I have download the driver instrument for hp8110a and it send me:
"The error -1073807342 occurred with hp8110a Initialize "

"Error -1073807342 occurred at an unidentified location

Possible reason(s):

VISA: (Hex 0xBFFF0012) Invalid resource reference specified. Parsing
error."

what it means? please answer me in english or in french, but answer
me.

thank you

Dennis Knutson

unread,
Oct 18, 2004, 11:33:23 AM10/18/04
to
Check to see if the VISA Resource Name on the VI that you're calling
matches the actual address of the instrument. The default of GPIB::10
may be different than what the instrument is set to. You can go into
MAX, right click on the GPIB board and select scan for instruments.
This will list all instruments found and their addresses. On the VI,
you can click the little arrow on the right side of the control and
change the resource name to any that are found.

JO

unread,
Oct 18, 2004, 12:10:10 PM10/18/04
to
I have make you said but the adress is good.
I don't know if the channel is good, how can I known that?
And it said now:
"VISA Open in HP81110A Initialize.vi->HP81110A Getting Started.vi"
thank you

JO

unread,
Oct 19, 2004, 4:04:16 AM10/19/04
to
When I try the apllication example, everything is good, but with
getting started the visa is bad. what's it means?

Dennis Knutson

unread,
Oct 19, 2004, 9:07:42 AM10/19/04
to
The only difference is that the getting started calls the initialize
function. Does the initialize return an error? If so, what is it? What
version of LabVIEW and VISA are you using?

JO

unread,
Oct 21, 2004, 7:58:58 AM10/21/04
to
Yes the initialise return en error, this Error : -1073807342, and I
use Labview 7.1 and I am not sure but I think I use VISA32. It is
good?

Dennis Knutson

unread,
Oct 21, 2004, 10:44:53 AM10/21/04
to
If you installed the version of VISA that comes with 7.1, then you've
got a current version. What I would suggest is running the initialize
VI with various combinations of the inputs ID Query and Reset On and
Off. That would help to isolate the problem. If you still get an error
with both set to false, then the error is coming from the VISA Open. I
don't know why you would get an error from VISA Open and the other
example would work. With current versions of VISA, you don't have to
call VISA Open. A VISA session will be automatically created so you
don't have to explicitly do it. Let me know what you find.

JO

unread,
Oct 22, 2004, 11:34:18 AM10/22/04
to
The error in the Initialize box come from the bundle in the structure
case True/False "Query Instrument ID". It send :
"Error -1074003951 occurred at HP81110A Initialize

Possible reason(s):

LabVIEW: (Hex 0xBFFC0011) Identification query failed. "
and when I change the green constant True/False at the bundle in
false, there is no error. It is normal?

Dennis Knutson

unread,
Oct 22, 2004, 2:09:33 PM10/22/04
to
It's not normal but it's not all that unusual. The query instrument ID
issues an *IDN? query to the instrument. The VI is expecting the
string "HP81110A" back. It's possible that, for whatever reason, your
instrument is returning something slightly different. You can probe
the output of the VISA read being done in the instrument query case
and see what is actually being returned. I often skip the query anyway
because I usually don't care. I seldom have problems with the wrong
instrument being in a system.

JO

unread,
Nov 4, 2004, 10:32:04 AM11/4/04
to
Now it is Ok!
I have change the constant :true/false inside ce box

thanks !!