I think it's caused by the pins of the amplifier.
On Monday, March 23, 2015 at 12:18:59 AM UTC+8, Jon Newman wrote:
This is very common and is likely due to a poor contact between the amplifier's gold pin and the MEA contact pad. Two things to try:1. Clean the MEA contact pads using a small amount of isopropyl alcohol with a cotton swab.2. When the MEA is in the amplifier, gently nudge the contact pad laterally (not up and down) using the wooden end of the cotton swab.
On Sat, Mar 21, 2015 at 11:28 PM, Reinhard <laos...@gmail.com> wrote:
Hi, Everyone,Did you ever meet the problem that some channels are not working? In my situation, the channel 2 is strange. When the power is on, the noise of the channel 2 can be reduced. When I input sine signal from functional generator, the channel 2 shows nothing. However, the problem will disappear sometimes. I checked the connection of the cables, nothing found. Is there anything I can try to solve the problem?
--
You received this message because you are subscribed to the Google Groups "NeuroRighter Users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to neurorighter-users+unsubscribe@googlegroups.com.
>>>> an email to neurorighter-users+unsub...@googlegroups.com.
>>>> To post to this group, send email to neurorigh...@googlegroups.com.
>>>> Visit this group at http://groups.google.com/group/neurorighter-users.
>>>> For more options, visit https://groups.google.com/d/optout.
>>>
>>>
>
> --
> You received this message because you are subscribed to the Google Groups
> "NeuroRighter Users" group.
> To unsubscribe from this group and stop receiving emails from it, send an
> email to neurorighter-users+unsub...@googlegroups.com.
To unsubscribe from this group and stop receiving emails from it, send an email to neurorighter-us...@googlegroups.com.
To unsubscribe from this group and stop receiving emails from it, send an email to neurorighter-users+unsub...@googlegroups.com.